Author Topic: Hioki DM7276-01 Precision DC Voltmeter (9 PPM)  (Read 28312 times)
Offline HighVoltageTopic starter
« on: June 03, 2021, 12:09:46 pm »
This HIOKI DM7276-01 has been added to my lab.
It is advertised as a 9ppm/year 7 1/2 digit DC Voltmeter with touch screen a very small footprint and lightweight.
Some specs:
- 100 mV ( ±120.000 00 mV) to 1000 V ( ±1000.000 0 V ), 5 ranges
- High-accuracy model with 1-year 9ppm Accuracy
- Capacitance contact check (using built-in C-monitor)
- Supports global production with built-in variable power supply
- Built-in EXT I/O, LAN, and USB
- Outstanding Long-term stability and Temperature characteristics
- Excellent noise performance approaching an 8-½ digit DMM
- Specifications ideal for everything from R&D to production lines at ¼ the conventional costs
I am just starting using this instrument and the first thing that is noticeable is its very fast warmup time from around 3 min, maybe less.
Stay tuned, first measurements should be interesting.
Anyone here having experience with this instrument?
The following users thanked this post: Andreas, Vgkid, doktor pyta, msliva, gamalot, Micke, bsw_m, ch_scr, MegaVolt, eplpwr
Offline MegaVolt
  • Country: by
Re: Hioki DM7276-01 Precision DC Voltmeter (9 PPM)
« Reply #1 on: June 04, 2021, 07:49:57 am »
A very interesting find !!
How much did it cost?
Can you share the documentation that is only available upon registration? "Instruction Manual", "Communication Command Instruction Manual".
I am already myself :))
« Last Edit: June 04, 2021, 08:28:47 am by MegaVolt »
Offline MegaVolt
  • Country: by
« Reply #2 on: June 04, 2021, 08:17:31 am »
Teardown? :)))))))))))))))
« Reply #4 on: June 04, 2021, 08:52:39 am »
Here I tested the warmup time of the DM7276.
My LTZ1000 reference was warmed up well, so no influence from that.
This little peak shortly before 2 min seems normal for this instrument.
It seems within 2 min this DM7276 is warmed up to get reliable values.
And this is even faster than the Keithley DMM7510.
The following users thanked this post: MegaVolt
« Reply #5 on: June 04, 2021, 08:58:53 am »
In this picture the Hioki DM7276 was hooked up to my LTZ1000A reference over night
Settings:
- 100 NPLC
- Smoothing filter ON with 5 average readings
The nominal value of my LTZ1000 reference is 7.065,90xV
For comparison I had the Keithley DMM7510 running in parallel to the Hioki.
The following users thanked this post: Mickle T., MegaVolt, Trader
Offline MegaVolt
  • Country: by
« Reply #6 on: June 04, 2021, 09:05:20 am »
It is twice as noisy as the 7510...
Offline Kleinstein
« Reply #7 on: June 04, 2021, 03:44:55 pm »
The factor is visible from the example data, but AFAIK the DMM7510 is very mixed perfomance when it comes to noise. For the data I have see, it is very good in the 100 mV range and also good with short integration like 1 PLC. However with 100 PLC the performance is not that great and a bit disappointing - at least with a lower applied voltage. With the test at 10 V the references could be a main factor.
Both meters are not really made for ultimate precision at 100 PLC. The main use is more like relatively fast production tests at 10 PLC or 1 PLC, often with a scanner.
The following users thanked this post: EE54
Offline Kleinstein
« Reply #8 on: June 04, 2021, 06:31:42 pm »
The Hoki meter is a voltmeter only - so no shunt and not current sources for resistance. This allows for a lower price.
Chances are the accurcy specs for the 10 V range are to a large part limited by reference drift. The external temperature regulation for the LTFLU in the DMM7510 has some disadvantage in this respect. The specs are also only estimates about future performance - the actual performance is usually better, but unknown upfront. Some estimates are more conservative than others.
