Hello, all. I'm looking into building a rig for experiments in resistance thermometry.
Goals:
Measure resistance ratios of pt100, pt1000, and thermistors
Measure resistance ratios near unity between UUT's and a standard resistor with uncertainties better than 40 ppm (e.g. .004
for a pt100), as well as wider ranges (3:1/1:3 and further)
Multiplex standard and UUT resistances
programmable current for zero-power compensation
Electrical test bench cost <700$, used equipment preferred
PC interface, for fully-automatic control.
Architecture selection:
As much as I'd like to play with one, there is a very small used market for automatic resistance thermometry bridges, and their prices reflect that (isotech f- series, fluke "super-themrometers" etc). Bridges requiring manual dialing of decade boxes to balance a circuit are right out, especially if they require manual lead/impedence/thermovoltage compensation. This leads me to DMM's with some supporting hardware. I'm not going to lie, I'm also a bit biased towards purchasing a metrology DMM, as this will also be my first DMM purchase with more counts than I can do with my hands, and will of course lead into other experiments.
I'm planning to build a current-reversing DC bridge, similar to the fluke 1570a/1595a utilizing a programmable current source , a switching/control unit (like an HP 3488a), and some pi/microcontroller action to tie it together.
Current Source Selection:
I think this will likely be scratch-built possibly tucked away in the 3488a. Absolute value, offset voltages, current leakage, etc will have a very small effect. The driving factors will be noise and very short-term drift (on the order of 1-20s). I will likely leave current reversing to the switching unit.
DMM selection:
The DMM should have excellent INL, as this will be the dominating uncertainty component. I think it would be preferable to have a DMM with built-in 100mV ranging, as I doubt I could build a signal amplifier that outperforms the DMM's internal range. Range should have similar INL when measuring negative voltages, as this will allow compensating for thermal voltages without allowing the UUT to cool off from its self-heating. Unfortunately, most meters in my price range will only specify overall 24h specs, each ratio measurement will actually only occur over 1-20s.
HP 3458a:
If I were a rich man... Sadly I don't think my fiance would be too happy with me taking a loan out on my car for my hobby...
HP 3457a:
pros:
- Native 300mV range allows nominal currents of 1mA over most reasonable temperature measurements
- Has excellent 24h specs, ~30 ppm on 100 mV range
- secret 7th digit
HP 34401a:
pros:
- AD linearity spec (2 ppm rdg + 1 ppm range), with data from others on this forum showing sub-ppm INL in the the native (10 V) range, with a fairly smooth curve
- Transfer spec (15 ppm of 100 mV range for readings within 10% of nominal)
What are your thoughts? are there other units I'm forgetting? data I haven't found yet?