Electronics > Metrology

Kalibrierlabor - Calibration LAB

(1/5) > >>

Nuno_pt:
Morning all,

While looking at some stuff relating to Calibrations, I've found this company in Germany "kalibrierlabor.de", does anyone know about this company?

Since I don't speak German, from what I can see on the Accreditation PDF < http://kalibrierlabor.de/urkunde.pdf > they use an HP3458, Datron 4700 (from what I see this is only good till 6 1/2 digits), Fluke 5220, but nothing else.

So with curiosity I've ask a quote to my K2000, and they quote it at 117.81€ with VAT, according to ISO 9000 (I can upload the PDF of the quotation if anyone want to take a look), but what I can't see is if they provide the data before and after the calibration and if that is on the price.

The prices seem a little on the low side.

Any input on this?


e61_phil:
The uncertainties seems to be quite high.


"Zu jeder Kalibrierung erhalten Sie einen Kalibrierschein nach DIN ISO 9000 ff mit einem Protokoll der Messwerte."

They provide the data.

Nuno_pt:
e61_phil, that was what I seeing and trying to figure out.

Why the uncertainties are so high? This can be related to the Datron and Fluke?

Kleinstein:
The uncertainties come from the instruments and how often and where they send there instruments for verification. If they are more into checking lower precision instruments at a lower precision instrument for a low price, this might be good enough. In Industry not many instruments need to be that precise.

Nuno_pt:
Just to compare I ask Tektronix in Germany a quote for the K2000, they quote 285€ with VAT for Cal + return shipping to Portugal, and 50€ for Data only.

So if I take about 35€ that is what should cost for Tek to send the K2000 to Portugal, the Tek price is 250€ with VAT + 50€ for the data = 300€.

Making the same math  117.81€ + 70€ (one shipping to Germany, one return shipping to Portugal) = 177.81€, so 177.81€ with VAT and Cal data.

The difference is 122.19€, would this make bigger difference?

I know you've to pay the brand of Tek, and of course low uncertainties, etc...

From the uncertainties I guess that for hobby use and 6.5d DMM this could be a low cost solution, but for the big boys 7.5, 8.5d DMM it's better go with the low uncertainties and with the brand.