Author Topic: Group-buy for 10V LTZ1000A reference? [Completed, no more boards].  (Read 12206 times)
Offline TiN
  • xDevs.com/live - 24/7 lab feed
    • xDevs.com
Considering Decided to make 10 completely assembled and tested modules to provide +10.00000 +/-4 ppm buffered output.
One of the challenges is cost and time involved to test and calibrate these things.
Device is based around xDevs.com FX module PCB, which have next design specifications:
Reference source: LTZ1000ACH with datasheet schematics, using LT1013ACN.
LTZ Temperature SP: 13K/1K which provide about +55C oven temp
Resistors: VPG BMF HZ for 120ohm, temperature SP. PTF56 5ppm/K for 70-80K bias resistors
Output 7V->10V stage: Chopper opamp ADA4522-2 in SO8, Hermetic resistor network or BMF resistors. Each unit adjusted by trim resistor for tight output tolerance.
Onboard supply: Linear LT3042 LDO, configured for +11V output to LTZ circuit.
PCB: 4-layer FR4 with ENIG plating
Input power requirements: +12V and -12V.
Power consumption : ~0.6W when operated with +24C ambient
Additional features : isolated section with I2C EEPROM with calibration data/config/history (RPI python app to be provided to maintain data), MAX6610 PCB temperature sensor for tracking
Calibration traceability: <4 ppm to 732B, uncertain with 0.6ppm.
Expected performance: <4 ppm/year long-term stability, <0.1ppm/K tempco over 18-28C ambient temperature.
Assembly type: Hammon metal box with Pomona 3770 low-thermal binding posts.
I'd like to check if there is any interest in expensive hobby-level reference among the members here.
If enough people would be willing to buy one, it will be easier for me to pull the trigger and tie in 4-5K$ into this project.
Estimated delivery date would be March 2018 October 2018, with final good cost $500 USD.
Specs above are on safe side, so the module likely to be 2-5 times more stable, however I wouldn't guarantee it just yet. There is no replacement to actually running reference for 3+ years to determine actual long-term annual stability.
Feel free to comment :)
Member list, who interested in module
montemcguire - 1 unit [CONFIRMED]
dr.diesel - 1 unit [CONFIRMED]
chriswebb - 1 unit [CONFIRMED]
PeLuLe - 2 unit [CONFIRMED]
MM - 1 unit [CONFIRMED]
bck - 1 unit [CONFIRMED]
USA cal club - 1 unit [CONFIRMED]
CM - 1 unit [CONFIRMED]
zucca - 1 unit
[CONFIRMED]
No more units available., unless somebody change their mind.
Prototype board overview, test data samples
Initial output after first assembly.
« Last Edit: August 29, 2018, 04:59:23 am by TiN »
YouTube | Chat room | Live-cam | Have documentation to share? Upload here! No size limit, firmware dumps, photos.
The following users thanked this post: dr.diesel, bck, BU508A
Online Vgkid
Re: Group-buy for 10V LTZ1000A reference?
« Reply #1 on: December 06, 2017, 05:24:58 am »
Sounds interesting, I would do it if I ever get the money. That would be around 6mo + from now.
Offline montemcguire
« Reply #2 on: December 06, 2017, 06:33:16 am »
I'd buy one - count me in!
Online lukier
    • Homepage
« Reply #3 on: December 06, 2017, 11:40:23 am »
Does it really need -12V? It would be much simpler to have just one battery (12V lead-acid or 4x18650 LiPo).
Online dr.diesel
« Reply #4 on: December 06, 2017, 12:29:10 pm »
Count me in, prepay is no problem to help offset costs.
Online maxwell3e10
« Reply #5 on: December 06, 2017, 01:33:45 pm »
Sounds interesting. What would be the specification for 0.1 to 10Hz rms noise? That is primarily what I am interested in.
Online mimmus78
« Reply #6 on: December 06, 2017, 01:45:32 pm »
I'm interested too but as I'm building my own version I'm not interested in buying one.
I have a DHL contact with good fee for international shipping ... I can offer few trips for free is some of this board wants to became a traveling standard among EU guys or for just a stop and go.
