If only for fast pulse generation, then what about Leo Bodnar's pulse generator?
+1
The main use of the fast rise pulse from the 284 is to perform the mid and high frequency compensation of the scope's vertical amplifiers (and maybe the probes as well). For that purpose, Leo Bodnar's pulse generator will work very well. It doesn't matter that it is a 10 MHz 50% square wave instead of
200 50 kHz. The rise time and pulse shape/flatness is most important. The output voltage is rather limited, only 1 Vpp back termiated into 50 ohms, or 2 Vpp not terminated, but that's higher than the 284 could do. You may need inline 50 ohm attenuators for use with the lower voltage vertical settings.
edit: I have a HP 8131A pulse gen that would work as well. The rise time is fast at <200 ps, but not quite as good as the Leo Bodnar unit, and it is a few orders of magnitude bigger. It has the benefit of adjustable repetition rate and width, as well as adjustable output voltage and offset, so it could be set up to emulate the 50 kHz, 1 us, 200 mV pulse of the Tek 284. It is digitally controlled and plenty accurate enough to use to calibrate the timebase/sweep on any analog scope. If you (or anyone) is interested in such a beast, let me know... it is one of quite a few items slated to fall victim to a general lab cleanup/downsize (mandated by the facilities manager, aka wife).