Yes, please.
Show us some pictures from the inside.
OK, quick teardown then.

It's a very simple device. 4 binding posts for force and sense, L&N made rotary switch and a bunch of wires/bars inside.

I got my unit from eBay very cheap ($45 total), so I'm not surprised it has a long history. High ohm resistors look a bit burned, probably someone tried to push too much current. But nothing is broken and the "event" happened probably long time ago so these resistors aged again.


I've measured them with my HP 3457A (NPLC=100, NRDG=10, OCOMP, AZERO) and Keithley 2001 (NPLC=10, OCOMP, AZERO, LINESYNC etc), nothing calibrated at a lab since I got these meters from eBay:
Range [Ohm] | HP 3457A Mean [mOhm] | HP 3457A StdDev [uOhm] | K2001 [Ohm] |
0.01 | 10.128 | 43.57 | 0.010000 |
0.02 | 20.103 | 45.47 | 0.020008 |
0.05 | 50.138 | 27.66 | 0.050006 |
0.10 | 100.127 | 44.48 | 0.100007 |
0.20 | 200.148 | 55.73 | 0.200056 |
0.50 | 500.116 | 50.82 | 0.50067 |
1.00 | 1000.901 | 42.52 | 1.000823 |
2.00 | 2000.852 | 53.89 | 2.000906 |
I've also tested this 0.001 Ohm, 1%, 25ppm/C shunt:
http://uk.farnell.com/powertron/fhr-4-2321-0r001-1-q/resistor-metal-foil-0-001ohm-3w/dp/2419239Range [Ohm] | HP 3457A Mean [mOhm] | HP 3457A StdDev [uOhm] | K2001 [Ohm] |
0.001 | 1.138 | 50.72 | 0.001008 |
I don't have an SMU, but I got curious how the L&N 4385 will behave under load (as multimeters only source ~1mA (3457A) - ~10mA (K2001) for 4W Ohms on this range), so I've selected 1.5A range (0.1 Ohm) and sourced 1A from Agilent 6632B, measured with K2001 as: 1.0007006A, connected that to the shunt box (assumed cold, as 1-10mA from previous measurements probably didn't heat the resistors a lot), sense outputs to K2001 in DC mode, with 1 NPLC and the voltage started with 100.120 mV and after 20 seconds stabilized at 100.105 mV. Therefore, assuming the current from 6632B stayed the same the resistance went from 99.423 mOhm to 99.408 mOhm, a delta of 14.89 uOhm. Now it is impossible to say anything, K2001 current ranges are not calibrated and not very accurate, 6632B was stated cold and might be unstable and the 14.89 uOhm delta is difficult to measure with my equipment anyway (see the 3457A StdDev).