« Last Edit: December 06, 2017, 01:50:43 pm by mimmus78 »
Offline nidlaX
« Reply #7 on: December 06, 2017, 01:54:15 pm »
I'm interested, but it's only a run of 10 units?
Offline chriswebb
« Reply #8 on: December 06, 2017, 04:17:57 pm »
I'd also be interested in picking up a unit if available.
Always learning. The greatest part of life is that there will always be more to learn.
Offline Kleinstein
« Reply #9 on: December 06, 2017, 05:35:10 pm »
The circuit should run with less than -12 V. It should be possible to make it run on just a single supply, maybe 14 V. For battery operation something like 12 V min would be nice, but this would not leave much room for driving both the positive and negative side. So getting it down to 4 Li Cells could be a little tricky.
Offline TiN
    • xDevs.com
« Reply #10 on: December 06, 2017, 05:47:53 pm »
montemcguire, dr.diesel, chriswebb - OK, I'll keep you guys in mind.
-3V can be fine too, I checked today, output change was less than 0.4ppm.
But I will be validating only +12/-12 +/-5%. Negative voltage is used on the output stage, not the LTZ-ref itself.
I expect ~3 uV perhaps. Will know better once I test first 3 prototypes I have now.
Even to make 10 boards, lot of time and money will be tied in, so I'd rather spend some resource for other projects, using this reference, instead of making just the reference. I originally did not plan to make any extra's at all, and still need to run the numbers to pull the trigger.
Also each output resistor network must match the used LTZ chip value, so it's NOT simple assemble/hook to fancy meter/write value/sell process.
« Last Edit: December 08, 2017, 11:42:15 am by TiN »
The following users thanked this post: dr.diesel, chriswebb
Offline pelule
  • What is business? It’s other people’s money
« Reply #11 on: December 06, 2017, 06:31:52 pm »
Count me for 2 (if such many left available), min. 1.
Prepayment possible.
BR
/PeLuLe
You will learn something new every single day
Online wn1fju
« Reply #12 on: December 06, 2017, 07:55:29 pm »
Could I buy one volt for $50USD ???
The following users thanked this post: bck
Offline VintageNut
« Reply #13 on: December 07, 2017, 12:36:41 am »
Hi TiN
I am curious about the 0.1 ppm/deg C temp dependency. If the temp changes +/- 3 deg C, then the output will change +/- 3uV?
The temp in my home/office can change +/- 5 deg, which is 10 deg p-p. That would translate to 10uV p-p or 1 ppm p-p.
My 731B which cost $250 is pretty much that good for temp dependency.
I think that if your LTC ref were in an oven like the KE DMM7510, THAT would be very attractive. \
One reason that I like my DMM7510 so much is that the ovenized reference makes the DMM rock solid stable in my home/office where the temperature is not stable.
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
Offline TiN
    • xDevs.com
« Reply #14 on: December 07, 2017, 01:00:57 am »
VintageNut
Will have to see if can do better. Current data looking like this:
Ignore the steep jump on the end, that shows reference ran out of oven thermal margin, with resulting +150ppm raise of zener voltage and output voltage.
« Last Edit: December 07, 2017, 10:22:28 am by TiN »
Offline texaspyro
« Reply #15 on: December 07, 2017, 02:00:24 am »
So 17 uV over 10 degrees C...
My ancient scrap heap salvaged EDC MV100 seems to drift around 20 uV over 3 degree C... not too shabby for an un-modified, non-ovenized 1N829 zener diode referenced clunker with a 6 decade adjustable divider... I've been looking at seeing what some LTZ1000 / LM199 / better chopper love could do for it.
Offline bson
« Reply #16 on: December 07, 2017, 03:44:21 am »
I'd be interested in a few (say 3) known to be genuine LTZ1000 references, just the parts without a board - if you can get a few extra!
Offline TiN
    • xDevs.com
« Reply #17 on: December 07, 2017, 04:27:08 am »
Actually more like 22 uV over 22.6C, here's live data.
So that's about 1uV/C, or -0.1ppm/K.
LTZ reference itself was tested last week to ~-0.03ppm/K over 20-50C span. TC is not exactly linear due to compensation.
