Finally getting around to some Space and the very much Harder part some Time to look at the 3458A I scored a FOUR years ago
Like most of us I don't need it and my two Calibrated 6 1/2 digit meters are more than I require.
It had a Cal 110 Error so I had assumed it was going to need a full memory fix and a dirty Cal by me and the wall of bits I have or a $$ trip to Keysight so it went on the shelf. The meter as far as I can find was never Calibrated and given how clean it is inside and still on its 1997 Memory never used or if it was very sparingly
Before pulling boards out to remove memory and add sockets etc I loaded up IanJ's WinGPIB software and actually got a dump of the Cal memory which I thought was curious but was it garbled or trash?
Found in some searching of the Xdevs site the 3458A/KE5FX dump software and managed to get what looks like something resembling data (copy of the text version below).
So on that same page
https://xdevs.com/article/hp3458a_gpib/#manual there is also the Query strings for the Cal Constants which also gave back sensible looking numbers (again below). Still cant get connection expert to recognize the 3458A but it still has the 1997 Firmware on it which might be the problem?
Can someone more familiar with the 3458A take a look at the dump and see if you think it is complete and valid? If so then any idea what might be causing the 110 error (destructive overloads?) or does is simply need a Cal to clear it when the new memory goes in or is the error just related to the likely dead or near dead battery?
Also thoughts on upgrading the Firmware, what would I gain?
TIA