Author Topic: General Radio 1403A 1000pf capacitor adjustment  (Read 3167 times)
Offline charliedeltaTopic starter
« on: May 18, 2021, 10:25:23 am »
I have a General Radio 1403A 1000pf capacitor. Its supposed to be a 1000pf capacitor with 0.1 % tollerance. When i measure this capacitor with  several accurate LCR meters I  am getting values readings of around 750pf.
I cant believe that such a well constructed capacitor can be so badly off. There appears to be no easy way of adjusting this capacitor.
I seem to think I am  misunderstanding how this capacitor was calibrated or measured. Can anyone shed some light onto why this capacitor is so badly off?
There is no signs of physical damage to the plates.
Thanks
Offline antintedo
Re: General Radio 1403A 1000pf capacitor adjustment
« Reply #1 on: May 18, 2021, 11:55:28 am »
What is the DC resistance between terminals? There might be whiskers shorting some of the plates.
Offline Stray Electron
« Reply #2 on: May 18, 2021, 12:51:54 pm »
  I might be wrong, but it looks like there is white corrosion on the plates and around the top where it joins the cylindrical shell. And it looks like there are dents and dings in the plates!!
« Reply #3 on: May 18, 2021, 01:03:14 pm »
  The connectors and the plates are simply stacked on metal rods and are held in place with nuts. Have you checked to see if all of the nuts are still tight?  You probably need to check the torque on them with a torque wrench instead of just tightening them with a pair of pliers.  But I have no idea what the torque requirement is. But you could probably find a recommended torque value for that thread size, pitch and material in Machinery's Handbook or other reference.
  To me, it looks like there are clear signs that the inside of that unit has been exposed to moisture. If so I suspect that there's some dissimilar metal corrosion between the aluminium plates and the stainless steel shafts and nuts and I expect that that's what is altering your readings.
« Reply #4 on: May 18, 2021, 01:09:11 pm »
   Did you already take one of the plate supporting rods out of the capacitor assembly?   it looks like there's one missing.  There is a empty hole shown near the top OF THE PHOTO in the bottom photo.
Offline Conrad Hoffman
    • The Messy Basement
« Reply #5 on: May 18, 2021, 01:38:12 pm »
It's hard to believe the connections would be so bad as to disconnect a plate. It's a solid screw system. Was it ever right? Did somebody repurpose it in the past by removing plates to get a new value? I think that's a trimmer screw on top, but it won't find you that much extra capacitance! If you have some decent vernier calipers you could measure the area of the plates and the spacing and maybe determine if the number of plates is grossly off. Did you get it cheap? That might be a clue! I assume you're measuring from center pin to center pin? I think you can also short one side out and measure across the other connector, but it will raise the value a bit.
« Last Edit: May 18, 2021, 01:40:34 pm by Conrad Hoffman »
Online chuckb
« Reply #6 on: May 18, 2021, 10:50:24 pm »
Like someone else mentioned, the capacitor has been modified. It is missing a large number of plates. See attached Pix.
Offline doktor pyta
« Reply #7 on: May 18, 2021, 10:56:00 pm »
chuckb was one minute faster :)
4 plates are missing so it looks like someone needed a custom value.
You can still make use of it by converting it into 100pF DIY standard.
The plates and standoffs are made most likely of aluminum so don't worry about the whiskers.
They were a problem in tin plated can of GR1404 nitrogen filled standards.
Here is mine:
« Last Edit: May 18, 2021, 11:06:14 pm by doktor pyta »
« Reply #8 on: May 19, 2021, 09:01:28 am »
Thanks all for the replies.
1. DC resistance, its open circuit. I tested it with the Fluke insulation tester.
2. DINGS, dents and corrorosion. There was a anti oxident goop that and it was  corroded in some places. The dings and dents were causes by me trying to open it to break the  badly corroded aluminium.
3.There may well be  corrosion. I will dissassemble it completely after first soaking it in isopropyl alcohol. Maybe thats what will fix it.
