Seriously? Calibrating the whole unit once ACAL was encountered is the solution provided? Wow. 
-branadic-
Hello branadic,
no it's meant differently.
ACAL only works correctly, if a proper DCV calibration (CAL) was done before.
Once working correctly, ACAL will always work correctly, w/o another CAL.
Some units failed, for unknown reasons, maybe CAL was not done right. In these cases ACAL caused an erroneous jump up to 20ppm.
Therefore, a new CAL does cure this problem on these units.
The ACAL procedure seems to work like this:
During CAL of the 5 DCV ranges, the nominal gain factors of the ADC / REF circuit, of the 100:1 divider and of the x10, x 100 amplifier are determined very precisely .
These gain factors will drift over temperature and over time.
ACAL then precisely re-measures these actual gains and corrects to the nominal ones, by using the internal voltage reference as a stimulus.
Such 10:1 ratio re-measurements are normally not possible to do precisely enough, when using an ADC which is linear to 6 digits only.
So they probably have invented something to make this transfer 10 times more precise, at these special calibration points... maybe by additionally determining the non linearity at these fixed ADC points during the calibration procedure.
This technique is also used in a similar manner inside the 5440B calibrator, which also requires to once calibrate the nominal ratios for the different ranges.
Frank