It was suggested by other forum participants that I start a new thread to discuss specs and features of AC & DC voltage references that I make and sell on eBay (as SQWARREL. Sample Link
https://www.ebay.com/itm/285789548856) where I kept things brief and non-specific. So, here is my attempt to do so. I appreciate your input and suggestions as this thread (and my device improvements) develop.
My intent is to make and sell very low-priced but USEFUL references for those of us who are not yet ready to resort to costly lab calibration. We may want to be able to perform basic, "ballpark" function checks of our DMM's to see if they're even working correctly or reasonably holding calibration. And, we can't justify spending $150 or more for higher quality devices like the DMMCheck+ and others.
AC Voltage Reference: Here's a device that generates a relatively pure, NON-FLUCTUATING, adjustable sine wave from 0 to 6VACrms @ 100Hz (fixed frequency) for those who do not have a quality signal generator or alternative. (100Hz was chosen as a DMM-friendly "sweet spot" for avoiding EMI-influence in the working environment.)
Remember, you can't do credible AC comparisons or checks by using isolation/step-down transformers, variacs or by sticking your probes into a mains socket because the line voltage is always jumping around and the sine waves don't look anything like sine waves because of all the noise and distortion.
The AC references can be pre-calibrated (buyer's choice) to between 100mV and 6VACrms @ 100Hz +/- 0.5% to cover near-full-scale readings of most popular DMM's, requiring only a single 12VDC regulated power supply or battery because they have an on-PCB ICL7662 inverter for the -12V rail.
There can be several variations (and prices) for such a device including lower sine wave amplitudes which allow using single supply op-amps and 9V batteries, etc., etc., but I would rather focus on minimum performance SPECIFICATIONS for device USEFULNESS at this time.
DC Voltage Reference: Rather than the untrustworthy cheap references, many using crappy components and USED AD586 parts and phony calibration stickers, I have developed a straightforward 10VDC reference based upon the LT1236 chip.
This DC reference currently uses a 14VAC 0.5A wall-wart transformer or other suitable 14-16VAC power supply to the input jack leading to an on-PCB 15VDC regulated power supply that then feeds the LT1236 circuitry. The power input can also be 18-24VDC through the same input jack and that also offers autopolarity protection due to the full wave bridge rectifier at the input on-PCB.
The LT1236 is trimmed to exactly 10.0000VDC. How does this happen? I use our brand new SDM3055 5.5-digit DMM (new production, direct from Siglent) which is "comparison calibrated" to a new DMM6500 that I have access to. Then, the SDM3055 is the basis for "comparison calibrating" my Fluke 87V's.
The DC reference trimming is done with the SDM3055 by powering on both for 24-48 hours before final trimming to 10.0000 which is as good as it gets until we buy a DMM6500 or similar in the near future.
However, after trimming one of the 10V references 2 months ago and keeping it powered on, and after re-checking the SDM3055 vs DMM6500 readings weekly, the 10VDC reference STILL reads exactly 10.0000, spot on, if that's worth anything.
Now, I realize that this whole discussion doesn't meet any volt-nut standards but, remember why we're here in the first place, as I set forth in my introduction.
I simply would like to improve the SPECIFICATIONS and USEFULNESS of these devices while keeping them low-cost to give the average DMM owner an alternative to spending hundreds of dollars and more on devices that may be of higher quality and accuracy but are of no more usefulness.
Thanks for any insight and help that comes to mind. A few pix attached for general reference...
Modify message