Sweet spot with nearly zero TC is to run with ambient temp 30 to 40C.
I'd suggest to move discussion about the design to thread.
There will be no parts/kits, only fully assembled and tested module is in question.
« Last Edit: December 07, 2017, 04:32:04 am by TiN »
Offline texaspyro
« Reply #18 on: December 07, 2017, 04:36:13 am »
I already have a couple of 732A's and a 732B,  but if you need a another taker to meet your minimum build I'm game.  Might even be interested, otherwise.
The main difficulty I see with the unit is the +/- 12V power.   I'd like to do a small unit with LiPo (18650) backup batteries.  There are some nice, isolated, low noise DC-DC's out there that should be able to take care of that.. hopefully without slathering the system with switching noise.
Online Echo88
« Reply #19 on: December 07, 2017, 09:49:33 am »
Maybe a capacitor charge pump suffices for the negative voltage while ensuring low noise? Or a big LED + solarcell.  ;D
Offline zucca
  • EE meid in Itali
« Reply #20 on: December 08, 2017, 10:54:16 am »
TiN let's the magic happen, it's Xmas time!
I am in. Yes Yes Yes!
PS: anyone with connections for custom resistor arrays/dividers (Fluke 2464707)
https://www.eevblog.com/forum/metrology/tin-livestreaming-layout-of-new-ltz1000a-10v-reference-module/msg1345273/#msg1345273
Offline TiN
    • xDevs.com
Re: Group-buy for 10V LTZ1000A reference? [Completed, no more boards].
« Reply #21 on: December 08, 2017, 11:26:58 am »
zucca
OK. What did you do with previous ref, btw? :)
ALL
No more boards, so far I got all allocated, unless somebody from the list wanna change the mind.
Ordering parts (LTZs from LT direct, resistors from VPG direct, rest are from Digikey) next week.
The following users thanked this post: dr.diesel, zucca, montemcguire
Offline zucca
  • EE meid in Itali
« Reply #22 on: December 08, 2017, 12:02:54 pm »
zucca
I needed first a HP 3245A, now I am going to upgrade it...
Following your path.. very slowly my friend.
Offline TiN
    • xDevs.com
« Reply #23 on: December 11, 2017, 04:49:16 am »
LTZs, opamps and LDO's ordered from LTC  ::).
The following users thanked this post: dr.diesel, zucca, montemcguire, chriswebb
Offline TiN
    • xDevs.com
« Reply #24 on: January 07, 2018, 03:05:14 pm »
Linear parts and Fluke network resistors arrived.
The following users thanked this post: dr.diesel, zucca, chriswebb
Offline TiN
    • xDevs.com
« Reply #25 on: January 23, 2018, 04:18:09 am »
Got quote from VPG resistors (13/1K network hermetic, 120R hermetic, 70K epoxy).
However lead-time is looking around 11-16 weeks, which makes it around May'18.
If anyone want to drop off waiting that long, let me know.
Offline Pipelie
  • Country: cn
« Reply #26 on: January 24, 2018, 01:44:19 am »
the lead-time of VPG resistors easily goes around 16-18 weeks :(, we ordered several times from Vishay in the past 3 years.
Offline ramon
  • Country: tw
« Reply #27 on: January 24, 2018, 02:43:27 pm »
Hi Tin! What is the price for the 13/1K hermetic resistor network? I am thinking about building two LTZ1000 boards .... someday. So I can wait.
Offline TiN
    • xDevs.com
« Reply #28 on: January 24, 2018, 04:32:26 pm »
It's around $70ish so far.
Offline Kleinstein
« Reply #29 on: January 24, 2018, 04:44:16 pm »
For a 10 V reference the 13K/1 K divider at the LTZ1000 is not that important. The divider for the 7 V to 10 V step is way more important (e.g. about by a factor of 50).  It would only be for an possible 7 V output that the stability of the temperature set point can matter. There is a kind of natural limit on how much to spend on the resistors - at some point it would be more effective to have a second ref. chip or a way to measure the voltage at the 13K/1 K divider - possibly to make it an extra 0.4xx V reference output.
Offline cellularmitosis
« Reply #30 on: January 24, 2018, 06:51:39 pm »
The VHD200 12.5/1K divider I ordered from Texas Components was $46.
