On your configuring your lab, bench, and planning - typically, you'd expect the 732A to be the traveling 10V reference - carrying the precious low-uncertainty 10V from a suitable (very important aspect) lab to your environment - that then can, as Frank suggests, serve aligning your 5440 with the truth of the lab you used.
While evaluating that pursuit, make sure the TUR and all credentials of the labs you use are proper and exceeding your own references. I've realized very quickly that my local preferred lab cannot calibrate my standards from a certain point of my trajectory (and standards at hand), and I've filtered the labs I trust vs. those I don't by how honest they are with their claims and how forthcoming with their assessments. The lab I like, but can't really use, is the one that volunteered all the relevant information, showing both competence and honesty in the process. I'd effusively recommend them any day for the appropriate application.