Author Topic: A story of one Datron Wavetek 4708  (Read 4502 times)
Offline wolfy007
Re: A story of one Datron Wavetek 4708
« Reply #25 on: November 22, 2018, 06:06:45 am »
WOW, well this article just helped me locate my Fail 1 error on my just arrived 4808 quickly. Was looking through schematics while googling 4808 and here we are! Thanks Doktor Pyta.
Will still go through and check/change caps.
Tek: 222A, 212, TAS465, TDS460A, 2465, 2465B, 2467, TM504, TM506, SG504, SG503, PG506, TG501, SC504, PS503a, FG501a, DM504a, PG501, 7603+7L12
BWD: 835, 539d, 521, 504, 141, 160a  Fluke: 8060A, 8024B, 8025A, 27/FM, 87, 87III   HP: 5315A, 3478a, 3468A, 3457a  Keithley: 196, 228a  Solartron: 7081 7151
Offline doktor pyta
    • RF Scientific
« Reply #26 on: November 22, 2018, 08:17:25 am »
@Johnny10
The noise level with the new fan is slightly lower but there is more important factor- due to worn out bearing old fan generated weird disturbing sounds, changed pitch etc.
At the moment the sound is definitely less annoying.
« Last Edit: November 22, 2018, 08:30:37 am by doktor pyta »
Offline doktor pyta
    • RF Scientific
« Reply #27 on: November 22, 2018, 08:19:53 am »
@wolfy007
Was it You who bought 4808 which appeared on Ebay two weeks ago (or so) for $5000?
Offline wolfy007
« Reply #28 on: November 22, 2018, 10:27:52 am »
@doktor pyta
Yup Im the mad bastard that purchased it on ebay...
And its looking like more than simply a few caps. I replaced the 4 tants and the problem persisted, themisters appear to be ok, but I have noticed someone before me has been at the amp assembly and the heatsink assemblies, so I guess Im going to have to go over it in a bit more detail.
Online syau
  • Country: hk
« Reply #29 on: June 18, 2019, 03:53:33 pm »
Managed to source one of those Datron cable set (total 3) from evilbay. The cover is made of plastic not metal. The set come with both side connector as of socket type  :wtf: Checking the serial number of the  able confirmed that they are from the same set with part number 40016-A, B, C
Anybody has idea what is the intended use of those cables ?
« Last Edit: June 18, 2019, 04:00:53 pm by syau »
Offline Testtech
« Reply #30 on: June 18, 2019, 04:19:55 pm »
The box/cable plugs in the front of the calibrator, the banana's then to the DMM under test.
I suspect that the Lemo cable may be one of the interconnects for the 4600 transconductance amp. A better picture of the ends would help confirm.
Offline TiN
  • Country: tw
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
« Reply #31 on: June 18, 2019, 04:54:48 pm »
That is cable set for Datron/Wavetek 4950. You can sell it to rare owners of such.
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
The following users thanked this post: syau