Edit: lead time is 3 weeks.  should be here by 1/29.
« Last Edit: January 24, 2018, 06:53:14 pm by cellularmitosis »
LTZs: KX FX MX CX PX Frank A9 QX
Offline chriswebb
« Reply #31 on: January 24, 2018, 08:16:03 pm »
Is there anything you need from me to get my order confirmed? Payment sent or something? I think I missed when you were getting the orders confirmed.
Offline TiN
    • xDevs.com
« Reply #32 on: January 25, 2018, 01:08:22 am »
Naw, you all good. Just sanity check for myself.
Offline 1anX
« Reply #33 on: January 25, 2018, 01:32:52 am »
A little off topic but how well does a AD584KH compare used in temp controlled (aircoditioned) lab?
LiPo battery powered versions are readily available https://www.ebay.com/itm/High-Precision-4-Channel-2-5-7-5-5-10V-Voltage-Reference-Module-AD584KH-NEW-coi/202035834876?hash=item2f0a462ffc:g:4D0AAOSw0ThZoOBg
Would this be a waste of money to use to check calibration of lab equipment?
Update: I did a forum search and found my answer! http://www.markhennessy.co.uk/ad584_references/
« Last Edit: January 25, 2018, 02:10:23 am by 1anX »
Offline Jxw6637
  • Country: cn
« Reply #34 on: January 28, 2018, 10:32:07 pm »
I am interested in getting one, is there an extra?
Offline TiN
    • xDevs.com
« Reply #35 on: January 30, 2018, 04:09:01 am »
Sorry, no more PCBs on this project.
Offline TiN
    • xDevs.com
« Reply #36 on: June 09, 2018, 01:58:25 pm »
Yay, finally something arrived.
70K resistors will be used in above-mentioned FX 10pcs run. 1K/13K VHP203T's are just for temporary testing, final boards will  have 1/13K VHD divider instead.
Rest of resistors (120R, divider) expected to arrive next month.
Resistors were ordered February 7, 2018, so this gives us ~17 week delivery time for these and ~20weeks or so for remaining.
Started tempco test already.
Offline ManateeMafia
« Reply #37 on: June 09, 2018, 02:04:53 pm »
you need more 3458A   >:D
Offline johartley2001
« Reply #38 on: June 19, 2018, 12:05:10 pm »
are these still available to buy ?
Offline TiN
    • xDevs.com
« Reply #39 on: June 19, 2018, 12:41:07 pm »
All FX units are allocated for members, and I'm still waiting for final part delivery - resistors. It is typical to have 17-24 weeks lead time order for precision resistors.
Perhaps somebody on EEVBlog have another LTZ reference to sell.
Offline eurofox
    • Music
« Reply #40 on: June 19, 2018, 01:04:43 pm »
Build one yourself, there is enough information in this thread to do it.
Mine in aging right now and start to be very stable, it include a 10V buffer, it is mounted in a teco box and I will add a oven control on it to bring the temperature to 35°C +/- 0;1° C.
In fact I build different versions most without buffer to 10V.
Only problem is final calibration  :horse:
eurofox
Offline SKPang
    • SK Pang Electronics Ltd
« Reply #41 on: June 28, 2018, 09:02:42 am »
Hi TiN,
Are you going to do another batch ? If so put me down for one.
Offline TiN
    • xDevs.com
« Reply #42 on: June 28, 2018, 12:23:13 pm »
Nope, due to high cost only 14pcs of these modules will exist in the world, sorry.
KX one is available on OSH Park for PCBs : https://oshpark.com/shared_projects/HfKcqjV3
(LTZ chips you can buy from ADI and resistors from Edwin Pettis here or VPG).
Offline TiN
    • xDevs.com
« Reply #43 on: August 08, 2018, 09:33:02 pm »
After 181 days of waiting, complete VPG order is finally here.
Now I can go ahead and build stuff!
The following users thanked this post: dr.diesel, bck, zucca, montemcguire, chriswebb
Online Vgkid
« Reply #44 on: August 09, 2018, 03:30:49 am »
Damn , all of those resistors cost more than everything I have bought in my lab... :-[   :scared:
Offline TiN
    • xDevs.com
« Reply #45 on: August 13, 2018, 12:29:53 pm »
Okay, Digikey box from USA arrived nice and safe.
