Author Topic: Electron Plus SPA100 Source Picoammeter / High Resistance Meter  (Read 2494 times)
Offline MegaVolt
  • Country: by
Offline HighVoltage
Re: Electron Plus SPA100 Source Picoammeter / High Resistance Meter
« Reply #1 on: November 26, 2021, 02:14:19 pm »
"<2fA resolution (200pA range)"
That is a bold claim especially on normal BNC cables.
Someone probably needs to buy one and perform a tear down.
Online Kosmic
« Reply #2 on: November 26, 2021, 06:57:44 pm »
Just contact the owner of Electron Plus. He is a member of the forum. Username is fcb https://www.eevblog.com/forum/profile/?u=25995
Offline fcb
    • Electron Plus
« Reply #3 on: November 26, 2021, 08:48:21 pm »
Fairly 'standard' transimpedance amplifier design with careful layout that's gone through a couple of iterations, lots of slots in PCB's etc..
The front-end and the bias generator are isolated with a transformer and opto-couplers.
We are updating our website at the moment (as well as manuals/datasheets/etc...) , so the picture is a little out-of-date, most of our products now have a label on-top and all have stainless steel front & rear panels.  Here's a picture courtesy of our US distributor, Saelig.
The following users thanked this post: TiN, Mickle T., HighVoltage, WN1X, 2N3055, MegaVolt
Offline MegaVolt
  • Country: by
« Reply #4 on: November 30, 2021, 01:21:27 pm »
Perhaps you can answer some questions?
1. BNC instead of triax? Why did you decide not to use them?
2. What was the noise level on the lower range?
3. Any pictures inside?
Offline fcb
    • Electron Plus
« Reply #5 on: November 30, 2021, 02:54:35 pm »
BNC's tend to be used on picoammeters (i.e. Keithley 6485), generally there is no requirement for a guard system - the inner and outer parts of the BNC are typically at the same potential; it can be thought of as essentially zero ohms.
Also cost is a major consideration - BNC's are low-cost and fairly readily available - triax connectors are very expensive (40-100x more), more difficult to obtain, etc..  Two triax connectors would equal the retail price of the SPA100.  I bought some recently for my Keithley 237 SMU, eye-wateringly expensive and that was without the cable.
Noise. On the lower range will depend on sample rate (2/10Hz) a little and sample (rolling) averaging (1x to 64x) - however it tends to be dwarfed by aliased mains hum pick-up.  We have another SPA100 batch finishing production this week, so will post a few screen captures of the o/c noise if you'd like.
Sorry no pix inside.
The following users thanked this post: MegaVolt
Offline bsw_m
  • Country: by
« Reply #6 on: November 30, 2021, 06:51:34 pm »
There is no need to use a triax connector for a transimpedance amplifier. The reason is simple, the potential difference between the ground and the input terminals is practically zero (to be precise equal to the amplifier offset). And this offset doesn't generate tangible leakage current.
But warning in noise that can be generated by not a very high quality insulator.
Offline fcb
    • Electron Plus
« Reply #7 on: November 30, 2021, 07:34:38 pm »
We haven't noticed any noise we can trace back to insulator materials. The insulated 50R BNC's are made from white polypropylene (so without black/carbon fillers).
On the lowest current range you can create a small transient shift in the reading by changing the bias generator output from say -40V to +40V,  but then waving your hand 3 feet away does the same when the nothing is plugged in.
Offline bsw_m
  • Country: by
« Reply #8 on: November 30, 2021, 08:33:34 pm »
This is the simplest way to test whether the electrometer input is still alive.  :-+
The dielectric noise will be lower than the noise of your amplifier. Therefore, it is not noticeable.
P.S.
Unfortunately, I am not able to test this amplifier.
Its parameters and design look very balanced.  :-+
But I don't believe in 1E-15A noise free, only if with a very "evil" filter, with a long settling time.
I think you used a 10GΩ feedback resistor for the low range?  ;)
« Last Edit: November 30, 2021, 08:43:24 pm by bsw_m »
The following users thanked this post: splin
Offline fcb
    • Electron Plus
« Reply #9 on: November 30, 2021, 09:10:53 pm »
P.S.
I don't quite understand the context of the last few lines of your post.
1G/PGA/ADC if that helps  ;)
Offline bsw_m
  • Country: by
« Reply #10 on: November 30, 2021, 10:13:12 pm »
Thanks.
My interest in the resistance of the Rf is only in the context of a rough estimate of the possible noise floor of the amplifier.
