3457A is cheap garbage IMO ;-)
Cheap plastic case, LCD that is hard to read, bad keyboard
Frontend is also weird, any normal DVM has 10GOhm input up to 10V range, like that:
100mV range, AMPx100=> 10V to ADC
1V range, AMPx10=> 10V to ADC
10V directly to ADC
100V div/10 to ADC
1000V div/100 to ADC
On 3457A, 10GOhm input is only up to 3V range and range amplifier/divider goes like that:
30mV AMPx333 => 10V ADC
300mV AMPx33.3 => 10V ADC
3V AMPx3.33 => 10V ADC
30V div/100, AMPx33.3 => 10V ADC
300V div/100, AMPx3.33 => 10V ADC
That's why 30V and 300V range and similar in Ohms mode, are so bad on that one
Not to mention the input goes more or less directly into the range switching hybrid, so if you toast one of the FETs or the range divider resistors in there, you're screwed and need a new hybrid, whereas in 3456A, you just grab your soldering iron and replace the JFET with a J113 or similar.