Offline MiDi

  • Regular Contributor
  • *
  • Posts: 197
  • Country: de
3458A Worklog
« on: January 07, 2019, 12:41:27 pm »
This unit is equipped with option 002 and was received with "RAM TEST 1 LOW" error.
It was first powered on on Nov 24th, on Dec 1st it was taken apart for further inspection and cleaning.
Besides a bit of dust on fan & filter it looked as it came straight out of factory.
The datecodes on the ics reveales a manufacturing date from late '89 to early '90.
Remarkable that the calram still seems to be in good condition and maintaining caldata after nearly 30 years.
Obviously it is high time for replacement of the 3 Dallas NVSRAMs - to save the calram a Linux-Tool for GPIB was used.

Next thing to check was heart of instrument: condition of adc (A3-board)
Manual data for cal 72 & 175 was captured between Nov 26th and Dec 9th, since then the acal dc data was captured by script every 5min.
The first two weeks were not promising: the SN18A check revealed drift of > 0.43ppm/day.
It appeared to become a member of A3-sickness club, but fortunately the drift went down to sub 0.01ppm/day for last week  :-+
The adc shows a TC of ~0.33ppm, that is below the stated 0.4ppm in HP Journal p. 13.
This seems to be in spec and shows why it is a good idea to do regular acal for precision measurements even in the 10V range - if the lab temperature is not rock stable (which home lab is?).
In conclusion the adc seems to be in reasonable condition after a month of settling and ~8ppm total drift.

In the meantime nvsram, caps and mains filter were ordered.
Late on Dec 27th it went down for replacement of nvsrams and update to latest FW.
This took a bit longer than estimated - as it revealed that uv light for e.g. developing pcbs (UVB) is not good for erasing eproms (UVC)  :palm:, but after 24h exposure all bits where cleared and newest FW version 9.2 could be programmed successfully.
After this the device turned on and showed a convergence error once, after reset everything was fine and on several power cycles no error came up anymore.

Things to do:
-ordering special caps (radial 8200µF & big 15mF) and fan
-replacement of electrolytics, mains filter & fan
-investigate input leakage, noise, stability, TC etc.

probably: reduce LTZ1000 temperature set point with parallel 100k to reduce drift even more
 
The following users thanked this post: Inverted18650

Offline MiDi

  • Regular Contributor
  • *
  • Posts: 197
  • Country: de
3458A Worklog: Part 2
« Reply #1 on: January 07, 2019, 12:41:57 pm »
On January, 25th the unit went down for another refurbishment.
It was planned to replace the caps, mains filter and fan, but only caps were replaced for now - explanation below.
There is an overview awailable at bottom with all original and replacement parts that were used - it can be easy copied to favorite spreadsheet program.

Replacing all electrolytic capacitors was straight forward and there were no surprises.
A thick copper wire was used to desolder both pins of the caps at the same time - desolder guns are overrated.
The old caps were quite in good condition as a quick check with LCR-tester revealed.
Just a bit more ESR and loss than replacements - would be nothing to worry about for now, but worth replacement for a couple of bucks for good sleep next years.
A4 C09 & A6 C8 were replaced with next higher value available for best fit - as these caps have wide tolerance, a bit more capacitance would not do any harm.
The hidden yellow C17 on A4 was missed, but will be replaced next time.
The visual check on A6 catched attention on the two 2.2nF Y caps C15, C16 near mains fuse, they look cracked like spider app.
They were dropped for now and are replaced next time - there are already Y caps in mains filter, so this should not be an issue.





Mains filter from Schaffner has no good reputation due to several fails, so better to replace it.
It was decided to do it with remaining things next time, as it would be good idea to drill the rivets with closed case from outside to keep the chips out.
Afterwards the case has to be opened again for replacement and all remaining work will be done in one step next time.

The fan introduced quite some vibrations and has this annoying metal sound similar to old noisy hard drives.
It is supplied with unusual 15V and it was intended to be replaced with 24V type that is rated down to 12V.
When comparing both fans it seems that the chosen one would be reasonable with similar current draw and airflow.
As it turned out there is no standard for the pitch and size for mounting holes of "standard" fans and the screws did hold the replacement fan only slightly.
It is more silent and introduces no significant vibration, but a slightly mounted fan seems no good idea.  :--
For now there came no option to mind how to solve this, but it was not intended to put the old fan back without any improvement.
There was fitted a bend copper wire into the slots for balancing and after some fiddling the best position and weight was found to minimize introduced vibration.
Maybe it will be replaced in future...



