Thanks for your help...
Difficult for me to take action on the devices, because they are only used partly and have to be returned close in time back to production

Today I had a 2nd 2558 on desk and there also the 100mV run out of spec. (but here not needed in production enviroment)
All other ACV are within spec... also ACI are in good condition.
Only have to check the 400Hz ACI on hi amps again, because frequency stable reference leaved for calibration

But I have more to come... so (if other work allows it) I maybe will have the chance to get at least one of them full functinal within spec and then get the others piece by piece fully back... but it will take time.
You know... managment decisions... "is it needed with all the effort or does partly calibration is ok in this case?" and you know the answer if they request needed time to get them full up.