The following users thanked this post: MegaVolt, E-Design
Offline guenthert
« Reply #9 on: June 06, 2021, 05:41:10 pm »
     It's even a DC-only voltmeter, which further simplifies the input stage and eases timing requirements (while the 3458A is often praised here for its precision and linearity, it was really its speed which puts it ahead, more so in '89).
     I was even wondering why they bother with a multi-range DC-only voltmeter.  Why not make it 10V only, simplifying the input stage even further (making it cheaper or more precise) and selling it as digitizer?  Seems they are eyeing the currently hot battery market.   :-//
« Last Edit: June 06, 2021, 05:43:38 pm by guenthert »
Offline aronake
« Reply #10 on: June 01, 2023, 07:13:09 am »
Do anyone have more experiences to share on this multimeter?
Maybe some teardown pictures?
Anyone know what voltage reference it uses?
Linearity?
With the collapse of the Japanese yen is is getting relatively inexpensive from Japan.
https://store.shopping.yahoo.co.jp/firstfactory/soku-hioki-dm7276-02.html
373K JPY is around 2.7K USD. Some freight etc added of course.
Offline aronake
« Reply #11 on: June 01, 2023, 07:15:07 am »
Similar price in Europe apparently.
https://www.calplus.de/hioki-dm7276-01.html
« Reply #12 on: June 01, 2023, 11:24:50 am »
While Keysight 34470A and keithley DMM7510 had a dramatic price increased in the last few years, this Hioki Voltmeter has been stable.
Since the warranty has expired on my DM7276-01, and if there is any interest, I could open it up and take some pictures from the inside.
The following users thanked this post: aronake, Nanitamuscen
Offline MegaVolt
  • Country: by
« Reply #13 on: June 01, 2023, 12:07:30 pm »
It is very interesting :)
« Reply #14 on: June 01, 2023, 03:57:44 pm »
It has an LTZ1000ACH on a separate small board with all SMD components.
First set of pictures
« Reply #15 on: June 01, 2023, 03:58:36 pm »
Second set of pictures
« Reply #16 on: June 01, 2023, 03:59:53 pm »
Last set of pictures
The following users thanked this post: Echo88, Mickle T., doktor pyta, coromonadalix, ace1903, aronake, ch_scr, MegaVolt, iMo, KT88, eleguy
« Reply #17 on: June 01, 2023, 04:23:38 pm »
Back together and still working.
The following users thanked this post: Mickle T., Andreas, ace1903, aronake, ch_scr, MegaVolt, iMo, Svgeesus, KT88, Detlev
Offline maxwell3e10
« Reply #18 on: June 01, 2023, 10:23:41 pm »
Settings:
- 100 NPLC
Thanks for the tests and teardown. Something seems wrong with Hioki displayed statistics. P-P cannot be less than twice S_n. P_P would be equal to twice  S_n if the distribution is purely bimodal, otherwise P_P should be more than twice S_n.
Offline laichh
  • Country: my
« Reply #19 on: June 02, 2023, 07:55:33 am »
Here are the inside of a Hioki DM7275, I don't see any different compared against a DM7276, seems identical to me.
« Last Edit: June 02, 2023, 07:58:15 am by laichh »
The following users thanked this post: Mickle T., HighVoltage, ace1903, ch_scr, MegaVolt, iMo, chekhov, Smms73
Offline laichh
  • Country: my
« Reply #20 on: June 02, 2023, 07:57:46 am »
more Hioki DM7275 pictures.
The following users thanked this post: Mickle T., HighVoltage, ace1903, ch_scr, MegaVolt, iMo, Smms73
« Reply #21 on: June 02, 2023, 09:15:01 am »
This was a quick and dirty setup, definitely not reliable.
Let me do a proper hookup and testing.
Offline Kleinstein
« Reply #22 on: June 02, 2023, 11:48:34 am »
The inconsistancy between the p-p and standard deviation values has nothing to do with the input signal it is a pure math thing.  For a given P-P value one gets the largest RMS value / standard deviation with a kind of square wave and this gets 50% of the p-p value. It is only the SD estimator for a small sample size that adds the extra factor of square root of (N/(N-1)) that can make it slightly larger.