Wait is over, we are finally ready to start assembly. Enjoy, that's 6K+ in parts...  :scared:
The following users thanked this post: Echo88, dr.diesel, bck, zucca
Offline zucca
  • EE meid in Itali
« Reply #46 on: August 13, 2018, 12:45:47 pm »
The following users thanked this post: ManateeMafia, dr.diesel
Offline TiN
    • xDevs.com
« Reply #47 on: August 14, 2018, 04:10:48 pm »
Got some parts for low-noise regulators assembled.
Close-up.
The following users thanked this post: ManateeMafia, dr.diesel, montemcguire
Offline TiN
    • xDevs.com
« Reply #48 on: August 16, 2018, 06:37:42 pm »
Had little chance to work on assembly tonight.
LDO regulators on 5 boards are good and quiet, now putting in LTZ section parts (except BMF resistors and LTZ itself, obviously) and output stage parts.
BMF resistors are still going thru sorting/binning so it will take some time till i get overall result table to pre-sort those before installing into boards permanently.
Offline branadic
« Reply #49 on: August 16, 2018, 06:48:31 pm »
The ceramic caps are quite a bit big for the footprint, aren't they?
-branadic-
Fluke 8050A | Prema 5000 | Prema 5017 SC | Advantest R6581D | GenRad 1434-G | Datron 4000A | Tek 2465A | VNWA2.x with TCXO upgrade and access to: Keysight 3458A, Keithley 2002, Prema 5017 SC, 34401A, 34410A, Keithley 2182A, HDO6054, Keysight 53230A and other goodies at work
The following users thanked this post: TiN, bck
Online Andreas
« Reply #50 on: August 16, 2018, 06:54:49 pm »
perhaps they are only soldered on the wrong edge  :-//
with best regards
Andreas
Offline TiN
    • xDevs.com
« Reply #51 on: August 16, 2018, 07:01:11 pm »
I got that one messed up in order, Caps are 1210 on 1206 pads.  ???
Offline TiN
    • xDevs.com
« Reply #52 on: August 26, 2018, 03:23:06 pm »
Almost done on matching resistors.  :)
The following users thanked this post: Echo88, dr.diesel, bck, zucca, montemcguire
Offline TiN
    • xDevs.com
« Reply #53 on: September 08, 2018, 01:50:38 pm »
Sets ready for tempco trim of 7V stage.
Ref 3 is are my "old" prototypes, unlabelled one is USA CalClub version.
Offline Dr. Frank
« Reply #54 on: September 08, 2018, 06:33:50 pm »
And what about this 7001 reference?
Is it defect, or could you determined its  oven setpoint?
Frank
Offline Philipp23
« Reply #55 on: February 23, 2019, 07:27:58 am »
Hello
I hope it is OK to write down this here. Is any one of the lucky owners of the 10V LTZ1000A reference willing to trade it against other precision metrology equipment? For example a National Instruments PXI-4071 7½-Digit FlexDMM or a very nice digital Multimeter that was way ahead of its time the Prema 8017 Multifunction Meter?
Kind Regards
Philipp
Online Echo88
« Reply #56 on: February 28, 2019, 03:07:33 pm »
@Kleinstein: Could you elaborate on the possibility to omit the Negative Voltage Supply for the bipolar output buffer stage which is used in the FX-reference?
Im at the moment testing this output-design and thinking about doing the same, but cant think of a solution which wouldnt add drift and noise when shifting the ground.
Im using an LT1054-based charge-pump for now, to generate the needed negative voltage, like its done in the 732B.
Offline Kleinstein
« Reply #57 on: February 28, 2019, 05:26:28 pm »
The circuit to provide separate negative side drive and sene pins only needs a relatively low voltage. Depending on the OPs use some 0.5 or 1 V could be already enough, if one limits is to compensating normal small lead resistance.  One can create such a small negative supply with just something like 2 forward diodes to get a small voltage drop. So the external supply negative side would be the negative supply, the reference circuit ground would be behind some 2 diodes, some 1.2 V higher.  This would need the external supply to be higher by the drop of the diodes. So 12 V external supply could be too little or at least quite close to the edge. A 14 V supply could be sufficient.