Another odd thing is a bend top cap for LTZ1000.
Not shure if this is critical and should be fixed, any recommendation?



ADC-Drift test is going on and gave unpleasant surprises, although it is still in spec.



Overview of replacement parts:

A4 - inguard power supply
part #descriptionpitch [mm]manufacturerreplacementmanufacturer
C07, C082,200µF 35V 85°C 18x35mm SMC7.5Nippon Chemi-Con2,200µF 35V 105°C 16x25mm FR Type APanasonic
C0915,000µF 16V 85°C 35x26mm SME10Nippon Chemi-Con18,000µF 16V 105°C 30x30mm LGUNichicon
C1722µF 15V 85°C ?x?mm3,5?22µF 50V 105°C 5x11mmPanasonic

A5 - outguard controller
part #descriptionpitch [mm]manufacturerreplacementmanufacturer
U121, U122DS1235YW-150 NVSRAM28PDallasDS1230Y-120+ on precision socketMaxim
U132DS1220Y-150 NVSRAM24PDallasDS1220AD-150+ on precision socketMaxim


A6 - outguard power supply
part #descriptionpitch [mm]manufacturerreplacementmanufacturer
C1, C11680uF 50V 85°C 16x22mm SXE7.5Nippon Chemi-Con680µF 63V 105°C 16x25mm LXZ (max 25mm - height limited by case)Europe Chemi-Con
C88,300µF 35V 85°C 25x67mm 53D70Sprague10,000µF 35V 85°C 25x61mm TVXNichicon
C200, C20147µF 63V 85°C 8x15mm SME3,5Nippon Chemi-Con47µF 63V 105°C 8x11,5mm EB Type APanasonic
C202330µF 25V 85°C 10x16mm SME5Nippon Chemi-Con330µF 25V 105°C 8x11,5mm FR Type APanasonic
C203220µF 16V 85°C 8x15mm SME3,5Nippon Chemi-Con220µF 35V 105°C 8x11,5mm FR Type APanasonic
C15, C162,2nF 250V~ Y2 40/085/56 3,5x13x7mm PME 27110RIFA2,2nF 300V~ Y2 55/105/56 4x13x9,5mm MKPWIMA

Case
part #descriptionpitch [mm]manufacturerreplacementmanufacturer
Fan60x60x25mm 12V 0.7W powered with 15VPapst60x60x25mm 24V 2.88W 109R0624D4011 (12-27.6V)Sanyo Denki
Mains Filter3A 47nF 2x750µH FN 32340Schaffner6A 47nF 2x800µH KFSSchurter
« Last Edit: February 12, 2019, 10:13:42 am by MiDi »
 

Offline MiDi

  • Regular Contributor
  • *
  • Posts: 197
  • Country: de
Re: 3458A Worklog
« Reply #2 on: January 07, 2019, 12:42:28 pm »
placeholder
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 3743
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: 3458A Worklog
« Reply #3 on: January 07, 2019, 03:31:02 pm »
Welcome to the club.  :-+
Now you need SR104, 732A/B/C and second 3458A.... rabbit hole only starts here...  :popcorn:

Quote
The first two weeks were not promising: the SN18A check revealed drift of > 0.43ppm/day.
It appeared to become a member of A3-sickness club, but fortunately the drift went down to sub 0.01ppm/day for last week  :-+
I would hold the breathe, as sometimes bad A3's give you little tease, before they go back driftin. Don't ask me how I know...
SN18A testing should be done weeks after long offline storage anyway, so you are sure data is valid, and not just initial settling of the unit (which often takes few weeks of 24/7 operation).

Quote
The adc shows a TC of ~0.33ppm, that is below the stated 0.4ppm in HP Journal p. 13.
Pretty horrid, if that TC is actually of the unit, and not the measurement setup. I'd say good happy 3458A should be <0.1ppm TC on main 10V range (no ACAL of course).
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 
The following users thanked this post: MiDi

Online Dr. Frank

  • Super Contributor
  • ***
  • Posts: 1550
  • Country: de
Re: 3458A Worklog
« Reply #4 on: January 07, 2019, 06:40:45 pm »


Quote
The adc shows a TC of ~0.33ppm, that is below the stated 0.4ppm in HP Journal p. 13.
Pretty horrid, if that TC is actually of the unit, and not the measurement setup. I'd say good happy 3458A should be <0.1ppm TC on main 10V range (no ACAL of course).