The test description indicates more like a longer test.
« Reply #23 on: June 03, 2023, 03:31:29 pm »
Here are the results of 100 measurements with 100 PLC and smoothing filter ON
The source is a stable LTZ1000A
« Reply #24 on: June 03, 2023, 04:57:58 pm »
1.5 hours later and 1394 measurements later
Same setting
« Last Edit: June 03, 2023, 04:59:37 pm by HighVoltage »
The following users thanked this post: coromonadalix, Smms73
Offline maxwell3e10
« Reply #25 on: June 03, 2023, 05:05:49 pm »
It is an impressive stability, only 1-2 uV. But the statistics is still wrong, maybe the smoothing filter messes it up. It also would be interesting to download the data and plot, see how many digits are used internally.
« Last Edit: June 03, 2023, 05:08:28 pm by maxwell3e10 »
Offline iMo
  • Country: li
« Reply #26 on: June 03, 2023, 05:21:52 pm »
It does smoothing with "moving average" - how many samples do you average?
PS: What is the architecture of the ADC in that nice meter? We can see the AD7175 ADC there, as well as parts used in the Mutlislope (even though no fpga or mcu in the inguard?)..
« Last Edit: June 03, 2023, 05:38:43 pm by iMo »
Offline Kleinstein
« Reply #27 on: June 03, 2023, 06:18:00 pm »
They say that they use a sigma delta ADC and chances are this would be the AD7175 and not an extra seprate build one. The AD7175 is a slightly off choice, more made for high speed, but still OK. The relatively fast ADC may be used to get still relatively fast trigger / start of a measurement. A slower SD ADC may have too much delay at the start.
Noise wise the AD7175 should be good enough, even with a background check of the ADC and ref. divder gain.
Chances are there is a µC for the ingurad part, as there is a 2nd crystal or clock generator besides the one at the ADC. Chances are the 16 pin chip next to it and close to the signal isolators would be something like a µC.
The following users thanked this post: iMo
Offline iMo
  • Country: li
« Reply #28 on: June 03, 2023, 07:17:42 pm »
Unbelievably complex inguard board, however.. My first thought had been it is a kind of multislope with 7175 as the rundown something..  :D
Offline Kleinstein
« Reply #29 on: June 03, 2023, 07:33:16 pm »
The circuit indeed looks quite complicated for just a simple single channel DC voltmeter.
As an extra function there is the contact checking mode via capacitance, but I would not expect that much circuit for this either.
There some relatively large ceramic capacitors connected to the four ADG1211 switches. This looks a lot like switching capacitors for voltage division.
So this gives quite a few chips to replace some recision resistors.
It gets a little more complicated from having some kind of ACAL - though still no reason for that much circuit (e.g. 1 x ADG1208, 1-2 relays and some resistors should be enough).
« Reply #30 on: June 04, 2023, 10:11:40 am »
Here the measurements have been running over night with the same settings.
The smoothing filter is set to Average over 10 measurements
Offline iMo
  • Country: li
« Reply #31 on: June 04, 2023, 12:10:24 pm »
FYI noise with RAND() and 1uV resolution.
« Last Edit: June 04, 2023, 12:15:47 pm by iMo »
Offline DavidKo
  • Country: cz
« Reply #32 on: June 05, 2023, 06:00:56 am »
I like Hioki products, but for home user their products are not too much interesting. When the products are out of the support, they do not calibrate the device nor send the calibration/service manual (same experience with Japan and Czech support). Compared to mentioned Keysight or Keithley, where the calibration adjustment procedure is available.
« Reply #33 on: June 05, 2023, 07:07:11 am »
That is very true.
Japanese companies in general do not like to give out service manuals or calibration procedures or schematics.
However their products are usually very good.
