Hello branadic,
That way, I can't extract no reasonable information.
these graphs show, that there is a gain (adjustment) error between the 3458A and the 4000A, as well between all three instruments.
So for an INL graph, you would need to normalize these gain (and offset) errors first, to get the INL over U.
What was your intention, in the end?
To determine the INLs of the 4000A and the R6581, assuming very low INL for the 3458A?
For the extraction of T.C.s that's also not so good, as all three T.C.s interfere, so your running T.C. measurement on the 3458A probably will let you solve for the other T.C.s
A wider temperature span like +/- 5 ..10 °C would have been better, I guess.
So a nice experiment, when you calculate the INLs correctly.
Frank