My main point of criticism is, that this CAL? 72 parameter definitely depends on the environmental temperature, which is not mentioned at all in SN-18.
Yes, but it's "easy" to find correlation once TEMP? data is collected together with ACAL cycles before/after recording CAL? data. I run this procedure on somewhat random period, here's plot of CAL?72 for my "rusty" 3458A. Be my guest to find temperature correlation in it
Frankly speaking, I just can't believe, that Illya has 3 different repaired 3458As, which all should have zero T.C. for their LTZ1000A reference, as well as for the U180.. Sorry Illya..
No need to apologize. But you don't see the complete picture. I've repaired more than 8 3458A's, and my own 4 meters, and over went thru over 10 A3's. So far my only
two, not three meters are 0.0x ppm/K TC, rest are more usual 0.2-0.3 ppm/K. I think we have over 20 meters in our little xDevs chat team to collect data from, if you'd like to join our talk.
I believe we are going over this again, just like number of previous diluted 3458A threads
. I do checks almost every month on my 3458s, because I cannot believe 0.0x ppm/K TC myself too, perhaps. Here's July data over 3 days, with two excursions from +23 +/-2°C to +29 +/-2°C. Ambient room temperature is black line (monitor by BME280).
DSV-source file with all points. DSV file also have TEMP? readings for all meters, which are not shown on the plot above. All equipment state in lab is unchanged (to remove doubts that heat gradients during test changed due to other gear around).
All seven meters are connected willy-nilly with low thermal cables to AC-powered Fluke 732B reference (calibrated vs PJVS at May 31, 2019).
Legend:
3458-2 - my first "rusty" meter, replaced A3 ADC 3 times.
3458-3 - my second meter, replaced A3 ADC 1 time to 2016 year one. Meter is modified, have lower DCV noise.
3458-10 - 4th meter, came with 1ppm/day drifty ADC. replaced A3 ADC 2 times.
3458-11 - 3th meter, came with 0.1 ppm/day drifty ADC. installed A3 ADC 1 time.
K2002-6 - one of the 2002's, no modifications. My other 2002 have much better TC, but was busy with other experiment at a time.
D1281 - fresh addition to the team, old but mighty Datronski. TC is so-so.
F8508A - nothing to explain here. Good TC , but worse than my "mythical" 3458s.
Worthy note -
both 3458-2 and 3458-3 have modified A9 reference. -3 unit have LTZ1000CH (not ACH) and 95kOhm S102 in parallel with 15K. -2 unit have original LTZ1000ACH chip, and 100 kOhm VAR BMF in parallel with 15K. -11 and -10 run stock A9's so far, with cooking LTZ temperatures.
There is also second test, with even warmer extreme, with 2 days of data.
DSV-source file with all pointsSame test condition. Test is not highly scientific, proper way would be to keep 732B (or the meters) in the isothermal climate chamber, but I don't have anything big enough to fit 732B, so we make
assumption that my ref meets it's factory specification <0.04 ppm/°C for 10V output. I have no much reason for doubt here, because my previous tests with xDevs FX reference in constant +23.00 °C oven show same results over 3458A/B (-3, and -2 units), so data correlate well.
Also my other past data show evidence that 10V range TC is not 100% relate to A3 ADC only, I have enough units and A3 boards to
prove see that now, but I'm not ready to disclose details about this yet, because this is part of bigger project study. Sorry about that.
I propose to make some more detailed investigations (over temperature) on a specific 3458A, before one claims it to have a 'faulty' A3 board!
While I do not disagree, that testing ultra-high performance DMM/ADC like 3458A can be challenging task, it is not that hard to sort faulty A3 vs good A3. If meter (as a whole) fails meet it 24 / 30 day specifications with ADC A and it meets them with known good ADC B (with guardband margin, let's say 60% just to be sure) - then there is nothing else to discuss, and ADC A is faulty. Yes, there is variation from reference noise, like Dr.Frank correctly highlights, but it's irrelevant even over few weeks data, leave alone a month.
Hence main conclusion is that one cannot just spend a day or two with 3458A and determine if it's faulty or good (as a whole), making all those ebay listings for "pulled from working environment, NIST calibrated" not worth too much. For lab on a budget (he-he) best way is to have known-verified good 3458A as a "reference" unit and do verification of unknown A3 ADC/A9 refs in it, so things can be tested one at a time, instead of trying to figure out if unknown instrument as a whole is good or not.