A little analysis will show that the OP-AMP minimum GBW Product requirement is:
GBW Product > F*(Q-2), where F is frequency of operation and Q is the Virtual Ground Quality Factor (Input Signal Magnitude/Virtual Ground) virtual ground of the input.
Example: A selected frequency of 10KHz and effective virtual ground Quality Factor selected is 10,000 (Vin =10VPP, Virtual Ground = 1mVPP):
GBW Product > 10K(10,000 -2) ~1GHz
Using the mentioned TL072 which has a GBWP of ~3MHz results in measurement frequency of < 300Hz, TLC272 < 200Hz and the ADA4891 <22KHz.
CAUTION, these results are VERY OPTIMISTIC since the GBWP is a
Small Signal parameter, and the
actual Large Signal GBWP is much less, usually less than 1/5 the specified
Small Signal GBWP.
So what does this mean in reality?? That the expected results of a Quality Virtual Ground need to be resolved with an OP-AMP that has a
Power or Large Signal GBWP of at least the numbers presented and since many OP-AMPs don't specify this parameter, one needs to increase the small signal GBWP requirement by at least an additional 5X, safer would be 10X!! This is why we developed the summing node way back in 1980

Best,