4.It was a surplus acquisition that was marked as faulty from the lab. But the fault was that the connector wires were broken from  spinning connectors.  It may well be that someone has removed plates to optimise the value for a specific application.
5. I have not  dissambled or removed any plates, its as is received. The original owner may have modified it.
6. Thanks for the pictures. Mystery solved, the unit has been butchered. Anyway I got a  10pf in the deal for 10 dollars so I have some value left in the deal. Thanks all for the help.
Its really sad when people butcher such excellent engineering. Nothing worst than ignorant butchers or tinkerers who think that they can reinvent the calibration wheel.
« Last Edit: May 19, 2021, 09:05:08 am by charliedelta »
Offline ahbushnell
« Reply #9 on: May 19, 2021, 03:35:02 pm »
Look at it this way.  If it doesn't do what you need but you can make it work for what you do need.  Modify.  I don't think it's a collectors item.
« Reply #10 on: May 19, 2021, 04:51:13 pm »
There was a parts unit on eBay, maybe create a single correct one out of two!
Offline Henrik_V
  • “ground” is a convenient fantasy
« Reply #11 on: May 26, 2021, 10:44:23 am »
C-D : I have a opened and ruined 1404 ... but shurely enough parts (plates, screws, spacer) to bring yours back to value... drop me note , would be shipping costs from Germany:)
Then ask Conrad to hook it up to his 1615 :D
Greetings from Germany
Henrik
The number you have dialed is imaginary, please turn your phone 90° and dial again!
Offline manganin
« Reply #12 on: May 27, 2021, 12:20:45 pm »
Probably requires a long storage in bad conditions, but I have seen that more than once with the 1403 fraction of a picofarad types. In those a small 3-terminal capacitance is created by an aperture in the guard plate between the electrodes.
Aluminium oxide is a good insulator.
« Reply #13 on: May 29, 2021, 09:02:40 am »
Hi Henrik
I would be interested in your parts unit. I have sent you a PM
Thanks
Offline Henrik_V
« Reply #14 on: June 14, 2021, 04:35:31 pm »
CD here is what you hopefully get...
Maybe check the torque while dismounting and use a torque wrench to mount ....
« Last Edit: June 14, 2021, 04:48:28 pm by Henrik_V »
Henrik
Offline Labrat101
  • Regular
  • Renovating Old Test Equipment & Calibration ..
« Reply #15 on: June 14, 2021, 10:23:02 pm »
Hi I would not try opening the nuts they are torque set .
If its that far out I would wash it very care fully in IPA . even better an ultrasonic bath .
If dirt dust has got between the plate it will have become static charged to the surfaces .
 so while cleaning it make sure the out terminals are shorted .
If you have loosened any of the clamping nuts . Find one that you have not opened and with a
stainless steel torque key measure the release pressure and the reset all the lock nuts to the same.
when these are assembled they put shims between the plates so they are all parallel to each.
That Last picture looks like it has been knocked and bent .
I would get hold of some Plastic feeler gauges and measure the good plate distances.
 These unit are also dependent on air . so altitude is also part of the calibration & temp.
« Last Edit: June 14, 2021, 10:30:33 pm by Labrat101 »
"   All Started With A BIG Bang!! .  .   & Magic Smoke  ".
Offline Henrik_V
« Reply #16 on: June 15, 2021, 07:07:02 am »
The one pictured is one of a batch made by Quadtech and cans only, most of them failed due to wisker and in some you hear something is jingling inside while shaking. I harvested them out of the dumpster.
We opened some cans (the 1404 cans are sealed and filled with nitrogene to avoid the influence of humidity) hot air didin't do the job, we used a torch  ;D
(and vented the cans first  >:D)
... the one showed dropped  on the floor :-//  some of the plates are bended. However still enough parts for CDs cap.
Maybe one day we rebuild some without solder plating in the housing and polish the plates (the genuine GR ones definitly look better than the successors)
« Last Edit: June 15, 2021, 07:10:26 am by Henrik_V »
Henrik
« Reply #17 on: June 15, 2021, 01:50:55 pm »
I thought the 1404 plates were invar, but not the 1403. Have to check that.