Here it helps that modern AZ OPs often are Rail to Rail. So just -1.2 V or even -0.6 V could be enough to compensate a few 10 mV of drop on the wires.
Online Echo88
« Reply #58 on: March 01, 2019, 02:12:45 pm »
Indeed it works, thanks.
Im wondering about the noise contribution of the diodes and would have expected that the output-noise increases, but it stays the same no matter if i use a dedicated -6V Supply or the diodes (Noise/Noise2.asc).
Simulations are attached. Maybe i made an error? Didnt simulate noise in a long time.
Offline Kleinstein
« Reply #59 on: March 01, 2019, 02:41:23 pm »
Creating the negative supply from the string of diodes makes not big difference over a real regulated -6 V. The OPs PSRR  should be quite good and the voltage over the diodes is usually quite low noise (most regulators are higher noise), just a little temperature dependent.
The noise simulation is not really correct, as Vout is relative to the ground symbol. So the negative side sense part is not correctly included. To get it better, move the GND point to the negative output.
The OPs seem to not include noise (at least not in the models used on my PC).  The noise model of the reference may also be not that accurate.
Online Echo88
« Reply #60 on: March 01, 2019, 03:35:44 pm »
Weird, if i connect the gnd-symbol to the negative supply (Q3 Collector) as you suggested, then ill actually get less measured noise than when ground-symbol is connected to the lower side of R3 across Vout.  :o This happens in both posted versions, with diodes and Voltage Source as negative supply.
Did you simulate your ground-symbol-suggestion? I use the normal Solver-Method and cant recall that i changed any other tolerance/etc. that might influence the simulation.
I should use other OP-Amp-Models if it doesnt include the OP-Amp-Noise and i was suspicious wether LTSpice would simulate diode-noise correctly.
I think the diode-solution might produce less noise than the cap-switcher which produces EMI, but ill have to read more about noise-simulation to get a better fitting simulation of the diode-based-solution.
Offline Kleinstein
« Reply #61 on: March 01, 2019, 04:00:38 pm »
I have not done the simulation with the moved GND.  Getting a lower noise with the correct GND point is not such a surprise, as the OPs are modeled without noise.  Though the difference should be rather small as it's mainly the diodes and a few resistors.
One can kind of add some of the OP voltage noise to the simulation with a high resistor in series to one input. Some, likely more modern models do include noise for the OPs too.
AFIAK the noise of forward biased diodes is modeled reasonably good. So the diode part is definitely low noise, just a more drift for the negative supply voltage.
The charge pump chips tend to produce quite some ripple / spikes, depending on the layout - not really something to see in a simulation. Chances are a switcher with inductor is lower EMI.
Offline TiN
    • xDevs.com
« Reply #62 on: March 19, 2019, 11:13:31 am »
Guess it's a big (sorta) day today.
First preliminary FX was shipped few hours ago.
Not too impressed with Pomona 3770 posts. While job getting done, they are still kinda flimsy to use. Certainly do not ooze quality as "proper" 25$ 5-way posts.  >:D
I'm still undecided if should use LEMO multi-way connector to provide input +/-12V power for module, isolated power for EEPROM/PCB temperature sensor and I2C data, or just stick with simpler dual BNC connectors for +/-12V only. LEMO option is nice but it's expensive and over the planned budget already. If somebody have better idea, feel free to tell.
The following users thanked this post: dr.diesel, zucca, Grandchuck
Offline zucca
  • EE meid in Itali
« Reply #63 on: March 19, 2019, 06:46:52 pm »
Thanks for the update, personally I would love the LEMO. I wanted a Ferrari, you are getting pretty close to perfection.
"Abbiamo fatto trenta, facciamo trentuno!"
It is a phrase that is historically attributed to Pope Leo X who in 1517 had to name some new cardinals. With great work, he came to draw up the final list, reaching as many as 30 names, already a substantial number. In the end, however, he realized that he had not added a prelate of great merit and fame, not being able to exclude any cardinal, he added the name to the list, pronouncing the famous words: "We did 30, we do 31!"
