Author Topic: Chasing PPMs? Put that cell phone away!  (Read 7089 times)
Online CalibrationGuyTopic starter
« on: May 29, 2024, 01:43:14 pm »
So I wanted to take a few pictures of our 732 array, and was reminded of something I would tell all my lab techs. PUT the stupid phone away! Here's what happened to a measurement I was doing to test a newly acquired 732a with our most stable reference (also a 732a).
Those peaks occurred when I moved close to take a picture of the "flatness" of the test. So yes, these measurements can be affected by your (and your cell phone's) presence.
Picture...
TomG.
« Last Edit: May 29, 2024, 01:45:24 pm by CalibrationGuy »
The following users thanked this post: boB, dietert1
Offline Andreas
Re: Chasing PPMs? Put that cell phone away!
« Reply #1 on: May 29, 2024, 03:03:50 pm »
Hello,
Not only mobile phones. Also (even unused) USB-cables near measurement setup
can be a source of unexpected measurement deviations. E.g. here:
https://www.eevblog.com/forum/metrology/lowest-drift-lowest-noise-voltage-reference/msg3738565/#msg3738565
with best regards
Andreas
The following users thanked this post: CalibrationGuy
Offline EC8010
« Reply #2 on: May 29, 2024, 04:38:20 pm »
And switch-mode power supplies to laptops. I've twice chased red herrings before remembering to unplug the laptop's power supply.
Online ITArchitect
« Reply #3 on: May 29, 2024, 05:54:38 pm »
Ah, chasing noise can be fun sometimes.  I found a similar issue with EMI coming from a wireless phone charging stand.  Pulled my favorites scope apart thinking I had issues.  Since then, all mobile devices have been banned from the lab.  |O :scared:
Offline mendip_discovery
« Reply #4 on: May 29, 2024, 07:51:42 pm »
What would be these PPMs you be talking of? Surely you mean µV/V.
My MD was harping on that I should be doing TikTok videos, this in front of the 17025 assessor. He didn't notice my comment about TikTok users being mostly kids and creepy old men. I feel the need to draw 4 red perpendicular lines, some with blue in and some with transparent ink.
Motorcyclist, Nerd, and I work in a Calibration Lab :-)
So everyone is clear, Calibration = Taking Measurement against a known source, Verification = Checking Calibration against Specification, Adjustment = Adjusting the unit to be within specifications.
« Reply #5 on: May 29, 2024, 08:44:49 pm »
Yes, of course, the joke being that a part per million in the 10Vdc, or 5Vdc references that we are building and/or testing is a tangible thing that can be acquired and/or kept! Hence calling them ppms in this context.
TomG.
Offline exe
  • self-educated hobbyist
« Reply #6 on: May 29, 2024, 09:15:36 pm »
Is this because of the phone, or your presence? Would it make a difference if you put the phone in flight mode?
The following users thanked this post: MiDi
« Reply #7 on: May 29, 2024, 09:26:48 pm »
If you use shielded cables, your presence does not usually affect the measurement unless you start touching things. Then, all bets are off. Yes, when the cell phone reaches out via radio or bluetooth, etc., that will be a larger disturbance than the EMI emitted by their new, powerful CPUs and also the inverters powering the displays, etc. It's just easier to leave them in the next room. Remember, the interference decreases by the square of the distance.
TomG.
Online dietert1
  • Country: br
    • CADT Homepage
« Reply #8 on: May 29, 2024, 09:54:48 pm »
The Fluke 732A is a design from a past when cellular phones weren't that abundant. One could even call it a legacy device. Would be interesting to which degree their revisions  (B and C) solved the problem.
« Reply #9 on: May 29, 2024, 10:09:10 pm »
Perhaps a savings grace is the fact that all the old stuff was built like a tank with lots of metal encasing everything, which would help somewhat. I'm not going to put my phone near my 732a reference gear to find out, though!
TomG.
Online David Hess
  • DavidH
« Reply #10 on: May 30, 2024, 12:28:04 am »
Test equipment used to be available with an "EMI" option for extra shielding.
In extreme cases, I have used bulkhead connectors and coaxial cable to keep RF out of sensitive signals.
Offline edpalmer42
« Reply #11 on: May 30, 2024, 01:05:00 am »
When I'm making sensitive measurements I always use double-shielded cable - preferably braid & foil rather than double braid.
I also collect the data from the equipment via GPIB to a computer that I access remotely via VNC.  I never go closer than about 2M from the equipment.
Offline Andreas
« Reply #12 on: May 30, 2024, 05:20:08 am »
The Fluke 732B specifies <0.18V/m as limit in the manual. See screen shot attached here:
https://www.eevblog.com/forum/metrology/emi-measurements-of-a-volt-nut/msg2684070/#msg2684070
A cell phone can generate easily 100-300 V/m when you put it into a cookies box and call it from outside.
I recommend metal shields (tinned steel) laying on the desk to shield somewhat the environment from the connection lines.
with best regards
Andreas
« Last Edit: May 30, 2024, 05:28:59 am by Andreas »
« Reply #13 on: May 30, 2024, 05:41:14 am »
with best regards
Andreas
I have an ISO17025 accreditation. I bring this up because our auditor was old school and a stickler for detail. When I showed him the 732a and the Fluke 5520a, he asked me to verify the EMI field strength and demonstrate that it was below Fluke recommendations. We had to get a new meter to prove that the EMI was below the threshold. Many labs have these meters precisely because you can never fully know every source without some means of detection. Also, if you rent space like we do, you don't always know what's in the walls and above the ceiling tiles.
TomG.
Online David Hess
  • DavidH
« Reply #14 on: May 30, 2024, 06:22:05 am »
I use double shielded cable when making sensitive receiver measurements, or RF will leak around attenuators, but have not needed it to keep EMI out of DC measurements.  I still often use it because the connector fastenings are more reliable.
