Hi group,
The first two lines of the Datron 1062 DC calibration call for connecting a 1M \$\Omega\$ resistor in parallel with a 1nF capacitor for the Lin step. A 10M \$\Omega\$ resistor in parallel with a 1nF capacitor is required for the IB step. Both networks are described as shielded and given the Datron part numbers 400391 and 400392 respectively.
I was unable to find any pictures of these fixtures. I had to take an educated guess at the design.
This what I came up with:
(https://www.eevblog.com/forum/metrology/datron-1062-calibration-fixture-400391-and-400392/?action=dlattach;attach=2726189;image)
(https://www.eevblog.com/forum/metrology/datron-1062-calibration-fixture-400391-and-400392/?action=dlattach;attach=2726181;image)
(https://www.eevblog.com/forum/metrology/datron-1062-calibration-fixture-400391-and-400392/?action=dlattach;attach=2726185;image)
I have connected the case and the shielding copper to the Guard terminal.
The networks are housed in a 1 inch length of aluminum extrusion.
Does anybody have the official fixtures?
Regards,
Jay_Diddy_B
Very nice.
I always just used some through hole parts and screwed them down in the terminals, seemed to work fine.