Online Andreas
« Reply #64 on: March 19, 2019, 07:58:12 pm »
For power supplies I usually use DIN (Diode) connectors.
e.g.
https://de.rs-online.com/web/p/din-steckverbinder/1564224/
https://de.rs-online.com/web/p/din-steckverbinder/0776088/
https://de.rs-online.com/web/p/din-steckverbinder/0776167/
But you could also use D-Sub connectors.
with best regards
Andreas
Offline guenthert
« Reply #65 on: March 20, 2019, 02:23:02 am »
Congrats!
Pity that they're not for sale, but I hope that, once I'm reunited with my gear, I can get on the US CalClub list and meet one of those units then ...
Offline TiN
    • xDevs.com
« Reply #66 on: June 20, 2019, 04:31:02 am »
Playing with matplotlib and python. Here I can present tempco plot, captured from eight xDevs.com FX LTZ1000A references (before trim, as is):
Purpose of the test is to establish baseline for TC before readjustment, to make sure we have no problems with temperature stability. My final goal is to have each ref 10V output deliver better than 0.03 ppm/K.
In test - each reference scanned twice, one value for raw LTZ1000A output (7.x-ish volt) and second value is 10V output (before trim adjustment). Except one output, channel 12 is connected to my xDevs/Fluke 792X FX ref +9.9999779(22) VDC, outside of chamber to verify stability of scanner setup. Worst ref = channel 14, 10V output -0.059 ppm/K.  :scared:
Scanner = Two Keithley 7168 nanovolt JFET cards. Detector : Two Keysight 3458.
Temperature swept by Keithley 2510 in DIY airchamber box, with 40W watercooled TEC and Xeon copper fansink inside as heat radiator.
Temperature in the box measured by Fluke 1529 thermometer and Honeywell RTD PT100 probe.
Used RAW data file, DSV format.
Python 2.7 app to plot graph. Depends on scipy, numpy and matplotlib.
« Last Edit: June 20, 2019, 04:35:38 am by TiN »
Offline TiN
    • xDevs.com
« Reply #67 on: October 05, 2019, 01:48:55 am »
Alright, after 95 weeks, time for long awaited update has come.
Few units are completed and delivered while ago as test drive, and feel happy so far:
MM - 1 unit [DONE]
USA cal club - 1 unit [DONE]
Now I need these people to send me their shipping information in PM, so I can proceed.
montemcguire - 1 unit [PENDING]
dr.diesel - 1 unit [TBD]
chriswebb - 1 unit [TBD]
PeLuLe - 2 unit [PENDING]
bck - 1 unit [TBD]
CM - 1 unit [TBD]
zucca - 1 unit [PENDING]
I understand it was a loooong wait, so If somebody changed mind and don't want a ref anymore, that is OK, just let me know so I update the list accordingly.
« Last Edit: October 05, 2019, 12:25:54 pm by TiN »
The following users thanked this post: bck, niner_007
Offline niner_007
« Reply #68 on: October 05, 2019, 10:01:17 pm »
If anyone changed their mind or you got extras, I’m interested, please let me know
Offline syau
  • Country: hk
« Reply #69 on: October 06, 2019, 07:01:44 am »
If anyone changed their mind or you got extras, I’m also interested, please let me know
Offline wolfy007
« Reply #70 on: October 06, 2019, 11:42:58 am »
Same here, if theres an extra, Ill take it.
Tek: 222A, 212, TAS465, TDS460A, 2465, 2465B, 2467, TM504, TM506, SG504, SG503, PG506, TG501, SC504, PS503a, FG501a, DM504a, PG501, 7603+7L12
BWD: 835, 539d, 521, 504, 141, 160a  Fluke: 8060A, 8024B, 8025A, 27/FM, 87, 87III   HP: 5315A, 3478a, 3468A, 3457a  Keithley: 196, 228a  Solartron: 7081 7151
Offline zucca
  • EE meid in Itali
« Reply #71 on: October 07, 2019, 10:04:09 am »
This is like good italian wine, the more you wait the better it gets.
PM sent.
The following users thanked this post: alex-sh