Author Topic: ESI Resistance Standard SR1010 and Standard Resistor SR104  (Read 18309 times)

0 Members and 1 Guest are viewing this topic.

Offline acbern

  • Frequent Contributor
  • **
  • Posts: 316
  • Country: de
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #25 on: January 16, 2016, 04:21:42 am »

I'm confused a bit, which one do you regard as your actual standard? Is it the Vishay VHPxxx?

If you want to have one of your standards calibrated, wouldn't it be better to send the SR104 to the cal lab?


Well, I do have a history with the homebrew one. So I will get that cal'ed to verify. I will then transfer this to the SR104. Since it has changed its value so much, I am hesitant to use it and want to veryfy it has settled. How do I know it does not change again when shipped (I may actually hand cary it there) to the cal lab (always assuming of course that the homebrew low drift rate is confirmed).
 

Offline acbern

  • Frequent Contributor
  • **
  • Posts: 316
  • Country: de
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #26 on: January 16, 2016, 04:38:42 am »

+3...4 ppm is typical for an SR104 of that era.

Let the unit cool down to +15...18 C and after that warm back to the room temparature while measuring the main resistor and the thermistor. The tempco should be very close match to the original chart. And there shouln't be any hystereresis left after reaching the original temperature.


So you say since its build date it should have drifted 3-4ppm until today, right? The 7 ppm drift I meaured was within a year or so.
From its build date until today (always provided my current measurement is right and my homebrew reference is not bad) it would have drifted from 10.000,005kOhm to 10.000,032 kOhm, or 2,7ppm, which looks ok.
I will do the measurement you propose soon (currently my thermal chamber is occupied with some other stuff). I am not expecting any deviation though, the current value measured looks nice, and I hope it will be confirmed when I have my standard re-cal'ed. The main problem is, whay was my measurement so much off when done in 2014.  Your explanation re. humidity sounds valid, but how do I proof (if possible at all) and how did it happen?
 

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 715
  • Country: us
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #27 on: January 16, 2016, 05:15:04 am »
@acbern

Since your SR-104 was made before 1990, have you taken into account the +1.69ppm correction in 1990? Not that it is relevant to the apparent drift, but my SR-104 was made in 1969 and I plan on sending it in for its first cal (for me). I was wondering what to expect on its calculated drift since its original calibration value 46 years ago.
 

Offline ltz2000

  • Regular Contributor
  • *
  • Posts: 103
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #28 on: January 16, 2016, 05:39:51 am »
So you say since its build date it should have drifted 3-4ppm until today, right?

10 kohm (1990) + 3...4 ppm nowadays.

When calculating the total drift you should include the drift of the US legal ohm from the date of manufacture until 1990 plus the 1990 change of definition already mentioned.
 

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 715
  • Country: us
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #29 on: January 16, 2016, 05:47:32 am »
Thanks.

It looks like PTB only adjusted by +0.56 ppm. I was referencing the NIST correction.
 

Offline lars

  • Regular Contributor
  • *
  • Posts: 132
  • Country: se
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #30 on: January 16, 2016, 07:01:18 pm »
About the SR1010: I have an SR1010-10LTC and an SR1010-1k from the sixties. I think they are excellent ratio transfer standards between 1:10 and 1:100 but absolute value suffers from seasonal variations. Both my SR1010´s have about 8-10ppm seasonal variations for about 30-40%RH variations. A regression on the 1k series (10k) for the last 5 years give +0.3ppm/%RH +2ppm/C and -0.5ppm/year against a GR1440-10k. The individual resistors in the SR1010-1k is between -1ppm/C and +6ppm/C.

Lars
 

Offline ltz2000

  • Regular Contributor
  • *
  • Posts: 103
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #31 on: January 16, 2016, 10:03:42 pm »
I have an SR1010-10LTC and an SR1010-1k from the sixties. I think they are excellent ratio transfer standards between 1:10 and 1:100 but absolute value suffers from seasonal variations. Both my SR1010´s have about 8-10ppm seasonal variations for about 30-40%RH variations. A regression on the 1k series (10k) for the last 5 years give +0.3ppm/%RH +2ppm/C and -0.5ppm/year against a GR1440-10k. The individual resistors in the SR1010-1k is between -1ppm/C and +6ppm/C.

