Hi, Mr. Mickle.T and Mr.Kleinstein ,
Here comes the trying these day.
1. The INL issue are nearlly the same even with the new cap. Please refer to the attached. There alway have a turn over point around 7-8V input. As I know , the 6581 only calibrate the positive 10V and negative 10/1/0.1V together with the zero point . It seems have no correction if the gain is not flat in the whole 10v range. Which might lead to that? The transfer error for 10v-1v is still in the 12-14uv range.
2. Futher more, I had replaced the U214/DG442 with reed relay, no obvious change viewed . I was worry about the leakage of the DG422 before.
3. I had chaneged the K006 to low EMF type reed relay.
Focusing on the ADC, but have no idea about the next step.
Hi, Tin,
Your INL issue is like mine. Hope to find solution!
Best Regards,
szszjdb