Hello TiN,

how do you measure that on your 3458As?

Per specification, about 0.5ppm/°C is the T.C.  w/o ACAL, that is mostly the R-ladder matching T.C. of U180, and 0.15ppm/°C is with ACAL, i.e. latter is mostly the LTZ1000As T.C.

I've determined about 0.23ppm/°C for CAL? 72 (i.e. purely U180).
Did an ACAL directly after power-up, in cold state, that extrapolation fits nicely with statistical data.
Combined T.C. of my box is about 0.4ppm/°C.

I think, MiDis 3458A is fine as well.

Frank

« Last Edit: January 07, 2019, 06:54:46 pm by Dr. Frank »
 
The following users thanked this post: MiDi

Offline TiN

  • Super Contributor
  • ***
  • Posts: 3743
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: 3458A Worklog
« Reply #5 on: January 08, 2019, 02:26:50 am »
Quote
Hello TiN,

how do you measure that on your 3458As?

Well, I remember we discussed this before at EEVBlog. Perhaps it's time we create separate thread "DMM temperature stability comparison project", not to pollute this fine thread?

I was interested in total tempco of the DMM (that is from terminals to the output data, end-to-end, rather than specific 3458A's component contributor) so I did not use CAL? 72 data to evaluate tempco but rather variated ambient temperature (aircon on, about +18-20c, aircon off about +30c ambient), while keeping stable aged LTZ1000 reference in isothermal chamber (+/-0.01c at all times). Resulting data (don't recall graph links right now, but they are somewhere public) show me that both of my 3458A's had no visible tempco (below 0.1ppm that is, down it the noise). After that test I stopped doing silly ACAL DCV every 1C of TEMP? change. I did repeat same test few more times, with equal result.

I'm willing to test this again once new A3 board for my 3458C arrives, so we can use more meters in the mix (and I still have 8508 to test too).
In fact I have somewhat indicative plot, just quick check of K2002 warmup stability from 1 week cold state. But temperature changed too, so you can also see invisible tempco of my reference 3458 units (STD and B box, green lines).

So that's why I say good 3458 should be 0.1 ppm/K or better, even though official spec (old and not updated) is 0.5ppm/K.
My standalone HP 3458A A9 (old STD version) reference also have tempco <0.02ppm/K over 24-50c range, so A9 contribution to tempco can be much much smaller.
During my own modules testing/assembly all LTZ references that fall outside of <0.05 ppm/K box at 18c-50c I consider as fail  ::).
« Last Edit: January 08, 2019, 02:30:52 am by TiN »
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 
The following users thanked this post: MiDi

Offline MiDi

  • Regular Contributor
  • *
  • Posts: 197
  • Country: de
Re: 3458A Worklog
« Reply #6 on: January 08, 2019, 03:51:52 am »
Well, I remember we discussed this before at EEVBlog. Perhaps it's time we create separate thread "DMM temperature stability comparison project", not to pollute this fine thread?
As there are placeholders and I appreciate all your input and feedback, it would be no problem for me to "pollute" this thread - far from it!

These valuable information backed my feeling, that this unit might be on the high side for adc tc.
Perhaps the newer adcs have improved tc/stability?
Collecting autocal data will go on for several weeks - lets see, how stability evolves.
For now I have a bit of confidence, as it shows a quite smooth settling curve.
 

Online Dr. Frank

  • Super Contributor
  • ***
  • Posts: 1550
  • Country: de
Re: 3458A Worklog
« Reply #7 on: January 08, 2019, 04:19:59 am »
Maybe we should really open another thread about the different T.C. specifications and HW versions of the 3458A.

I was not aware, that TiNs 3 DMMs all should have such low T.C.s, and we did not discuss that in detail, either, if I recall that correctly.
But I also can hardly believe that, for some reasons.

The LTZ1000 references inside the 3458A have never been trimmed for lowest T.C. by hp, as far as I can tell, even up today, in the 'new' version of the 34470A.
I really doubt that hp has ever selected these references for zero T.C., either.

Then, the T.C. of U180 is the sum of about 6..8 different resistors, and even if these all are sitting on the same hybrid / chip, I really doubt that such ultra low T.C.s are really systematically possible.I really doubt, that  KS has really redesigned U180 after 2000. Probably they just live from a last time build of this hybrid...as they live from bodges and All Time Buy from other terminated components..

