Dr. Frank, I'm glad you stopped by to comment also. I was having a host of problems with my 5440B.
https://www.eevblog.com/forum/repair/fluke-5440b-psu-fault-check-gaurded-power/msg1509067/#msg1509067
I have started preparing to get it up and running again, especially since I have something I can
calibrate it with (732A). If you go and have a look at the link above, it will show some of the
problems I came up against. Any feedback would help since the 5442 is similiar to the 5440.
Hi,
apart from some additional functionality, the 5442A is virtually identical in HW & SW.
The stability in reality seems to be the same as the better specified 5440, especially after years of use.
I don't know of any hardware related difference to achieve the better stability, maybe apart from a pre-selection of the reference.
I had a look into your error description, but there is too few details given, sorry. I assume you have one or several digital problems.
Have you checked systematically all the power supplies? Please provide a full list of all the test points, outguard and inguard which you might also find in the service manual.
Your error messages, do they arise from the internal analog and digital self test procedures, or are they appearing directly, when you power your unit on?
Does your unit output anything in OPERATE mode?
When I in school, 45 years ago, I have designed and built several measurement instruments, especially a 7digit precision universal counter, based on a grave of TTL ICs. Unfortunately, I have later torn it apart again for part salvaging, as it devlopped more and more loose connections.
Today, I really regret this deeply.
Therefore, I would never let go such a device like your 5440B. You can always fix it on a later point in time, when you have gathered more electronics experience and systematics in error analysis.
Also, does anyone know how long it takes the 732A to stabilize? I have heard months and I have
heard weeks. This unit is show -15uV on the 10V output over 93 hours of logging. If I am
understanding correctly, this -1.5ppm. Is that correct?
I second bdunham7's statements:
-15µV / 10V = -1.5ppm, that's right.
Against which other reference did you check this, and how was your setup?
How can you tell, that you are not only seeing the drift of your other reference?
(Or were these simply your 34401A or 3456A?)The 732A is based on the SZA263 as well, therefore, I expect that your instrument would recover to its default output and its normal drift rate (~3ppm/Year) within a few days at most.
Whenever I power up my 5442A, it needs about 4h only to reach exactly its former 10V output, within +/- 0.2ppm, compared to my whole, very well characterized reference group.
So I can tell, that my 5442A drifted less than 1ppm in 5 years, and that the SZA263 does not show any noteworthy on/off hysteresis or lengthy recovery time.
Have you made sure, that the 732As oven is running properly? Then, any room temperature changes during those 93h can only have an influence on that other reference you're using. The T.C. of the 732A is specified < 0.05ppm/°C.
If I use my 3458A for such a 24h comparison, I have to make sure that R.T. stays constant within +/- 0.2°C, which is the characteristics of my basement lab. A 3°C change of R.T. would already give 1.5ppm deviation for my 3458A, as it's characterized total T.C. is about +0.45ppm/K w/o ACAL. Your DMMs are on the same T.C. order of magnitude.
Frank