Author Topic: ASR-103  (Read 2404 times)
Offline maxampTopic starter
ASR-103
« on: December 06, 2024, 05:53:35 am »
Has anyone used or done a tear down of a ASR-1**?  I can't find anything about it in EEVblog...
Mouser store link to ASR-103 $6,204.55:
https://www.mouser.com/ProductDetail/Alpha-Electronics/ASR-103?qs=hWgE7mdIu5R25LS1yp8SdQ%3D%3D&srsltid=AfmBOoolc1-lgZ9fvI4DQWtX7C0wGJsJcDNH76tcIUk59h36N9-eJ48-
±3 ppm per year
Paper about the evaluation of a ASR-103:
https://static1.squarespace.com/static/5525d010e4b00afb743a318d/t/55bbd5a8e4b002b451be6a92/1438373288134/ASR-103-report-NMi.pdf
Seems like you would be better off putting a Y607110K0000T9L in a box or spending the money on a SR104.
Digi key Y607110K0000T9L $80.71:
https://www.digikey.com/en/products/detail/vpg-foil-resistors/Y607110K0000T9L/6029609
Shelf life stability: 2 ppm for at least 6 years
(unaffected by humidity)
SR104 $7,728.00
https://www.ietlabs.com/10-000-ohm-primary-resistance-standard.html
Highest accuracy (1 ppm/yr) and stability (±0.5 ppm/yr typical) of any resistance standard
You can get the 1K version for the low low price of $3,699.00 on ebay:
eBay auction: #204820149899
Online Overspeed
Re: ASR-103
« Reply #1 on: December 06, 2024, 06:58:54 am »
±3 ppm per year
SR104 $7,728.00
Hello
That not rocket science , there is some pictures of similiar products ( other brand on this forum ) generally that a main resistor and one in series / parallel to adjust the value , I have put in the post a link with Xdev DIY resistor that very close to the design except Xdev use a random value and not an adjusted value.
Depend who you are as If you are a big company with budget that not a problem but perhaps that possible to be clever ( estimated prices include shipping )
a nice machined CNC alu enclosure can be in the 700 USD ( CAD / CNC machining / surface treatment ..)
a set of Pomona banana connectors = 5x30 = 150 USD
internal insulation bespoke cut = 100 USD
internal shield = 100 USD
temp sensor well = 50 USD
various parts 25  USD
after the main component as the '' resistor '' is in the 150 x 4 = 600  USD max and can be tested / calibrated  before installation and after installation
final calibration to complete the project a quite costly service following the 2019 NIST prices list
Xdev have interesting things on his website :
https://xdevs.com/article/bsw104/
a DIY transfert low cost from Xdev
https://xdevs.com/article/lcrbox/
BE AWARE of EBAY standard as there is no proof of evidence as the resistor have not been over heated by using wrong current level
Regards
OS
« Last Edit: December 06, 2024, 05:16:30 pm by Overspeed »
The following users thanked this post: maxamp
Online Overspeed
Re: ASR-103
« Reply #2 on: December 06, 2024, 07:52:46 pm »
Hello
I link a test report of the ASR-103
Regards
OS
The following users thanked this post: Alex Nikitin
Online Overspeed
Re: ASR-103
« Reply #3 on: December 07, 2024, 08:45:30 am »
Hello
a Fluke 742 autopsie on Xdev website
https://xdevs.com/article/f742a/
Xdev work quality is impressive
Fluke 742 by using two parallel copper plates ( slotted / drilled ) is a very good design which allow a simple setup and easy DIY without sophisticated tools .
Regards
OS
« Last Edit: December 07, 2024, 10:12:43 am by Overspeed »
Offline MiDi
  • Country: ua
Re: ASR-103
« Reply #4 on: December 07, 2024, 01:24:42 pm »
There are some papers about HRU standard resistors available.
Those have different design than PWW (old SR104, F742A) or BMF (VHP100/200/VHA series, new SR104?).
Seems ASR does not use same technology as HRU/USR-SF, more like classic BMF: Alpha Electronics Standard Resistors
Be careful with DS of VPG resistors, their typical values may be far from what you get IRL: Study of temperature coefficient on 260 precision resistors.
