Bonjour, Tim Fox and PhysicFan: A few notes please:
0/ Tim, the 4 term RLC meters generally are used with both sides floating but the low side terminals are near ground. For HP 4332A, both terminals must float.
1/ The smaller EL meters indeed have two floating terminals. The large units 20-100KV and up must have a ground due to a metal case.
2/ We used the Sensitive Research Electrostatic VM in applications of measuring 12 kV Avionic CRT power supply, with 100 uA current capability. The leakage was never an issue.
3/ The large meters were for testing arc lamp ignitors up to 55 kV of RF bursts 100k - 10 MHz. Again leakage was not relevant, only capacitance.
4/ Ultra high resistance needs a special bridges and guarded electrodes. all of these apply a high test voltage, the GenRad to 1000V
We have several methods,
GenRad Megohm Bridges and meters series 1862, 1863, 1844A
https://www.ietlabs.com/pdf/Manuals/1644_im.pdfDanbridge (Danish) JP30A 30 kV insulation tester, up to 30 kV, 1 uA FS indications can read ~ 2-10 nA least)
Lab HVPS, with Keithley elecrometer set on current range in series at return end of DUT.
5/ The Sensitive Research devices are excellent BUT dust, dirt film and exposure to humidity naturally degrade the leakage.
We cleaned with 99% iospropyl alcohol every insulator before every measurement, had room humidity below 40% , used Teflon or similar insulated wires and jumpers.
6/ All electrometer voltmeters are VERY delicate, move with care, store upright and NEVER drop even a few inches! CAREFULLY LEVEL AND ZERO ADJUST BEFORE EVER MEASUREMENT!
I hope this info and experience can be interesting and benefit your work.
Bon Chance,
Jon