Author Topic: Hioki DM7276-01 Precision DC Voltmeter (9 PPM)  (Read 46215 times)
Offline maxwell3e10
Re: Hioki DM7276-01 Precision DC Voltmeter (9 PPM)
« Reply #25 on: June 03, 2023, 05:05:49 pm »
It is an impressive stability, only 1-2 uV. But the statistics is still wrong, maybe the smoothing filter messes it up. It also would be interesting to download the data and plot, see how many digits are used internally.
« Last Edit: June 03, 2023, 05:08:28 pm by maxwell3e10 »
Online iMo
  • Country: sm
« Reply #26 on: June 03, 2023, 05:21:52 pm »
It does smoothing with "moving average" - how many samples do you average?
PS: What is the architecture of the ADC in that nice meter? We can see the AD7175 ADC there, as well as parts used in the Mutlislope (even though no fpga or mcu in the inguard?)..
« Last Edit: June 03, 2023, 05:38:43 pm by iMo »
Offline Kleinstein
« Reply #27 on: June 03, 2023, 06:18:00 pm »
They say that they use a sigma delta ADC and chances are this would be the AD7175 and not an extra seprate build one. The AD7175 is a slightly off choice, more made for high speed, but still OK. The relatively fast ADC may be used to get still relatively fast trigger / start of a measurement. A slower SD ADC may have too much delay at the start.
Noise wise the AD7175 should be good enough, even with a background check of the ADC and ref. divder gain.
Chances are there is a µC for the ingurad part, as there is a 2nd crystal or clock generator besides the one at the ADC. Chances are the 16 pin chip next to it and close to the signal isolators would be something like a µC.
The following users thanked this post: iMo
Online iMo
  • Country: sm
« Reply #28 on: June 03, 2023, 07:17:42 pm »
Unbelievably complex inguard board, however.. My first thought had been it is a kind of multislope with 7175 as the rundown something..  :D
Offline Kleinstein
« Reply #29 on: June 03, 2023, 07:33:16 pm »
The circuit indeed looks quite complicated for just a simple single channel DC voltmeter.
As an extra function there is the contact checking mode via capacitance, but I would not expect that much circuit for this either.
There some relatively large ceramic capacitors connected to the four ADG1211 switches. This looks a lot like switching capacitors for voltage division.
So this gives quite a few chips to replace some recision resistors.
It gets a little more complicated from having some kind of ACAL - though still no reason for that much circuit (e.g. 1 x ADG1208, 1-2 relays and some resistors should be enough).
Offline HighVoltageTopic starter
« Reply #30 on: June 04, 2023, 10:11:40 am »
Here the measurements have been running over night with the same settings.
The smoothing filter is set to Average over 10 measurements
Online iMo
  • Country: sm
« Reply #31 on: June 04, 2023, 12:10:24 pm »
FYI noise with RAND() and 1uV resolution.
« Last Edit: June 04, 2023, 12:15:47 pm by iMo »
Offline DavidKo
  • Country: cz
« Reply #32 on: June 05, 2023, 06:00:56 am »
While Keysight 34470A and keithley DMM7510 had a dramatic price increased in the last few years, this Hioki Voltmeter has been stable.
Since the warranty has expired on my DM7276-01, and if there is any interest, I could open it up and take some pictures from the inside.
I like Hioki products, but for home user their products are not too much interesting. When the products are out of the support, they do not calibrate the device nor send the calibration/service manual (same experience with Japan and Czech support). Compared to mentioned Keysight or Keithley, where the calibration adjustment procedure is available.
« Reply #33 on: June 05, 2023, 07:07:11 am »
That is very true.
Japanese companies in general do not like to give out service manuals or calibration procedures or schematics.
However their products are usually very good.
Keithley and Keysight are going more and more in the same direction and I would not be surprised, if they also would not supply calibration procedures in the future.
Offline ap
    • ab-precision
« Reply #34 on: June 05, 2023, 01:23:02 pm »
Unless its broken, and given the fact that this Hioki is DCV only, one can get away with verification. This way the drift history is not covered either by re-adjustment.
