I have a 40 years old SR104 and I am very interested by the comparison chart of 10 SR104 but I have 2 questions :
1. Why some SR104 show no drift at all ? Mesurement errors ?
2. Most important, how do you insert in the graph the 1990 change in the ohm definition ?
"The U.S. representation of the ohm based on the quantum Hall effect to be maintained at NIST starting January 1, 1990, using the internationally agreed-upon or conventional value of the von Klitzing constant RK-90 = 25 512.807 ? exactly. From quantized Hall resistance measurements it is calculated that ?(NIST-90) will exceed ?(NBS-48)01/01/90 by 1.69 ppm"
There must be a jump in the various drift lines or did you recalculate the drift without the jump???
1. I could be the combination of low drift(or right drift) of the resistors with the change in the definition. It is for certain that most good wire-wound resistors are drifting upwards(value increase) if properly made and not constrained.
2. If a very stable resistor is constantly compared to the NBS/NIST standard, there should be a jump in 1990 when graphed.
NBS Ohm was based on the mean value of five Thomas resistors before 1990, and this definition is not constant by nature.
According to an IEEE paper in 1989(
http://www.nist.gov/calibrations/upload/eee-38-2.pdf), the drift rate of NBS Ohm was -0.053ppm per year, which means that the five resistors were drifting at +0.053ppm per year before 1990. That change of 1.69ppm in 1990 is not random value in my opinion, rather, it is the correction of NBS Ohm back to the right track.
If a stable resistor was placed into dormancy during 1980 to 1995 for 15 years, it will not aware of all these changes. That is exactly the case for most of the SR104s in my chart. I plotted the chart with dots of data and connect them in straight lines.
Someone has asked me before how I obtain those data on the chart, it is simple, I gather data from the web. Sometime it is already charted like that Japanese SR104(
http://www20.tok2.com/home/daisuken/esisr10401.html) but most other times I extract data from calibration stickers(
http://lionelectroniclabs.com/esi-sr-104-10k-resistance-standard/) or descriptions(
http://www.amplifier.cd/Test_Equipment/other/Widerstands_Normal.htm).