Electronics > Metrology
Hysteresis with LM399 and ADR1399
NaxFM:
Hi everyone, I'm trying to develop ad ultra stable reference using the ADR1399, I've been studying circuits for weeks, even those using the LTZ1000, and I'm getting everything figured out, also thanks to countless simulations (I can't believe how useful LTSpice is)
However, there's one thing that bothers me: hysteresis effects. I know that to have an extremely stable voltage one must keep the device powered on like the famous Fluke 732 standard, but that would be very inconvenient.
I was wondering if you happen to have any information on the hysteresis effects with the LM399 or the ADR1399, especially the effects of repeated power cycles on the long and short term stability of the zener.
I have already ordered an evaluation board to do some experiments, but i really don't have the equipment right now to start counting ppms. All I have is a calibrated 34401A, but soon i'll buy a DMM7510 or even a 3458A if I find a good deal.
Andreas:
Hello,
cannot tell for the ADR device since I have no measurements up to now.
Especially if you want to use the SMD version which might have a influence from the PCB due to humidity swelling.
My LM399 have rather low hysteresis during normal operation at room temperature.
Even a power outage of ~10 days (see around day 2840 in the diagram) does not change much.
The larger problem is to sort out the devices with "popcorn noise". (See the dark blue device in the diagram).
with best regards
Andreas
NaxFM:
--- Quote from: Andreas on August 17, 2022, 05:11:00 pm ---Hello,
cannot tell for the ADR device since I have no measurements up to now.
Especially if you want to use the SMD version which might have a influence from the PCB due to humidity swelling.
My LM399 have rather low hysteresis during normal operation at room temperature.
Even a power outage of ~10 days (see around day 2840 in the diagram) does not change much.
The larger problem is to sort out the devices with "popcorn noise". (See the dark blue device in the diagram).
with best regards
Andreas
--- End quote ---
Thank you! This was exactly what I was looking for.
It looks good, except for the blue one which is really bad.
Interesting to see that some of the little spikes occur at the same time for all the devices.
Do you have any explanation for that?
Andreas:
Hello,
no measurement instrument is perfect.
So probably some common mode effect like popcorn noise / (not completely compensated) temperature drift on the measurement instrument.
Standard deviation of the instrument used (ADC13) is around 0.25 ppm for a 7V reference.
with best regards
Andreas
Chance92:
--- Quote from: Andreas on August 17, 2022, 05:11:00 pm ---Hello,
cannot tell for the ADR device since I have no measurements up to now.
Especially if you want to use the SMD version which might have a influence from the PCB due to humidity swelling.
My LM399 have rather low hysteresis during normal operation at room temperature.
Even a power outage of ~10 days (see around day 2840 in the diagram) does not change much.
The larger problem is to sort out the devices with "popcorn noise". (See the dark blue device in the diagram).
with best regards
Andreas
--- End quote ---
Hi Andreas. Are these measurement taken automatically?