Keithley and Keysight are going more and more in the same direction and I would not be surprised, if they also would not supply calibration procedures in the future.
Offline ap
    • ab-precision
« Reply #34 on: June 05, 2023, 01:23:02 pm »
Unless its broken, and given the fact that this Hioki is DCV only, one can get away with verification. This way the drift history is not covered either by re-adjustment.
Metrology and test gear and other stuff: www.ab-precision.com
Offline aronake
« Reply #35 on: January 21, 2024, 05:16:55 pm »
The voltage reference board read BT4560 in the corner, indicating that it is the same voltage reference board as in BATTERY IMPEDANCE METER BT4560:
https://www.hioki.com/in-en/products/resistance-meters/battery/id_5897
There are sellers on Taobao who claim to have ADR1000 for sale.
https://item.taobao.com/item.htm?abbucket=5&id=761537983785&ns=1&spm=a21n57.1.0.0.5256523cWu7jic
They are on very similar voltage reference boards with an ADR1000 but read DM7275. So it looks like ADR1000 may be used in newer models, and as has been indicated that Linear is now selling it for companies using it in products.
The following users thanked this post: HighVoltage
Online gamalot
  • Correct my English
    • Youtube
« Reply #36 on: January 21, 2024, 07:20:08 pm »
They chose OPA2140 on their LTZ1000 board.
I'm a poet, I didn't even know it. |  https://youtube.com/@gamalot | https://github.com/gamalot
Online gamalot
    • Youtube
« Reply #37 on: January 22, 2024, 04:05:04 am »
Sorry to go off topic again, are these resistor values 1.58K? I have never seen such top marking before.  :-//
Offline Lajt
  • Country: hr
    • Lajtronix
« Reply #38 on: February 21, 2024, 10:56:37 pm »
If anyone is interested, I've made a DIY temperature probe for Hioki DM7275, DM7276 & RM3544.
The build is documented on https://lajtronix.eu/2024/02/21/diy-hioki-dm7275-76-temp-probe-z2001/
It has been tested and is working great.
Br, Lajt
The following users thanked this post: HighVoltage, ch_scr
Offline laichh
  • Country: my
« Reply #39 on: February 22, 2024, 03:14:28 am »
1.58 kohm is correct, these resistors are likely made by KOA.
Online gamalot
    • Youtube
« Reply #40 on: February 22, 2024, 11:40:24 am »
I have a lot of small light blue KOA Speer chip resistors and most of them don't have top markings, even the 0603 size ones.
Online gamalot
    • Youtube
« Reply #41 on: February 22, 2024, 11:42:11 am »
Br, Lajt
I happen to have a Z2001 sensor that came with my RM3545, maybe I can do some simple measurements on it.
Offline Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #42 on: April 19, 2024, 11:49:46 pm »
Well, I've purchased the DM7275-03 for my home lab. So far I am quite happy with what this small unit does. I like that it is completely quiet, seems to be very stable and as arrived did differ about -3ppm from my usual standards (which I count as about 10ppm accuracy in any case). The input current measures about -20pA and seems to be stable and not noisy. The temperature measurement feature is very handy and useful. I am already getting some good data on the low frequency noise of my HP3245A and other sources, with the passive filter the baseline noise is about 600nV p-p in 0.1-10Hz range at 1NPLC.
I have a question though - does anybody knows a calibration company in the UK which would calibrate this unit with traceability, data and uncertainties (to my surprise, the unit came without any calibration data from the manufacturer).
My next home project now could be to build a universal shunt box for the Hioki and the HP3456A, to measure DC currents from about 100uA FS to 1A FS. I have good 10K and 1R0 references but it would be much more convenient to have a box with proper range switching and fuses.
