I’ve done several videos repairing these units, my experience is that:
- All electrolytic caps have to be replaced.
- The plastic bolts holding the +/-15V regulators to the heat sinks degrade and break, so the regulators over heat and cause the 15V rails to turn off and on when they get too hot.
- The two round bridge rectifiers often fail and need to be replaced.
- ICs need to be reseated in the sockets to freshen up the connections.
- Tantalum caps like to randomly blow and short out the 15v rails (always replace with 35v, not 25v)
- The wires on the 5V regulator can be badly corroded by the flux that was used at manufacture, check it isn’t just about to fall off.
Check voltage rails, more than once, the last unit I did had tantalum caps blowing and keep on shorting the +/- 15V rails out.
You can narrow down faults by unplugging boards, it will run without the resistance, current, GPIB, AC boards plugged in, don’t forget the ribbon cables snd side rail connectors.
There are tantalum caps on the front panel between buttons, never seen those fail but there could always be a first time.
Some units have the front/rear input switching unit mounted in the back which has relays on it, that has tantalums too which I have seen fail.