As proposed by Kleinstein i adjusted the preregulator replacing the 6K8 adjustment resistor by 1K8 and the first mains transformer now gets 209 Vac, still in 220 V configuration. All low voltage regulators have their headroom.
I implemented many other mods to the meter received from Ismsanmar, including Pomona copper binding posts at the front and a TFT display. Also the reference board was wired with soldered copper wires replacing board connector. Still want to recap the low voltage supplies and mod the input MUX.
In order to test that R6581T i connected it to our ADR1399 and a Keithley 3706. The log at 6 second intervals (100 NPLC with Autozero + 0.76 seconds for R6581T internal temperature measurement) exhibits a point to point standard deviation of about 3.5E-8. In other words: about 250 nVrms noise at this timing (roughly 0.05 Hz).
Internal temperature of the R6581T is about 39 °C at 20 °C ambient. Ambient temperature coefficient is 0.06 ppm/K for the R6581T, yet there are additional deviations depending on speed of temperature change. About 2 uV shift after turning on heaters with a 2 °C ambient temperature rise within 30 minutes.
Image shows one small R6581T malfunction observed during two days of logging, reason unknown (K3706 noise larger than deviation).
Regards, Dieter