Cheers
Alex
« Last Edit: April 19, 2024, 11:52:03 pm by Alex Nikitin »
Offline alm
« Reply #43 on: April 21, 2024, 10:35:26 am »
I have no suggestion for a lab, but you'll need a lab with sufficiently low uncertainties (check their scope of calibration) that doesn't mind doing something they may not have a procedure for. The calibration procedure in the manual (see attachment) should be straight-forward for any cal lab with the right equipment. So I don't see why it would be difficult for them to produce a traceable certificate with data and uncertainties. Adjustment may be more tricky, but do you really need adjustment? Can you not live with just knowing the offsets?
I'm a bit surprised there is no linearity check like a negative voltage or say 50% of range in the calibration procedure. I didn't study the design, but is the linearity guaranteed by design?
    • A.N.T. Audio
« Reply #44 on: April 21, 2024, 11:35:57 am »
I did some checks on zero volt, range and polarity change, and in general errors are not more than 1LSB (which for the range change could be 1ppm of value, measuring +/-1V at 10V range). The zero drift on 100mV range with a shorted input is below 50nV for 10min, the offset noise on Slow speed is ~350nV p-p (and that is without offset correction). Which means this small thing might be as good a Null-Meter as the HP3458A (and might be even better as it stays cool unlike most accurate bench meters).
On calibration I am not worried too much especially as the difference between the meter at 10V and my own references is slowly getting smaller in time and now below 2ppm. However it might be nice to have a way of adjusting the unit in the future, and that should increase the value, as HIOKI policy on second hand units is simple - they would deal only with the original owner, as far as I understand.
Cheers
Alex
« Last Edit: April 21, 2024, 11:37:36 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #45 on: April 21, 2024, 03:40:31 pm »
I am already thinking about building an "extender" for the Hioki, a box with relays and resistors controlled from the I/O connector of Hioki and providing DC current and resistance measurement options. It is not that Hioki can not measure DC current on its own (see the photo below  ;) )
Cheers
Alex
Offline Kleinstein
« Reply #46 on: April 21, 2024, 04:04:00 pm »
It is indeed odd that the calibration points are only 0 and full scale. Checking the negative side would be good and not that difficult. It's not a strickt INL test, more a crude check the amplifier / supply is not broken.
The ADC is an AD7175 high end SD ADC chip. So the linearity can be pretty good and normally no separate adjustment needed for the positive and negative side. If the design is OK it is unlikely the the meter would develope high INL over time.
At the expected linearity level it is anyway a bit tricky to check, as the specs for the usual calibrators are limited (the actual performance may be better).
    • A.N.T. Audio
« Reply #47 on: April 21, 2024, 10:34:01 pm »
DM7275/6 has an interesting feature, called scaling, where it is possible to use a linear scaling factor and offset, plus rename the units if required. The scaling coefficients A and B (for A*x +B expression) can be entered from the front panel or remotely with 7 digits precision and later the combination can be saved as a setup and recalled when needed. In essence it makes the unit very convenient to use with current shunts, as long as the shunt is stable enough, the tolerance doesn't matter much as it can be corrected out. The unit also offers a (linear) parametric correction from temperature sensor readings, up to +/-1000 ppm/C. I did make a nice 1 Ohm reference recently using a precision hermetic high-current JRL 1 Ohm resistor, the same resistor as evaluated on xDevs. The resistor I've used measured about + 560 ppm and I just found almost exact correction resistor value. However with the HIOKI one can use one of these JRL resistors as is and just calibrate the error out by scaling when use it as a shunt to measure current.
Actually the scaling can be used as a user calibration option to correct for the reference drift if needed one day ;) .
Cheers
Alex
    • A.N.T. Audio
« Reply #48 on: April 23, 2024, 10:02:10 pm »
I did some measurements on the capacitive coupling of my bench meters to the mains ground (i.e. to the meter case in all three units) for another thread on the forum here. The newcomer came on top by not a small margin, to my surprise:
Hioki DM7275-03
Positive input to mains ground - 10V range and below <30pF, 100V and 1000V <10pF
Negative input to mains ground  <30pF
Between input terminals  ~1500pF for 10V range and below*, <10pF for 100V and 1000V ranges.
