Until you have some history you are limited to just published specifications.
Did you take into account the thermal EMF caused by the material contacts?
I would avoid having your UoM written as 0.0000938000V, first, there needs to be a space between the number and the V. But I would refer it as 93.8 µV but you need round to two digits so 94 µV.
I just looked at my UoM for my Transmille and I have 7.5 µV at 100 mV. I will double-check this once I have cooked dinner.
EDIT:
Now that dinner has been dealt with.
Taking assumption,
9.3 µV/V + 0.8 µV for 1yr 95% spec.
0.01 µV Resolution
0.0316 µV Repeatabilty
1.2 µV to cover 0.3 µV/°C over a ± 2 °C swing of temp
I don't know your Imported UoM so I can take a stab at 4.9 µV/V from
Fluke's Schedule in the UKI get an Uncertainty of 2 µV at 100 mV. But it would be best to quote yours at FS, so 2.2 µV.
Remember it is fine having a low number but you have to prove you can achieve this and maintain it. So it is advisable to add a bit extra at first just so you have room to move if it starts to drift out.
Start doing a monthly cross-check between the multimeter and calibrator. Having some data will help you.