Offline bsw_m
  • Country: by
« Reply #11 on: December 01, 2021, 12:34:21 pm »
Measuring current generation after physical impact (connection) and noise. Pay attention to how long it takes to settle (More that 2 hours). Also You can view noise of this triax connector.
Case and inner ring(guard ring) are connected to ground, central pin connected to electrometer input. Direct connection without using a cable. The input connector of the electrometer served as a measuring chamber.
In addition to the graph, I also attach the raw data in the archive.
« Last Edit: December 01, 2021, 12:37:46 pm by bsw_m »
Offline fcb
    • Electron Plus
« Reply #12 on: December 01, 2021, 01:03:46 pm »
Interesting. What picoammeter/electrometer are you using, and what did the test setup look like?
Your data is approx -0.28fA to -0.04fA (40aA!) - what makes you think this is purely noise from the triax insulator and not from elsewhere?
Offline bsw_m
  • Country: by
« Reply #13 on: December 01, 2021, 01:26:59 pm »
For this measurement I'm use my V7-45 electrometer. Self noise of this electrometer in attach.
Also measurement setup looks like at the photo in the attachment.
In these photos, I not connect the ground. As I quickly reassemble it to show how it looked.
I think the above description and the attached graphs and photographs describe why I am confident that I have measured the noise and parasitic current of a standard  triaxial connector with PTFE insulator.
Offline fcb
    • Electron Plus
« Reply #14 on: December 01, 2021, 01:42:42 pm »
Great.  Looks like a beast of instrument.  I'm assuming these are only available used in FSR countries?
I'm assuming the V7-45 is a two box unit.  Any pix of the insides?? ;D
Offline bsw_m
  • Country: by
« Reply #15 on: December 01, 2021, 01:47:26 pm »
Also, if we refer to the table of the maximum permissible errors for  performance verification procedure. We can see the checked point 20aA with a maximum permissible error of 39.4% or ±8aA. (screen in attach).
You right!
Some pics and video You can see in this topic: https://www.eevblog.com/forum/metrology/v7-45-electrometer-made-in-belarus-(welcome-in-attoamps-world)/
And if You have interest, little info about ultra low current and ultra high resistance standard, that able calibrate this electrometer: https://www.eevblog.com/forum/metrology/nk4-1-low-current-and-high-resistance-calibrator-(made-in-belarus)/
« Last Edit: December 01, 2021, 01:52:32 pm by bsw_m »
Offline fcb
    • Electron Plus
« Reply #16 on: December 01, 2021, 01:52:09 pm »
bvi = does this connect to the head?
kop = control/computer?
vb1chod analog vb1y = analog output?
Offline bsw_m
  • Country: by
« Reply #17 on: December 01, 2021, 01:55:37 pm »
this connector for connect external measurement block (yes, head)
- this is GPIB IEEE488.1
Yes.
Offline fcb
    • Electron Plus
« Reply #18 on: December 01, 2021, 02:07:12 pm »
Love those vintage eastern block instruments. Performance looks amazing - did they build an updated unit or did it fissile (pun intended) with the end of the cold war?
So do you have different heads for current and voltage?
Interesting lunch-time reading:
https://ampnuts-com.translate.goog/v7-45/?_x_tr_sl=auto&_x_tr_tl=en&_x_tr_hl=de&_x_tr_pto=nui
Offline bsw_m
  • Country: by
« Reply #19 on: December 01, 2021, 02:21:06 pm »
The updated device was called B7-57 / 2, produced after the collapse of the USSR. Its production was piece and the device was unobtanium.
I'm restore one. Found it in a very poor condition and unfortunately not complete. Some photos "as received" in attach.
No, one head can measure currents, voltages and charges.
« Last Edit: December 01, 2021, 02:25:14 pm by bsw_m »
Offline snik
« Reply #20 on: December 07, 2021, 02:55:10 pm »
Does anyone know what's inside it? Any pictures?
https://electron.plus/product/spa100-source-picoammeter
https://www.ebay.com/itm/124791936727
Here some pictures :
[attach=1]
[attach=2]
[attach=3]
[attach=4]
Didn't use or test it until now.
The following users thanked this post: Kean, HighVoltage, bsw_m, ch_scr, MegaVolt
Offline fcb
    • Electron Plus
« Reply #21 on: December 07, 2021, 03:51:02 pm »
The 'dust' at the corners of the PCB is from opening the enclosure and is a mixture of the blacking of the taptite screws and the aluminium extrusion - it isn't there when assembled.
Worth tapping the empty-box on a desk before reassembling, and brushing the dust off the PCB.