Here I can add that I also measured the DC leakage to ground from measuring terminals with my Keithley 617 electrometer: for +100V above the ground the leakage is 2.5pA, for -100V it is -6pA.
Cheers
Alex
* - possibly something to do with the contact checking feature, even if it is OFF. The cure is simple (if the input C is a problem) - to put a 10K resistor in series with the positive lead and use Hi-Z, on 10V and below the meter has ~20pA input current or ~0.2uV added error with the resistor.
The following users thanked this post: HighVoltage, doktor pyta, ch_scr
    • A.N.T. Audio
« Reply #49 on: April 24, 2024, 09:52:04 pm »
Six days continuously powered on, the HIOKI is settlings down nicely, the difference with my reference meters and Fluke 731B is now essentially gone, the initial difference of about -3ppm has disappeared in about 3 days time and the meter appears to be stable - however I will continue to monitor its stability. Here just as an illustration the HIOKI measuring the Fluke 731B for 60 min when the temperature in my home lab was reasonably constant (the TC of 731B is about -2ppm/C, so the 0.2ppm rise over 60 min means the temperature has dropped by about 0.1C ). The red line is 30s running average, vertical scale is 1uV = 0.1ppm /div. It is good to see two low noise references (LTZ1000A and RefAmp) working well together.
Cheers
Alex
P.S. - I've corrected the graph's title, the measurements were made at NPLC10, 1s intervals, not NPLC1 .
« Last Edit: April 25, 2024, 10:34:27 am by Alex Nikitin »
    • A.N.T. Audio
« Reply #50 on: May 07, 2024, 09:16:21 pm »
Here is another measurement result using the DM7275, in this case I was trying to do a 1000:1 transfer of resistance, from Fluke 742A-10K to my 10 Ohm DIY shunt (made using a TO-3 packaged Vishay Foil resistor). The Keithley 263 is used as 1mA current source to supply the Fluke and 10R shunt in series, the voltage on the Fluke was monitored by HP3456A and was stable at 10.00046V (double checked with the Hioki), the Fluke 742A-10K has a measured resistance of 10.00005K so the current was about +41ppm from 1mA. As a control option the ESI SR1010 resistor was used as a 10R shunt in place of my DIY one. Despite some meters of cables and numerous connections, the circuit offset at the shunt was measured below +/- 50nV average, 250nV p-p variations (the Keithley 263 has some useful features to change the current polarity and also switch it on/off so it was easy to measure offsets without disturbing the wires and connectors). The graph below shows two 15min runs with two 10R shunts in this configuration. The temperature variations (measured by Hioki) were below 0.5C). A linear trendline was added to each run to indicate a possible drift in time. Vertical scale is 100nV /div (=10nA/div, or 10ppm of value) , horizontal is samples/time , 1min/div, (=15 samples).
Cheers
Alex
« Last Edit: May 07, 2024, 09:22:18 pm by Alex Nikitin »
« Reply #51 on: May 08, 2024, 09:37:51 am »
Your 1000:1 transfer results are very impressive with the old Keithley 263 as a 1mA source !
Please show a picture of your complete setup.
    • A.N.T. Audio
« Reply #52 on: May 08, 2024, 01:01:58 pm »
I had to tidy up the cabling a bit  :D and let it settle for 10 min or so. Also today the lab temperature is somewhat higher and the HP3456A shows 10.00044V on 10K Fluke (and that is the HP tempco shows, the Keithley is stable as Hioki shows 10.000463 on the Fluke).
The Keithley 263 is a very stable and capable instrument even at higher currents, my unit is repaired and fully recapped, and for most part I use it over the newly acquired HP3245A, mostly for convenience reasons but it is not that far in DC performance and similar in the LF noise area.
Cheers
Alex
P.S. - on the photos, the first one is measuring the offset only (the current source is OFF), the second is +1mA current, the third is -1mA current. The Keithley is very symmetrical in polarity switching, due to its architecture.
