Well, I did it. I stuck the Keysight display window in.
More shiny!
...
I'll install it Sunday night, once the timer for one week continuous operation has elapsed as specified in the drift calculation in service note 18A, so I don't ruin the test.
Then I can compare ACAL constants to check for the common drift issue, at least for the 1 week test. Longer testing will follow once it's installed back in my rack.
Congrats, and have fun!
As Ian is currently in the thread, let me ask again, whether your 18A drift measurements of U180 (by means of Ians program) take its temperature coefficient into account, i.e. compensate for this effect?
I guess, I'll append my latest measurements for comparison. The U180 drift nearly vanished over the past few years.
Frank
I always monitor CAL? 72 (DCV gain) and CAL? 175 (ACAL DCV temperature).
The first diagram then calculates the T.C. of U180, which fits to the 3 T-points taken during warm-up.
The T.C. was later confirmed by separate T.C. measurements of the overall T.C. minus the T.C. of the LTZ1000A reference.

When I plot the U180 gain constant over time, and substract the T.C. influence, I get a seemingly consistent, temperature-compensated drift curve for U180.
In 2019, I made the latest 10V DC calibration, which automatically gives a small jump of CAL? 72.
The drift was already very small, 0.5ppm/y or 0.002ppm/d and practically vanished for the last 2 years.