That is normal. The SR1010 is a transfer standard.

ESI made also an oil filled sister version, which was much better from that point of view.

A proven trick with the dry version is to seal the enclosure and put silica gel bags under a perforated fake bottom.
 

Offline zlymex

  • Frequent Contributor
  • **
  • Posts: 556
  • Country: cn
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #32 on: February 23, 2016, 04:12:05 pm »
That chart of "Drift of 5 SR104 standard resistors at 23°C" was plotted by me more than 8 years ago when I only had two SR104s.
I've been trying to keep track of every SR104 I can find and now I updated the chart to ten SR104s

Edit: update chart to 12 SR104s.
« Last Edit: June 12, 2016, 11:41:10 am by zlymex »
 

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 715
  • Country: us
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #33 on: February 23, 2016, 04:24:55 pm »
I hope to get my 1969 era SR-104 sent in for its first cal (by me). I will gladly pass that on to you.

On a side note, do you know if any of the owners of these resistors ever had the alpha and beta constants recalculated? I have read conflicting information that they change as they get older. The SR-104 would be a good candidate to confirm or debunk this theory.
 

Offline zlymex

  • Frequent Contributor
  • **
  • Posts: 556
  • Country: cn
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #34 on: February 23, 2016, 06:07:40 pm »
I hope to get my 1969 era SR-104 sent in for its first cal (by me). I will gladly pass that on to you.

On a side note, do you know if any of the owners of these resistors ever had the alpha and beta constants recalculated? I have read conflicting information that they change as they get older. The SR-104 would be a good candidate to confirm or debunk this theory.
One of my SR104 tested by NBS in Dec 1982, alpha=-0.10ppm/C, beta=-0.028ppm/C2, as shown in green line.
I tested it in Dec 2009, alpha=-0.09ppm/C, beta=-0.033ppm/C2, as shown in red line.
ref: http://bbs.38hot.net/thread-1352-1-1.html

« Last Edit: February 23, 2016, 06:10:32 pm by zlymex »
 

Offline ManateeMafia

  • Frequent Contributor
  • **
  • Posts: 715
  • Country: us
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #35 on: February 24, 2016, 01:18:56 am »
Thanks for the information.
 

Offline jfphp

  • Regular Contributor
  • *
  • Posts: 121
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #36 on: June 12, 2016, 02:41:51 am »
I have a 40 years old SR104 and I am very interested by the comparison chart of 10 SR104 but I have 2 questions :
1. Why some SR104 show no drift at all ? Mesurement errors ?
2. Most important, how do you insert in the graph the 1990 change in the ohm definition ?
"The U.S. representation of the ohm based on the quantum Hall effect to be maintained at NIST starting January 1, 1990, using the internationally agreed-upon or conventional value of the von Klitzing constant RK-90 = 25 512.807 ? exactly. From quantized Hall resistance measurements it is calculated that ?(NIST-90) will exceed ?(NBS-48)01/01/90 by 1.69 ppm"
There must be a jump in the various drift lines or did you recalculate the drift without the jump???
 

Offline zlymex

  • Frequent Contributor
  • **
  • Posts: 556
  • Country: cn
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #37 on: June 12, 2016, 11:33:39 am »
I have a 40 years old SR104 and I am very interested by the comparison chart of 10 SR104 but I have 2 questions :
1. Why some SR104 show no drift at all ? Mesurement errors ?
2. Most important, how do you insert in the graph the 1990 change in the ohm definition ?
"The U.S. representation of the ohm based on the quantum Hall effect to be maintained at NIST starting January 1, 1990, using the internationally agreed-upon or conventional value of the von Klitzing constant RK-90 = 25 512.807 ? exactly. From quantized Hall resistance measurements it is calculated that ?(NIST-90) will exceed ?(NBS-48)01/01/90 by 1.69 ppm"
There must be a jump in the various drift lines or did you recalculate the drift without the jump???

