Author Topic: DM3068 calibration  (Read 1192 times)
Offline jopl
DM3068 calibration
« on: October 29, 2018, 09:49:21 am »
Dear DM3068 gurus,
doesn't someone of you know, how to calibrate DM3068 multimeter? There is the service guide for DM3000 available, where I can find entering password for calibration, but this is not unfortunately case of DM3068, where is just notice in service guide manual, that DM3068 does not support user calibration ... Is there any suggestion, how I could calibrate multimeter with with help of 5522A Fluke Multi-Product Calibrator, like in other multimeter models from Rigol? Thank you for anybody's reply in advance. jopl
Offline bck
Re: DM3068 calibration
« Reply #1 on: October 29, 2018, 01:05:13 pm »
~ Alexander Becker
Offline jopl
« Reply #2 on: October 29, 2018, 03:10:22 pm »
Dear bck,
thank you for your answer. Unfortunately there is no visible option of password in the "Utility" - "T/C" buttons sequence as well as no options "SecrOff" or "Cal" are there. I though DM3058 and DM3068 are very similar but as I can see now, it is different in this menu option. Or am I wrong or missing something in the beginning, i.e. within start of the system for example? Thx for reply. jopl