I bought my device end of last year from eBay.
It is built in 89. with a main controller 03457-66511 Rev.B and an AC board labeled 09-8950.
Playing around with it, i noticed that in ACV, on all ranges,
whith shorted input, i had a quite "big" reading.
In the ACV 300V range this was something like 0.4V
The AC measurement in the 3457 is done by an RMS to DC converter the AD637 (U305).
I measured a small DC voltage at pin 13, which was present even with shorted inputs.
If only AC were needed, a coupling capacitor would solve the problem.
But in 3457 there is also the AC+DC mode and this is propably the reason for the DC coupling.
Searching around i found that i could bring it to zero by changing slightly the input offset
of the U308 opamp.
This improved the situation significantly.
This opamp is part of a DC canceling feedback for the input of the AD637.
For ACV it is configured as an integrator with a time constant
τ1=200k*C323 or τ2=200k*(C320 C323) depending on Q300 (HFAC).
(This is propably controlled by ACBAND command).
The "fast" AC mode gives 4.4ms, the slow AC mode 48.4ms.
Again this is not an "AC-path".
To make it short, the next thing was to put a socket in U308 and replace it.
I had a chop OPAMP in my drawer the MAX420. The two 100nf caps, which are necessary for the MAX,
were placed and wired on the tot of it.
An absolute 0 reading, even with 0V DC, cannot be reached, as the input is relative wide (>1MHz) and
all the noise will be treated as an AC signal.
With shorted inputs, i have the following readings after the change:
ACV 30mV : ~6uV
ACV 300mV: ~25uV
ACV 3V : ~0.3mV
ACV 30V : ~2.4mV
ACV300V : ~25mV
All above values were ~15 times higher.
The high-end values and AC response are not influenced by that change.
This "tweak" makes only sense, if the low end readings are unacceptable high