Online Kleinstein
« Reply #15 on: May 30, 2024, 06:38:03 am »
TomG.
With cell phones  and Wifi there is now more EMI from higher frequencies. A simple shield can work well for the lower frequencies, but it only needs a relatively small slit to let the higher frequencies in. Old style shielding may not be made for the modern high frequency environment. Modern PCBs tend to be multilyer with a ground plane - this can effectively act like a shield too, just look less impressive.
Cell phones can from time to time send rather powerfull pulses (e.g. to synchronize / reconnect to base)  - enough to be audible with some audio gear. A measurement may well miss those rare pulses that only happen a few times per day.
Whatching out for EMI problems is defintely a good idea. In this context I wonder, why there is no common fiber optic computer interface with sensitive instruments.
« Reply #16 on: May 30, 2024, 07:45:31 am »
@Kleinstein: I built a poor man's fiber optic interface from a project I saw in I think Radio Electronics (yes, I'm that old). It was RS232 to infrared led through a fiber optic cable back to RS232.
Also, you are correct. Despite the massive metal enclosures, these devices also had big slots everywhere to dissipate the heat.
TomG.
Offline EC8010
« Reply #17 on: May 30, 2024, 08:36:05 am »
With cell phones  and Wifi there is now more EMI from higher frequencies. A simple shield can work well for the lower frequencies, but it only needs a relatively small slit to let the higher frequencies in. Old style shielding may not be made for the modern high frequency environment. Modern PCBs tend to be multilayer with a ground plane - this can effectively act like a shield too, just look less impressive.
I repeatedly failed EMC compliance testing due to a 0.1mm wide slit that let 400-600MHz in. Once spotted, a tiny piece of copper foil inside fixed the problem. Self-adhesive copper tape (sold cheaply to gardeners for deterring slugs) is your friend; put it over all gaps. I now do all my testing (mostly audio frequencies) inside earthed biscuit or tobacco tins - if it's good enough to keep food fresh it's good for EMC. Beware that many tins are epoxy coated and need an explicit electrical bond between lid and tin. External connections are made via chassis-mounting BNCs. And that includes power. When I'm really nervous, the power comes from 12V lead-acid batteries also in earthed biscuit tins. All my BNC cables have a clamp-on ferrite at each end to assist common mode rejection. Screening close to the DUT or the source works best, hence the use of ground planes.
I've done a number of experiments that responded to my body heat in the room and to draughts.
And if you're really unlucky, light is a problem. It took me four redesigns before I was able to measure the I-V relationship of an amber LED down to single-digit fA because ambient light easily causes an LED to produce nA of current. The only thing that stops light is metal. Black PVC tape etc attenuates but doesn't stop. Insulated BNCs let light in. Try putting a powerful torch on one side of a thick wooden door - the results will surprise.
The following users thanked this post: Echo88, exe, pdenisowski
Offline pdenisowski
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
« Reply #18 on: May 30, 2024, 12:12:17 pm »
This was a big issue with GSM, but I haven't noticed audio artifacts with 3G and higher cellular technologies .... although that could be my hearing getting worse with each new generation of cellular technology  :)
For anyone who's either nostalgic or too young to have ever heard it, an audio sample is attached.
Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8
Offline HighVoltage
« Reply #19 on: May 30, 2024, 12:34:54 pm »
Best way I found to see if there is any unwanted disturbance, hook up a Keithley 617 electrometer and measure a 100 G \$\Omega\$ resistor. If that measurement is stable, your lab is clean.
The worst disturbance I had once was from a neighbor that had a broken power supply in his server rack.
https://www.eevblog.com/forum/chat/43-mhz-burst-noise-every-63us-in-the-lab-(how-to-find-the-source)/
The following users thanked this post: MiDi, MegaVolt
Online tggzzz
« Reply #20 on: May 30, 2024, 02:21:06 pm »
You will get artefacts with all pulsed radio systems, i.e. non-CDMA systems. It will audibly be worst with a TDMA system where the pulses are in the audio band, e.g. GSM at 217Hz.
Equipment has been forced to accommodate such pulses, by better shielding and/or electronically filtering out offending pulses (e.g. a 217Hz notch filter).
Obviously older equipment may be more susceptible, since that threat was never envisaged.
Online dietert1
  • Country: br
    • CADT Homepage
« Reply #21 on: June 01, 2024, 07:28:11 am »
Just saw an ebay offer with "Advantest R8240 Digital Elektrometer + Omron Z3GB optical GP-IB Link". The fiber isolator seems to exist as a product, yet rare. Don't know how they get the supply but remember that i once modded a GPIB socket, using one of the Gnd pins to take +5 V for an adapter.
Regards, Dieter
« Last Edit: June 01, 2024, 07:32:48 am by dietert1 »
« Reply #22 on: June 06, 2024, 02:38:16 am »
Yes, I see that they exist. If you have the money, here's one: https://www.artisantg.com/TestMeasurement/84017-1/National-Instruments-GPIB-140-Fiber-Optic-GPIB-Extender OOF, the price starts at $2,300, I think.
TomG.
Online dietert1
  • Country: br
    • CADT Homepage
« Reply #23 on: June 06, 2024, 10:09:19 pm »
Is that the price for a pair of those?
No, one should make that. e. g. using two small FPGA boards and two fiber transceiver modules. Except the GPIB bus hardware interface will be a bit nonstandard to make the connection simulate a bus cable. It is possible, see the isolating I2C drivers from TI. They also have open collector outputs on both sides.
Regards, Dieter
« Reply #24 on: June 06, 2024, 10:40:45 pm »
The price is for a pair of transceivers and fiber optic cables, but still.
TomG.
