Thanks
@rodpp, 5 ppm stands here just for 5E-6 , and yes this is a very low Q value.
a) It is still interesting to notice that the required Q precision is asked at 25 ppm (as stated by the manual) and that at the same time, Genrad provides examples of values that are 100 X lower (5K97 -> Q= 0.22ppm).
b) I had a look to what kind of TE could be used to calibrate your reference resistors.
The precision
E4980A LCR meter (successor of the HP4284a) seems to be unable to measure such low Q values.
The datasheet (see page 17, tables 23 to 25) tells that the TEs 'phase angle of impedance θ' is calibrated with a precision of 1E-4 radians at 1kHz for impedance ranges from 0.1 to 1 k Ohms. However, the displayed mesurement range starts much lower at ±1 arad (page 7).
If we take the previous example (R =375 Ohms, X = 1.875 mOhms, Q = 5 ppm) the angle of impedance θ is 5 µrad, so quite 2 orders or magnitude lower than the precision of calibration of the E4980A.
I don't know if there's a way to make a kind of relative measurement to take advantage of the very small display resolution provided by the E4980A.
c) I would also agree with
@zrq when he suggests to raise up the measurement frequency. This is also mentioned in the article he linked to (see page 6).
1. Find resistors of the same type and dimensions as the reference resistors inside the instrument. GenRad used those in the calibration resistors too.
d) Of course, gathering the genuine feedback resistors of a defective unit would be the best choice. As the 1689 seems to accept different cal resistors values than those in the genuine cal kit, a less than perfect way could be as follows :
- Design the PCB that fit into the standard Digibridge connector (4 wires + guard ?). Connexion areas have to be gold plated for best long term corrosion resitance
- Buy the best precision resistors you can find for an acceptable price
- Meet all of the soldering requirements (reflow curve for SMD); Use leaded solder with 4% Silver for through hole parts
- Assemble your own cal kit (4 standard resistors + O + S). Additionnal or duplicates values can serve as 'verification kit' for extended confidence
- Optionnaly build a 3D printed bracket + brass hand screws that holds the boards vertically
- Optionnaly add some potting around the resistors and PCB to avoid any moisture ingress (before potting, 'bake' the devices in your kitchen kiln for 24 hours at 50°C) (*)
@Electrole gave you very nice resistor combinations to build your cal kit. Here are some others (not so close to the genuine values).
I selected the best precision resistors from Mouser (Tolerance <= 0,01 % ; Tempco <= 2 ppm/°C) so that you gain simultaneously a reasonable long term stability feature (ideally not worse than ± 0.01 %/year). The manufacturers provide reliable long term stability data only for (some)
high precision resistor series.https://www.mouser.fr/c/passive-components/resistors/?resistance=20%20Ohms%7C~350%20Ohms~~400%20Ohms%7C~6.8%20kOhms%7C~9%20kOhms%7C~20%20kOhms%7C~25%20kOhms%7C~68%20kOhms~~100%20kOhms%7C~1%20MOhms&temperature%20coefficient=0.2%20PPM%20%2F%20C~~2%20PPM%20%2F%20C&tolerance=0.005%20%25~~0.01%20%25&instock=y&rp=passive-components%2Fresistors%7C~Temperature%20Coefficient%7C~Resistance%7C~ToleranceHaving 0.01 % resistors is nice but it's still far from the 20 ppm precision asked by the SM. So a measurement of your DIY cal kit with an 'in specs' 8.5 digits Ohmmeter is probably needed.
To replace the 95.3 kOhms standard, use 68 k + 25 k = 93 k (100 k // 1 M = 90.91 k might also be interesting if 100 k is 0.005 %)
https://www.resistance-calculator.com/advanced-resistor-calculatorTo replace the 5,97 kOhms standard, use 6.8 k (or 9 k // 20 k = 6.21 k)
To replace the 374 Ohms standard, use 400 Ohms 0.005 % (350 Ohms 0.01 % is a bit cheaper)
To replace the 24.9 Ohms standard, use 20 Ohms
Have also a look at other distributors (Digikey,
Newark, RS) if they source other values or better tolerances.
If the parasitic inductance value is provided in the datasheet, a ballpark of the impedance at 1 kHz could be calculated.
Keep your cal kit in a ziplock bag. Add some
silica gel to avoid any moisture.
For about 300 bucks, you should be in business with a DIY cal kit that is an acceptable replacement of the genuine IET Labs/Genrad item.
Maybe other forum members could be interested in having/buying/building such a cal kit if you publish the Gerber and STL files.
(*) at PTB, they dried the parts with deep vacuum. Personnaly, I would have been a bit reluctant in vacuuming the resistors, as the low pressure could alter their internal structure. But as they did it that way, it's probably safe. At least for SMD parts.