Does anyone have a picture or schematic of the internal circuit of these testadapter for the HM8118 LCR-bridge?
Are there any additional circuitry inside to make valid transformer measurements?

Many thanks in advance.

 

Offline F64098

  • Contributor
  • Posts: 38
  • Country: de
Re: Hameg HZ186 4 Terminal Transformator Test Cable - Internal Circuit
« Reply #1 on: October 16, 2018, 08:36:23 pm »
Hello,

when i read all accesible manuals, there is real nothing in this adaptor.
The main thing of this adaptor is the description of the connectors, so that you connect the DUT in the right way.

Best regards

Frank
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 1485
  • Country: de
Re: Hameg HZ186 4 Terminal Transformator Test Cable - Internal Circuit
« Reply #2 on: October 16, 2018, 09:12:34 pm »
There is nothing inside, it is just a 4W connection to both terminals. I once disassambled one of them, as I had to develope an application specific test adapter.

-branadic-