Author Topic: DMM does not stabilize when measuring resistance  (Read 10106 times)
Offline pwnellTopic starter
Re: DMM does not stabilize when measuring resistance
« Reply #25 on: December 05, 2021, 01:53:00 am »
Also Z202 max TCR spec is 2.8ppm\K, not 0.2 ppm\K. Don't fall for "typical" specmanship from VPG on these. You cannot expect to have every random resistor meet typical spec. Short term rift of your DMM is also expected to be much bigger than one from resistor itself.
Where do you see that in the data sheet?
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #26 on: December 05, 2021, 02:12:08 am »
Page 1, Table 1. Also my statement above comes from experience with testing hundreds of resistors (both VPG metal foil type, wirewounds and metal films).
I did explain this a bit here.
Offline joeqsmith
« Reply #27 on: December 05, 2021, 02:30:28 am »
Last one is 4 wire
Without anything to show how it was setup, I assume based on:
So here are some graphs - all same test setup:
that you are still mounting the part with the BNC inside the copper and using a 4-wire to the BNC.  All a pure guess as I can't see what you are doing.
« Reply #28 on: December 05, 2021, 04:23:25 am »
Very informative article, thanks.
« Reply #29 on: December 05, 2021, 04:26:14 am »
Sorry, it was a separate post and I used the bnc for the current wires and separate wires to the resistor from the sense terminals for that last graph only. So 4 wire, no copper tube.
Offline HighVoltage
« Reply #30 on: December 05, 2021, 12:02:46 pm »
Picture please .
« Reply #31 on: December 05, 2021, 05:57:21 pm »
Picture please .
Sure - BNC is only used because it is convenient.  BNC carries current, the other two cables are the sense leads connected close to the resistor's body.  Meter is in 4 Wire mode.
Offline mendip_discovery
« Reply #32 on: December 07, 2021, 05:41:14 pm »
I see people have taken into account the errors of the resistors but you also need to be wary of the effects that your meter has.
For example, the specs talks about its accuracy specifications over 24hrs with a room temp kept within ±1°C. So a few of the ppm drift may well be because of the meter settling.
Even the test leads might have a bit of time to adjust.
Be very cautious of thermals.
But experiment, the more you look the deeper you go. Just don't end up like n00py ;-)
Motorcyclist, Nerd, and I work in a Calibration Lab :-)
So everyone is clear, Calibration = Taking Measurement against a known source, Verification = Checking Calibration against Specification, Adjustment = Adjusting the unit to be within specifications.