Papers (may be available on a popular open source science hub):
Characterization of 1 kΩ Metal-Foil Standard Resistors and Continuing Drift-Rate Evaluation of 1 Ω and 10 Ω Standard Resistors
Novel 100-Ω Metal Foil Resistor
Characterization of 100-Ω Metal Foil Standard Resistors
Online Overspeed
Re: ASR-103
« Reply #5 on: December 07, 2024, 02:09:29 pm »
Hello
This one is interesting with internal pics
Regards
OS
Re: ASR-103
« Reply #6 on: December 24, 2024, 11:41:19 pm »
Be careful with DS of VPG resistors, their typical values may be far from what you get IRL: https://xdevs.com/article/tcr_test/#vpg2018
I put the VPG Y607110K0000T9L resistor I got from Digi-key on top of a peltier hot/cool plate and a 3458A.  Temperature cycled from 19-27C and got about 0.31ppm/C which is off but close to the 0.2ppm/C spec.  It could be due to my setup rather than the resistor. The xdevs article you linked is more about temperature coefficient rather than long term drift.
The following users thanked this post: Andreas
Offline Andreas
Re: ASR-103
« Reply #7 on: December 25, 2024, 08:29:54 am »
Hello,
What is the X-Axis: time in seconds or number of measurement (@ which NPLC).
For me it looks like there is some lagging effect (resistance reacts delayed to the temperature setpoint)
and/or some ageing drift (start and stop at the same temperature differ somewhat).
So I would:
- stay at the temperature extremes for a longer time. (plateau)
- cycle several times to see eventually occuring ageing drift.
By the way:
after my experience most of the heat transfer inside the resistor is done by the resistor wires.
So the wires should have a good thermal coupling to the temperature block.
Do you have the possibility to also measure the actual temperature value?
with best regards
Andreas
Offline bck
Re: ASR-103
« Reply #8 on: December 25, 2024, 09:15:12 am »
just a guess from my side: the "lag" is him running ACAL after an SP change.
~12min for a full ACAL iirc => about 45min each step
Alex. B.
Offline Andreas
Re: ASR-103
« Reply #9 on: December 25, 2024, 10:57:41 am »
Hello,
that is not what I am seeing:
from lowest temperature between 20 and 22 deg C there is no change in resistance.
Then at 23 deg C a small movement and at 24 deg C a large jump.
So the delay is more than one ACAL cycle.
with best regards
Andreas
Online Kleinstein
Re: ASR-103
« Reply #10 on: December 25, 2024, 12:31:24 pm »
Unless the measurement cycle is really long, it may not be a good idea to add ACAL calls in between. They also add some noise and over a short time the meter is more stable than the ACAL performance. It is still more the choice of no ACAL at all, or ACAL relatively frequent, like 2 x per step.
Re: ASR-103
« Reply #11 on: December 26, 2024, 09:42:38 am »
X-axis is time in seconds.  I'm taking an ACAL with the temperature change which flat lines the readings.  I put the temperature sensor on the plate, put the resistor leads over the plate (rather than to the side), switched to 8508 to avoid ACALS  and 1 hour per step.  I'll post results when it's ready.  Thanks for the feedback.
Online aronake
  • Country: hk
Re: ASR-103
« Reply #12 on: December 26, 2024, 10:47:50 am »
Could you show a picture of you "plate setup"? If it is what it sounds like, a peltier with heatsink on one side and you just put the resistor on top of that in open air, i could see this being a big source of unreliable measurements. Better to build a box that the peltier are cooling. There are lots of such build projects posted here.
Re: ASR-103
« Reply #13 on: December 26, 2024, 05:53:32 pm »
I attached what I have measured so far.  Keep in mind I started everything cold.  It now looks more like 0.83ppm/C.
I got this really cheap off ebay.  Its built like a tank.  I'm sending it serial commands.  It only has a set resolution of 1C.
https://torreypinesscientific.com/product/digital-shaking-chilling-heating-dry-baths/
Re: ASR-103
« Reply #14 on: December 26, 2024, 06:21:04 pm »
Yes, I realize something like this would be more optional.  I just haven't put the time into building it yet.
https://www.eevblog.com/forum/metrology/diy-large-thermal-chamber-for-metrology-tm-testing/msg3474756/?topicseen#msg3474756
Offline Andreas
Re: ASR-103
« Reply #15 on: December 26, 2024, 07:40:51 pm »
Looks better now.
the best results on my measurements I got with a setup like here:
https://www.eevblog.com/forum/metrology/t-c-measurements-on-precision-resistors/msg695463/#msg695463
(note the 2 temperature sensors in good contact to the heat spreader near the resistor wires).
with best regards
Andreas
Re: ASR-103
« Reply #16 on: December 28, 2024, 07:25:11 am »
Here is the longer run data.  Linear fit is 1.04PPM/C.