1. I could be the combination of low drift(or right drift) of the resistors with the change in the definition. It is for certain that most good wire-wound resistors are drifting upwards(value increase) if properly made and not constrained.

2. If a very stable resistor is constantly compared to the NBS/NIST standard, there should be a jump in 1990 when graphed.
NBS Ohm was based on the mean value of five Thomas resistors before 1990, and this definition is not constant by nature.
According to an IEEE paper in 1989(http://www.nist.gov/calibrations/upload/eee-38-2.pdf), the drift rate of NBS Ohm was -0.053ppm per year, which means that the five resistors were drifting at +0.053ppm per year before 1990. That change of 1.69ppm in 1990 is not random value in my opinion, rather, it is the correction of NBS Ohm back to the right track.

If a stable resistor was placed into dormancy during 1980 to 1995 for 15 years, it will not aware of all these changes. That is exactly the case for most of the SR104s in my chart. I plotted the chart with dots of data and connect them in straight lines.

Someone has asked me before how I obtain those data on the chart, it is simple, I gather data from the web. Sometime it is already charted like that Japanese SR104(http://www20.tok2.com/home/daisuken/esisr10401.html) but most other times I extract data from calibration stickers(http://lionelectroniclabs.com/esi-sr-104-10k-resistance-standard/) or descriptions(http://www.amplifier.cd/Test_Equipment/other/Widerstands_Normal.htm).
 

Offline ap

  • Regular Contributor
  • *
  • Posts: 242
  • Country: de
    • ab-precision
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #38 on: June 12, 2016, 07:26:50 pm »
@zlymex
Nice work on this overview. What you could do is to take the 1.69 ppm into account in a second version of your chart on the pre-1990 data, so that the real drift would be shown.
Metrology and test gear and other stuff: www.ab-precision.com
 

Offline zlymex

  • Frequent Contributor
  • **
  • Posts: 556
  • Country: cn
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #39 on: June 12, 2016, 10:57:26 pm »
@zlymex
Nice work on this overview. What you could do is to take the 1.69 ppm into account in a second version of your chart on the pre-1990 data, so that the real drift would be shown.

Of course, I should have done that long time ago ^-^
Here it is.
The formula is:    =IF(S1>=32874,U1,U1+1.69-(32874-S1)/365.24*0.053)
Where 32874 is the date code of 1-Jan-1990 when the -1.69ppm, U1 is the value calibrated on S1
 

Offline ap

  • Regular Contributor
  • *
  • Posts: 242
  • Country: de
    • ab-precision
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #40 on: June 12, 2016, 11:38:21 pm »
That was fast! Thx!
Metrology and test gear and other stuff: www.ab-precision.com
 

Online doktor pyta

  • Frequent Contributor
  • **
  • Posts: 262
  • Country: pl
    • RF Scientific
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #41 on: August 27, 2018, 06:28:15 pm »
Sad things also need to be posted. Below pictures of my DIY shorting bars after over 2 years. Bars were used maybe 10 times and were always touched by the edges.
I think the gold and silver layers were too thin.
Also there is interesting .pdf in the attachment. Guildline 'borrowed' a picture from my website and put it into their brochure.
That's not elegant for such company. Haven't they resources to take their own photo ?

Offline MK

  • Regular Contributor
  • *
  • Posts: 185
  • Country: gb
Re: ESI Resistance Standard SR1010 and Standard Resistor SR104
« Reply #42 on: November 05, 2018, 04:33:21 am »
...
Also there is interesting .pdf in the attachment. Guildline 'borrowed' a picture from my website and put it into their brochure.
That's not elegant for such company. Haven't they resources to take their own photo ?
Why not send them the fee for using your photo?
 