« Last Edit: May 08, 2024, 01:44:15 pm by Alex Nikitin »
    • A.N.T. Audio
« Reply #53 on: May 08, 2024, 01:11:15 pm »
Here is a longer run (1mA from K263 into 10R DIY shunt) in the night time, the temperature change was about 1C over that period. The orange line is averaging for 5min (75 samples) . Vertical scale 50nV/div, horizontal - time.
Cheers
Alex
« Last Edit: May 08, 2024, 01:24:14 pm by Alex Nikitin »
    • A.N.T. Audio
« Reply #54 on: May 10, 2024, 03:31:11 pm »
Here in an interesting result with the same 10R DIY shunt and Keithley 263 supplying +/-100nA, so the voltage measured is +/- 1uV, with the HIOKI on FAST (NPLC1) speed (~30s/sec), 10 sec OFF, 10 sec +100nA, 10sec -100nA and last 10 sec OFF. The orange line is 5 samples moving average.
Cheers
Alex
« Last Edit: May 10, 2024, 03:34:10 pm by Alex Nikitin »
The following users thanked this post: chuckb, eplpwr
    • A.N.T. Audio
« Reply #55 on: May 14, 2024, 06:50:04 pm »
I rather like the temperature measuring feature. Strangely, if you copy data to a USB stick, the temperature measurements are what is on the screen, with 0.1C resolution. However if you request the data remotely, the resolution is 0.01C which does make a lot of difference! Below the results of an overnight run with the Hioki measuring the Fluke 731B 10V output and the room temperature at the same time. I plan to use the DM7275 together with a small DIY "temperature chamber" for tempco measurements on resistors and references.
Cheers
Alex
Offline EC8010
« Reply #56 on: May 22, 2024, 06:46:47 pm »
Going back to the lovely internal pictures, the difference between the 7276 and 7275 is that the black dot is in the A box on the voltage reference board whereas it's in the C box on the 7275. Looks like they graded the stability of the reference board. Which seems entirely reasonable. Perhaps more significantly, it might mean that the DMM's drift is dominated by the voltage reference rather than input stage and ADC noise/drift? If so, it might be possible to improve a 7275.
« Reply #57 on: May 23, 2024, 11:03:06 am »
There is almost a $1000 price difference between the DM7275 (20 ppm) and the DM7276 (9ppm) models
Resolutions are the same.
You might be right, same as the differences in the Keysight 3458A models.
BTW, I used the Contact Check Function of the 7276 for the first time consciously.
Offline chekhov
  • Country: by
« Reply #58 on: May 23, 2024, 11:45:19 am »
Eh, having so much space on display they show "NoCntct" instead of "NoContact", lame  :palm:
Also, from the first seconds this is a little bit of cheating on low voltage input - "guessed" DMM in 10M input mode.
If it was in auto or was a different similar class DMM - it's evident when the input is not connected by readings jumping all around. Of course, for automated measurements that could be much more versatile function.
Offline MegaVolt
  • Country: by
« Reply #59 on: May 24, 2024, 09:22:22 am »
How does it work? What would it show if the battery was working?
Offline chekhov
  • Country: by
« Reply #60 on: May 24, 2024, 10:17:43 am »
Here is a little description from 158 page of user manual, provided in the beginning of this thread:
2238148-0
Offline MegaVolt
  • Country: by
« Reply #61 on: May 24, 2024, 10:43:08 am »
(Attachment Link)
If we put a thin insulator between the terminal and the battery, will the test pass?
Offline wanghar
  • Country: cn
« Reply #62 on: July 06, 2024, 10:10:09 am »
How to calibrate a DM7276/7275?  Does this descriped anywhere in the manual?
    • A.N.T. Audio
« Reply #63 on: July 06, 2024, 07:25:59 pm »
Calibration is described on page 212 of the user manual, however Adjustment (next page) is password protected and not accessible. It would be good to know the adjustment password!
