While measuring one of my LTZ references @25°C in a thermal chamber over a longer periode with my R6581D I noticed that there is quite some t.c. visible. Therefor the internal temperature was logged together with the voltage readings. It turned out this t.c. is due to the meter itself and about 0.8ppm/K. A measurement of the t.c. of the internal reference itself revealed, that this t.c. is not due to the reference, but something else. Time to upgrade R6581!
Several improvements are documented here:
http://bbs.1ppm.cn/topic/23/advantest-6581磨机-zy的得意杰作
http://bbs.38hot.net/forum.php?mod=viewthread&tid=28698&highlight=6581It just so happened that pipelie organized a group buy to replace R234 which boosts the zener voltage to 10V and R200 as part of the voltage to current converter.
R234 original: Alpha Electroncis SLD (7k34 X F 10k0)
R234 replacement: Vishay VHD200, hermetically sealed
R200 original: Alpha Electronics SS103F20k00 XF
R200 replacement: Vishay 300193ZT
The resistors arrived on friday so the weekend was used to replace them. Unfortunately there is no drop in replacement for R200, so the leads of the 300193ZT needs to be bend slightly into the correct position, while the VHD200 fit perfectly.
Now that the replacement is done it‘s time to let the meter settle with the new components installed, which is said to be about a month and see how this modification turns out.
Not to hide anything, removing the networks is a real pita and needs careful work and a lot of solder to heat up all pins equally. The vias are also very small and hard to clean. So unless you are well experienced in such work I would not recommend it, as you can damage a lot.
I use drill needles to finally remove last solder pieces inside the vias while heating the pad.
Also note that after that modification the meter needs calibration. To be continued ...
-branadic-