Offline Labrat101
  • Regular
« Reply #18 on: June 15, 2021, 01:56:33 pm »
I am glade you rescued them from near death . ..
 There are a few rules when working with this type of aluminium . Most people don't realises that
aluminium oxidizes in air .Oxygen  just scratching aluminium under a high power microscope the
surface of the scratch looks like its alive and the oxidisation starts happening in a few seconds.
aluminium oxide is not a true insulator more like a semi conductor and the particles are very small and
are always particle charged . Under clean room involvement they can cause a lot of problems in the hyper
filter as the static charge created by the dust can under certain frequencies produce 10kv plus.
so the the plates should only be handled with cotton gloves as not to add oils or other contaminates to
the anodized surface . aluminium is the metal form of Ruby and we all know when ruby is excited we get
light Lazar .
So polishing the plates will cause more harm . clean them with ultra pure DI water . Or IPA .
 And when reassembling at home I use to do it inside a large plastic bag  which helps to keep dust and
moisture limited . (heavy breathing & coffee )
One of those units looked very nice probably the one that was nitrogen filled The pressure is normally
1 bar (14.7 psi) which will give the calibration @ Sea level and wont be influenced by outside pressure
changes .
I'm sure you will get these back within spec with a Lot of Love & tender Care .
Offline Henrik_V
« Reply #19 on: June 15, 2021, 03:43:28 pm »
Don't know for the 1403, but the 1404 and 1406 are invar.
I don't think there is any aluminium in the 1404 ... haven't found some :)
Our plan is to polish the invar plates and maybe have them (and the bolds,spacers...) gold plated  ;D 8)  , more important is the removal of the tin/solder plating of the brass(?) housing parts to prevent whisker.  Maybe will be tricky for/around the throug-hole plating ...  But a very good workshop and galvanic is inhouse :)   also vacuum/ Nx refill / clean room/ temper  all there... but time and motivation ....since we have nice&stable working 1404 .
I attached a picture of one of those and marked a wisker... probably posted before ..
Henrik
Offline Labrat101
  • Regular
« Reply #20 on: June 15, 2021, 04:22:04 pm »
Invar is magnetic check with a small magnet
« Last Edit: June 15, 2021, 04:25:52 pm by Labrat101 »
Offline Labrat101
  • Regular
« Reply #21 on: June 15, 2021, 04:49:14 pm »
Interesting idea about Gold plating . I would check that this is feesable as in the use of 2 Dissimilar Metals
 I think that the extra cost of gold plating during manufacture would not increase the cost of these units very much
 I guess that this process is not good when it comes to high tolerance caps . I would check on what the overall effects are.
 I have never seen gold used in this way .. except in Aerospace and that's for a different reasons.
Offline Henrik_V
« Reply #22 on: June 16, 2021, 09:24:11 am »
I expect that one or more additional layers of some other elements will be needed, but that up to our galvanic experts :)
Henrik
« Reply #23 on: June 16, 2021, 08:09:08 pm »
I used to work for a place that made optical equipment out of invar and had the parts plated. Only plating shops that really know what they're doing are successful and I still don't trust the stuff. Even with a good flash of copper or whatever else they use, adhesion over time can be tricky. I've seen a lot of lifted and peeling plating. Invar has so much nickel that it should be pretty good if the environment is good. Even lying around in the uncontrolled shop, the stuff would slowly develop a haze, but not really corrode. I suspect a polished surface, something we never did, would be resistant to oxidation.
The following users thanked this post: Labrat101
Offline Labrat101
  • Regular
« Reply #24 on: June 16, 2021, 10:50:37 pm »
You right what you say also Invar & Gold is in the wrong place on the periodic table
which is why  copper or any plating will peal as the dielectrics of Invar has a very good temperature stability and copper or silver are the opposite and will peal off.
separate due to the chrome content .
Invar as you said has a very resilient surface and pure chrome is almost as tough as titanium .
« Last Edit: June 16, 2021, 10:57:31 pm by Labrat101 »