I also really doubt that they ever had budget, resources and brains to make any fundamental redesign / improvement on the 3458A in the last 20 years, apart from these slight compatible changes due to terminated components, basically A5 only.

The only obvious component change regarding T.C. was the 40k reference resistor, which had been changed from the old Vishay technologies with about 1..2 ppm/K typ., non -oil filled to the VHP101, about 0.2 ppm/K, oil filled type, somewhere around 1995.

In the end, I think that the old specifiaction from 1988, which is still valid, applies to the old instruments, before 1996 at least, but obviously also to my instrument from about 2000, apart from OHM TC and stability, maybe.

So to say, that T.C. > 0.1ppm/K is nOK, is to my opinion not correct.

Maybe TiN really has unbelievable luck in getting 3 A3 ADCs with exceptional low T.C., multiplied by the unbelievably low T.C. of his 3458As  LTZ references, or agilent really has changed U180 to better technology in the last years, or maybe we should really define a reliable method of determination of the different T.C. parameters of the 3458A.

MiDi, your instrument really seems to be be ok, and fine.. don't expect too low a T.C., even the more stable 8508A has 0.3ppm/K as best T.C. only, and that really sounds much more realistic.

Frank
« Last Edit: January 08, 2019, 04:38:27 am by Dr. Frank »
 
The following users thanked this post: AG7CK

Online sokoloff

  • Frequent Contributor
  • **
  • Posts: 931
  • Country: us
Re: 3458A Worklog
« Reply #8 on: January 08, 2019, 04:56:36 am »
Maybe we should really open another thread about the different T.C. specifications and HW versions of the 3458A.
I think there's value in that; I also think there's value in a basic "so you bought a good bench meter and want to know where it stands; here's the first 5 things to do and why..." with sub-threads of "here's how to do that on model XYZ".
 
The following users thanked this post: MiDi

Offline TiN

  • Super Contributor
  • ***
  • Posts: 3743
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: 3458A Worklog
« Reply #9 on: January 08, 2019, 05:02:18 am »
Okay, let's discuss here.
I think it will benefit us all to get down this rabbit hole.

Perhaps I got bit rude, using horrid word for 0.3ppm/K TC. It's just my level of acceptance on 3458A is high after all that effort put into units. Official spec on 3458A 10V range is 0.5ppm/K without help of ACAL
« Last Edit: January 08, 2019, 05:39:43 am by TiN »
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 
The following users thanked this post: MiDi

Offline AG7CK

  • Regular Contributor
  • *
  • Posts: 100
  • Country: th
Re: 3458A Worklog
« Reply #10 on: January 08, 2019, 02:55:54 pm »
...
 the drift went down to sub 0.01ppm/day for last week
...

If I am not mistaken, this means that all digits on the meter are steady. For instance readings varying between [9.999 999 9] and [10.000 000 1] flickers +-0.01ppm.

I have never seen a 8.5 digit meter showing 7 or 10 volt DC without flickering in the last (at least 1 or 2) digits in any video from any member here or elsewhere. It would be interesting if you could explain and show how you do your "sub 0.01 ppm" measurements.
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 3743
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: 3458A Worklog
« Reply #11 on: January 08, 2019, 03:01:04 pm »
0.01 ppm/day is not meter measurement, but the meter ADC gain stability. It does not include measurement noise.

Sent from my Lenovo K900_ROW using Tapatalk

YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
 
The following users thanked this post: engiadina, Inverted18650

Offline MiDi

  • Regular Contributor
  • *
  • Posts: 197
  • Country: de
Re: 3458A Worklog
« Reply #12 on: January 27, 2019, 04:21:09 am »
Short update: managed replacement of elkos on friday, details coming soon.

For now, only two things I stumbled over:

The two Y mains caps C15, C16 next to mains fuse look weird and I decided to drop them for now until replacement arrives.
There are Y caps inside mains filter, so this should not be an issue.






The second thing is a bend cap for LTZ1000, not shure if the gap is an issue  :-//

« Last Edit: January 27, 2019, 04:23:21 am by MiDi »
 

Offline MiDi

  • Regular Contributor
  • *
  • Posts: 197
  • Country: de
3458A Worklog: Part 2
« Reply #13 on: February 12, 2019, 10:26:26 am »
Took a bit longer than estimated, but now here is part 2.
 
The following users thanked this post: Andreas

Offline TiN

  • Super Contributor
  • ***
  • Posts: 3743
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Re: 3458A Worklog
« Reply #14 on: February 12, 2019, 11:36:07 am »
Nice :)