I got a 5065A some time ago, the unit was not used for quite some time, and with unknown working status. But it was nearly for free.
Befor switching it on, I replaced all the tantaliums, most of the did not leak at all, and there was no damage on any of the PCBs. The unit is a late one, serial 2740A01540. It has already the 10811 crystal oscillator.
Then I switched it on (yes, the special power cable came with the unit), and let it warm up for 2 hours.
I could not get it to CO, so I did the cell flooding procedure from the manual for a week, Nothing changed.
Next I checked the lamp assembly, and the 1.33KOhm resistor was open circuit. After replacing it with a 1,3kOhm, feeding it with 20V, the lamp started after some seconds. and the assembly took 137-140mA. I did polish the mirror.
But still nothing changed.
Readings were now:
Supply 40
Lamp Oven 38
Cell Oven 40
Osc Oven 36
Photo I 14 (was the same with the non-working lamp)
5MHz 16
Control 0
Error 0
2nd Harmonic 6
1 MHz 38
100KHz 40
But I do not see anything on A7 TP2, just very little noise. and it does not go into continuous operation.
Assuming it did work until the 1,33KOhn resistor went open, I think it should run.
Any suggestions?
Thanks,
Peter