Author Topic: DC calibrators noise 1. Fluke 8520A v HP 3458A  (Read 961 times)

0 Members and 1 Guest are viewing this topic.

Offline jfphpTopic starter

  • Regular Contributor
  • *
  • Posts: 238
DC calibrators noise 1. Fluke 8520A v HP 3458A
« on: May 19, 2020, 11:20:02 am »
I intend to mesure the noise of various DC calibrators. Generaly they are specified in two bands : 0,1 to 10 Hz and 10Hz to 10 Khz but almost all (except Fluke --before the 5730--) manufacturers don't explain anything about specs verification. Other manufacturers (Datron, Wavetek...) speak about average in 1 or 10 line periods (???) and exotic bands (DC 2Hz null detector...) without, too, verification process. I  would like to make a systematic approach of the problem and I have several questions.
1. First question for the lowest band (0,1 to 10 Hz): Fluke 5700 manual as 5440B uses a 8520A to mesure the standard deviation of several mesurements but I don't understand the relation between the sample rate (they use max, 20 per second) and filter set at 1000 ms. User manual is very unclear speaking about analog and digital filtering enabling optimum reject of the mains period.
How can I get the same mesurement with a 3458A ? Standard deviation, no problem, reading rate per second, no problem but filter and NPLC ? And, the last but not the least, how to make the difference between the noise of the 8520A/3458A and the noise of the calibrator ? All DC serious mesurements of the 3458A use 100 NPLC...
« Last Edit: May 19, 2020, 12:33:09 pm by jfphp »
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14210
  • Country: de
Re: DC calibrators noise 1. Fluke 8520A v HP 3458A
« Reply #1 on: May 19, 2020, 03:09:42 pm »
The Fluke 8520 can use an analog filter at the input, while the 3458 does not have this option. So it can not easily  recreate the same noise bandwidth as the 8520 with the filter enabled. If at all this would be in the non-auto zero mode, but this would add extra 1/f noise so it may not be adequate. The additional noise due to the lacking filter would be in the 25 Hz , 2.5 Hz or similar range depending on the PLC setting used.
 

Offline guenthert

  • Frequent Contributor
  • **
  • Posts: 712
  • Country: de
Re: DC calibrators noise 1. Fluke 8520A v HP 3458A
« Reply #2 on: May 19, 2020, 03:48:36 pm »
  I have doubts that those instruments are the best choice to measure noise.  The HP3458A is comparatively fast, but the Fluke 8520A will take its sweet time (~20s!) to measure at best resolution.  Worse, they'll integrate over a long time (100 or more PLC, as you already stated), so some 'equivalent time sampling' technique won't work.  You could add some fast S&H circuity, but how much noise will that introduce?

  What's wrong with the traditional approach (DC blocking high quality capacitor in series with a LNA and a modest, but quick sampling DMM or oscilloscope)? 
« Last Edit: May 19, 2020, 03:50:58 pm by guenthert »
 


Share me