Metrology and test gear and other stuff: www.ab-precision.com
Offline aronake
« Reply #35 on: January 21, 2024, 05:16:55 pm »
The voltage reference board read BT4560 in the corner, indicating that it is the same voltage reference board as in BATTERY IMPEDANCE METER BT4560:
https://www.hioki.com/in-en/products/resistance-meters/battery/id_5897
There are sellers on Taobao who claim to have ADR1000 for sale.
https://item.taobao.com/item.htm?abbucket=5&id=761537983785&ns=1&spm=a21n57.1.0.0.5256523cWu7jic
They are on very similar voltage reference boards with an ADR1000 but read DM7275. So it looks like ADR1000 may be used in newer models, and as has been indicated that Linear is now selling it for companies using it in products.
The following users thanked this post: HighVoltage
Offline gamalot
  • Correct my English
    • Youtube
« Reply #36 on: January 21, 2024, 07:20:08 pm »
They chose OPA2140 on their LTZ1000 board.
I'm a poet, I didn't even know it. |  https://youtube.com/@gamalot | https://github.com/gamalot
Offline gamalot
    • Youtube
« Reply #37 on: January 22, 2024, 04:05:04 am »
Sorry to go off topic again, are these resistor values 1.58K? I have never seen such top marking before.  :-//
Offline Lajt
  • Country: hr
    • Lajtronix
« Reply #38 on: February 21, 2024, 10:56:37 pm »
If anyone is interested, I've made a DIY temperature probe for Hioki DM7275, DM7276 & RM3544.
The build is documented on https://lajtronix.eu/2024/02/21/diy-hioki-dm7275-76-temp-probe-z2001/
It has been tested and is working great.
Br, Lajt
The following users thanked this post: HighVoltage, ch_scr
Offline laichh
  • Country: my
« Reply #39 on: February 22, 2024, 03:14:28 am »
1.58 kohm is correct, these resistors are likely made by KOA.
Offline gamalot
    • Youtube
« Reply #40 on: February 22, 2024, 11:40:24 am »
I have a lot of small light blue KOA Speer chip resistors and most of them don't have top markings, even the 0603 size ones.
Offline gamalot
    • Youtube
« Reply #41 on: February 22, 2024, 11:42:11 am »
Br, Lajt
I happen to have a Z2001 sensor that came with my RM3545, maybe I can do some simple measurements on it.
Offline Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #42 on: April 19, 2024, 11:49:46 pm »
Well, I've purchased the DM7275-03 for my home lab. So far I am quite happy with what this small unit does. I like that it is completely quiet, seems to be very stable and as arrived did differ about -3ppm from my usual standards (which I count as about 10ppm accuracy in any case). The input current measures about -20pA and seems to be stable and not noisy. The temperature measurement feature is very handy and useful. I am already getting some good data on the low frequency noise of my HP3245A and other sources, with the passive filter the baseline noise is about 600nV p-p in 0.1-10Hz range at 1NPLC.
I have a question though - does anybody knows a calibration company in the UK which would calibrate this unit with traceability, data and uncertainties (to my surprise, the unit came without any calibration data from the manufacturer).
My next home project now could be to build a universal shunt box for the Hioki and the HP3456A, to measure DC currents from about 100uA FS to 1A FS. I have good 10K and 1R0 references but it would be much more convenient to have a box with proper range switching and fuses.
Cheers
Alex
« Last Edit: April 19, 2024, 11:52:03 pm by Alex Nikitin »
Offline alm
« Reply #43 on: April 21, 2024, 10:35:26 am »
I have no suggestion for a lab, but you'll need a lab with sufficiently low uncertainties (check their scope of calibration) that doesn't mind doing something they may not have a procedure for. The calibration procedure in the manual (see attachment) should be straight-forward for any cal lab with the right equipment. So I don't see why it would be difficult for them to produce a traceable certificate with data and uncertainties. Adjustment may be more tricky, but do you really need adjustment? Can you not live with just knowing the offsets?
I'm a bit surprised there is no linearity check like a negative voltage or say 50% of range in the calibration procedure. I didn't study the design, but is the linearity guaranteed by design?
    • A.N.T. Audio
« Reply #44 on: April 21, 2024, 11:35:57 am »
I did some checks on zero volt, range and polarity change, and in general errors are not more than 1LSB (which for the range change could be 1ppm of value, measuring +/-1V at 10V range). The zero drift on 100mV range with a shorted input is below 50nV for 10min, the offset noise on Slow speed is ~350nV p-p (and that is without offset correction). Which means this small thing might be as good a Null-Meter as the HP3458A (and might be even better as it stays cool unlike most accurate bench meters).
