What is the magic point in hiding 2823 in S/N number? It's all same for all US-made old HP 3458A's.
Interesting calibration report, few things cought my eye:
* 90 day spec used as verification criteria. I'd rather expect 24 hour spec, however that is harder to reach for unknown history unit
* ANA mode for ACV instead of higher accuracy SYNC.
* AC HF calibration was not performed, based on equipment list.
* Giggle at using Rb for frequency check of 3458A.
As of 10K reading, that's likely not the true one.
Execute OCOMP ON, DELAY 5, OHMF 10E3 , connect 4W, wait for hour or two, and then we see
. Yes, it's slow
.
Thanks for the work,
MasterTech and
Adrian .