I have always used a Fluke 5500 , 5520, or 5522A.  Do any of you have any alternatives with similar functionality with good quality?  Interested in hearing opinions.
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: Fluke 5522A Alternatives
« Reply #1 on: June 07, 2018, 10:10:39 pm »
Datron/Wavetek 9000 and 9100 are very similar in functionality and performance.  Their interface isn't very intuitive and they are generally designed for DUTs with 2 terminals, like handheld meters.

I much prefer the 5520A over either.   We have 3 here at ENI that I support, they are real workhorses!
All your volts are belong to me
 

Offline dl1640

  • Regular Contributor
  • *
  • Posts: 222
  • Country: cn
Re: Fluke 5522A Alternatives
« Reply #2 on: June 08, 2018, 10:40:03 am »
May be the Transmille 3000 or 4000 series
I used to use 9100, now use 5502A w/ Scope option
 
The following users thanked this post: itsbiodiversity

Offline itsbiodiversityTopic starter

  • Regular Contributor
  • *
  • Posts: 99
  • Country: us
Re: Fluke 5522A Alternatives
« Reply #3 on: June 14, 2018, 12:54:58 pm »
I had this model saved in my bookmarks.  They seem "affordable" compared to their counterpart in Fluke.  I have used the 5522a for years and I know they are great - but as a new laboratory they are also as expensive as another employee. 

Does anyone have direct experience with the Transmille calibrators? 
 

Offline dl1640

  • Regular Contributor
  • *
  • Posts: 222
  • Country: cn
Re: Fluke 5522A Alternatives
« Reply #4 on: June 15, 2018, 10:25:52 am »
I was told by sales person that 9100 design team made Transmille, do not know further.


通过我的 PRA-AL00 上的 Tapatalk发言

 
The following users thanked this post: itsbiodiversity

Offline seebeck

  • Contributor
  • Posts: 32
  • Country: cn
Re: Fluke 5522A Alternatives
« Reply #5 on: October 30, 2018, 10:13:58 am »
Clarke-Hess, the factory manufactures the best phase meter such as the 5600 precision phase standard (phase accuracy : 5m Degrees) all over the world. C-H also produced their multifunction calibrator. For example, the 8080 MFC. The specifications as below:
https://www.clarke-hess.com/multifunction-calibrator-8080/
BTW, I remember Chroma (an instrument manufactory from Taiwan, focus on Hi-pot tester、PSU、etc...), they has manufactured a MFC called 16801  as same as the Clarke-Hess 8080 MFC, but it is discontinuous product for Chroma now. 
If DC function is enough, maybe KH523 is a good solution (http://www.krohn-hite.com), another one is 2701C from Valhalla (https://valhallascientific.com) for your refer.
Update the Chroma 16801 datasheet as attached file
« Last Edit: October 30, 2018, 02:00:42 pm by seebeck »
 
The following users thanked this post: MegaVolt, itsbiodiversity

Offline dacman

  • Frequent Contributor
  • **
  • Posts: 411
  • Country: us
Re: Fluke 5522A Alternatives
« Reply #6 on: November 06, 2018, 12:46:42 am »
Not many calibrators have input protection like the 5522A and 5502A have.
 
The following users thanked this post: itsbiodiversity

Offline mirzavanyaa

  • Newbie
  • Posts: 2
  • Country: pk
Re: Fluke 5522A Alternatives
« Reply #7 on: November 07, 2018, 12:29:48 pm »
I'm considering the 3 output DVR1 voltage reference, RS1 resistance standard and CS1 capacitance standards.  So I'd be spending $600-750 depending on which models I got..  Are there comparable units I should look at or teardown photos of these?
ahsan
 

Offline Sitta

  • Newbie
  • Posts: 1
  • Country: cz
Re: Fluke 5522A Alternatives
« Reply #8 on: March 20, 2022, 12:36:38 pm »
 

Offline Ugur

  • Regular Contributor
  • *
  • Posts: 84
  • Country: tr
Re: Fluke 5522A Alternatives
« Reply #9 on: June 08, 2022, 08:32:49 am »
Hello,
There is no real equivalent to the Fluke 55XX series. Yes Transmille is not a bad device but it can't be a Fluke.
It has some pros but also cons.
For example, note that every range may not give the same measurement uncertainty.
I've used both devices, Fluke is more reliable.
A brand new Fluke can be very expensive, but you can opt for a used Fluke.
Remember, the Fluke 5500A,5520A is no longer manufactured and technical support has been terminated.
If you are not going to use scope or PQ options, the E series is cheaper. But it cannot output both volts and current at the same time.
 

Offline Arhigos

  • Regular Contributor
  • *
  • Posts: 88
  • Country: gr
Re: Fluke 5522A Alternatives
« Reply #10 on: February 05, 2023, 02:32:02 pm »
Just in case, fluke released a new 55-series calibrator. 30A current range, 2Ghz oscilloscope option and 4:1 test uncertainty ratio to calibrate 6.5 digit multimeters  8)

https://www.fluke.com/en-us/product/calibration-tools/multifunction-calibrators/5500x

 

Offline MegaVolt

  • Frequent Contributor
  • **
  • Posts: 917
  • Country: by
Re: Fluke 5522A Alternatives
« Reply #11 on: February 06, 2023, 09:06:34 am »
to calibrate 6.5 digit multimeters[/img]

11 ppm per 2 year..... that's really cool...

 

Offline mendip_discovery

  • Frequent Contributor
  • **
  • Posts: 845
  • Country: gb
Re: Fluke 5522A Alternatives
« Reply #12 on: February 06, 2023, 03:28:29 pm »
The 95% 1yr spec is rather nice. It is a shame my boss won't let me spend the money buying stuff like this for the lab. But as they say, if I work hard etc. then this time next year my boss will have a new car.
Motorcyclist, Nerd, and I work in a Calibration Lab :-)
--
So everyone is clear, Calibration = Taking Measurement against a known source, Verification = Checking Calibration against Specification, Adjustment = Adjusting the unit to be within specifications.
 