On calibration I am not worried too much especially as the difference between the meter at 10V and my own references is slowly getting smaller in time and now below 2ppm. However it might be nice to have a way of adjusting the unit in the future, and that should increase the value, as HIOKI policy on second hand units is simple - they would deal only with the original owner, as far as I understand.
Cheers
Alex
« Last Edit: April 21, 2024, 11:37:36 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #45 on: April 21, 2024, 03:40:31 pm »
I am already thinking about building an "extender" for the Hioki, a box with relays and resistors controlled from the I/O connector of Hioki and providing DC current and resistance measurement options. It is not that Hioki can not measure DC current on its own (see the photo below  ;) )
Cheers
Alex
Offline Kleinstein
« Reply #46 on: April 21, 2024, 04:04:00 pm »
It is indeed odd that the calibration points are only 0 and full scale. Checking the negative side would be good and not that difficult. It's not a strickt INL test, more a crude check the amplifier / supply is not broken.
The ADC is an AD7175 high end SD ADC chip. So the linearity can be pretty good and normally no separate adjustment needed for the positive and negative side. If the design is OK it is unlikely the the meter would develope high INL over time.
At the expected linearity level it is anyway a bit tricky to check, as the specs for the usual calibrators are limited (the actual performance may be better).
    • A.N.T. Audio
« Reply #47 on: April 21, 2024, 10:34:01 pm »
DM7275/6 has an interesting feature, called scaling, where it is possible to use a linear scaling factor and offset, plus rename the units if required. The scaling coefficients A and B (for A*x +B expression) can be entered from the front panel or remotely with 7 digits precision and later the combination can be saved as a setup and recalled when needed. In essence it makes the unit very convenient to use with current shunts, as long as the shunt is stable enough, the tolerance doesn't matter much as it can be corrected out. The unit also offers a (linear) parametric correction from temperature sensor readings, up to +/-1000 ppm/C. I did make a nice 1 Ohm reference recently using a precision hermetic high-current JRL 1 Ohm resistor, the same resistor as evaluated on xDevs. The resistor I've used measured about + 560 ppm and I just found almost exact correction resistor value. However with the HIOKI one can use one of these JRL resistors as is and just calibrate the error out by scaling when use it as a shunt to measure current.
Actually the scaling can be used as a user calibration option to correct for the reference drift if needed one day ;) .
Cheers
Alex
    • A.N.T. Audio
« Reply #48 on: April 23, 2024, 10:02:10 pm »
I did some measurements on the capacitive coupling of my bench meters to the mains ground (i.e. to the meter case in all three units) for another thread on the forum here. The newcomer came on top by not a small margin, to my surprise:
Hioki DM7275-03
Positive input to mains ground - 10V range and below <30pF, 100V and 1000V <10pF
Negative input to mains ground  <30pF
Between input terminals  ~1500pF for 10V range and below*, <10pF for 100V and 1000V ranges.
Here I can add that I also measured the DC leakage to ground from measuring terminals with my Keithley 617 electrometer: for +100V above the ground the leakage is 2.5pA, for -100V it is -6pA.
Cheers
Alex
* - possibly something to do with the contact checking feature, even if it is OFF. The cure is simple (if the input C is a problem) - to put a 10K resistor in series with the positive lead and use Hi-Z, on 10V and below the meter has ~20pA input current or ~0.2uV added error with the resistor.
The following users thanked this post: HighVoltage, doktor pyta, ch_scr
    • A.N.T. Audio
« Reply #49 on: April 24, 2024, 09:52:04 pm »
Six days continuously powered on, the HIOKI is settlings down nicely, the difference with my reference meters and Fluke 731B is now essentially gone, the initial difference of about -3ppm has disappeared in about 3 days time and the meter appears to be stable - however I will continue to monitor its stability. Here just as an illustration the HIOKI measuring the Fluke 731B for 60 min when the temperature in my home lab was reasonably constant (the TC of 731B is about -2ppm/C, so the 0.2ppm rise over 60 min means the temperature has dropped by about 0.1C ). The red line is 30s running average, vertical scale is 1uV = 0.1ppm /div. It is good to see two low noise references (LTZ1000A and RefAmp) working well together.
Cheers
Alex
P.S. - I've corrected the graph's title, the measurements were made at NPLC10, 1s intervals, not NPLC1 .
« Last Edit: April 25, 2024, 10:34:27 am by Alex Nikitin »