I'm not sure if this belongs more to repair or metrology but anyways..
34401A failed on AC volts calibration at 50hz with really big error (1% low on all voltage ranges)
Adjustment routine runs fine until adjustment point at 10Vac 10hz give error 733 "AC low frequency convergence failed"
Manual wont tell more than that about error 733.
Any ideas or has someone actually stumbled on the same problem?
Oddly the VAC range measures ok-ish at 10Hz with less than 0.1% error and 1kHz is spot on.
There is a dip around 50Hz on all ranges and it is about 1% of reading.