As attached........comments welcome.
PS. The docs are online but the case is still recorded as open so maybe thats why you couldn't see.
Ian.
Ian, thanks again, it's really interesting / puzzling...
Obviously, your instruments passed Ohm completely after adjustment (As Completed), also this 2W zero error of 1.6 .. 2.4Ohm has been successfully 'removed' by adjustment.
As they had to actuate the front/rear switch for that check, and as the zero reading was ok afterwards, maybe the switch itself might be ok, but the crimping not.. depends a bit on when they checked the zero reading, either directly after adjustment, or afterwards, in a 2nd attempt.
As I already assumed, this error does not affect at all the gain calibration of the Ohm mode, 10k and 100k are spot on to < 1ppm after adjustment.
Same goes for the current ranges, which were "Undetermined" before.
It's a bit strange, why they set status Undetermined in several modes/ranges during As Received, and even after adjustment.. there's no obvious failure.
Frank