After buying one 34970A unit here on the forum, and
attempting to replace the "jumpy" LM399H reference with a nice low noise ADR1399 (which actually increased the noise at 1NPLC 2-3 times), followed by a sort of an investigation to make some sense of how the DCV noise depends on the reference (mostly unsuccessful, with some possible clues), followed by purchasing two more 34970A (one on eBay, as spares and one again here), I have now compared all three for a DCV noise with a rather strange results. The first unit (Agilent #1) is now very steady after the reference replacement, no jumps and fairly low noise, the second unit (Agilent #2, from eBay) is fully working with a very similar noise levels to the first (only the display is almost dead), however the #3, Keysight, almost new looking (made in 2017, with an original calibration certificate still available from Keysight), exhibits an even stranger noise pattern, with 1NPLC noise similar to other two units (and no jumps) but a considerably higher noise (x1,5 - x2.5) on longer integration periods (10 and 100NPLC). I plan to investigate this at some point (not urgently though) and this thread is to exchange information/ideas/clues on the reasong for such a strange DCV noise behaviour and perhaps how to fix/improve it.
The table shows the DCV noise (SD, in ppm of the range) for 1, 10, 100NPLC and all DCV ranges, all 3 units, 0V input.
The graph is a representative of 10NPLC measurements, 100V Range, 60sec for all 3 units.
Cheers
Alex