Cheers
Alex
The following users thanked this post: wanghar
Offline wanghar
  • Country: cn
« Reply #64 on: July 08, 2024, 10:01:56 am »
 Is this Hioki's style that not allowing users  to do calibration  by themselves?  :o
    • A.N.T. Audio
« Reply #65 on: July 08, 2024, 05:33:14 pm »
Not just that, so far I can not get any calibration information on my new meter from them or from the seller (Farnell UK) :palm: . And I am not even sure where I could send the meter for calibration! (OK, calibration is simple and any decent calibration lab should be able to do it, however if the meter needs adjusting, it might be a problem) .
Also the 3-year manufacturer's warranty on these meters is not transferrable from the original owner, so if you sell your meter or try to buy one second hand, there is no warranty even for a relatively new meter. Don't get me wrong, I like my DM7275 , use it all the time and won't swap it for almost any other meter, including the HP3458A, however the customer service from Hioki Japan is essentially non-existent and even from the distributor I can not get much sense so far in regard of calibration data.
Cheers
Alex
« Last Edit: July 08, 2024, 08:49:22 pm by Alex Nikitin »
« Reply #66 on: July 08, 2024, 07:33:08 pm »
I had tried a few times to the password from Hioki Japan but so far I had no luck.
The following users thanked this post: Alex Nikitin
    • A.N.T. Audio
« Reply #67 on: July 12, 2024, 02:12:38 pm »
I've measured the DCV output of my HP3245A from -10V to +10V in 1V steps with the HIOKI DM7275. The measurements were made at 10V range, SLOW speed, 10 averages per point, with the input cables polarity reversed for the second run. Vset = voltage output setting on HP3245A, V = measured voltage with the correct polarity, Vr = measured voltage with the reversed polarity. V+Vr is the difference between measuring the same output voltage in two polarities. There is a small meter/cables offset about -2.5uV which I did not null. Vertical scale is 0.5ppm = 5uV per division.
Cheers
Alex
« Last Edit: July 12, 2024, 02:21:16 pm by Alex Nikitin »
    • A.N.T. Audio
« Reply #68 on: July 13, 2024, 04:05:46 pm »
And here is a 24h measurement run of 10V output from HP3245A, there is a visible tempco of about -0.5ppm/C for the two units and connecting wires. The room temperature (orange line) varied from 24.5C to 25.7C. Right now I'm running the same measurement for 1V output.
Cheers
Alex
« Last Edit: July 13, 2024, 04:14:45 pm by Alex Nikitin »
Offline aronake
« Reply #69 on: November 26, 2024, 06:36:45 pm »
Has anybody managed to figure out how to adjust a DM7275/DM7276 except for using the mentioned scaling hack?
Offline aronake
« Reply #70 on: November 26, 2024, 06:41:56 pm »
The adjustment menu is mentioned on page 203 in the manual. But they also say: "The adjustment screen accessible on the MENU > SYSTEM screen is used by Hioki for repair and adjustment purposes. It is not for customer use."
www.itm.com/dm7276-01/manual/dm7276-01-manual.pdf?srsltid=AfmBOootKJ4X2_ElRKbZ8K__E4QlMGcIEXuv_FPCWucDw7tLfBVfxRff
Offline nwind
« Reply #71 on: May 17, 2025, 06:19:56 pm »
Did anyone figure out what LTZ1000 temperature is set?
Offline laichh
  • Country: my
« Reply #72 on: Today at 04:14:00 am »
Likely hiding behind the resistor network R4008.  :palm:
Offline nwind
« Reply #73 on: Today at 04:57:46 am »
Would be quite interesting to know for long term stability
Offline Andreas
« Reply #74 on: Today at 05:44:00 am »
R4008 looks to me as one resistor with 4 (2 pairs) of pins.
The 1K resistor should be R4005.
P4004 is the test point at the tap. (Setpoint voltage)
with best regards
Andreas
Offline laichh
  • Country: my
« Reply #75 on: Today at 05:50:50 am »
R4008 is an RM3225 resistor network by Susumu.
https://www.susumu.co.jp/english/product/category.php?cid=2