I have hard time getting standard LTZ1000 but I managed to order one ADR1000 from Analog Devices site which should perform similarly.
I came up with R4 12k, R5 1k, R2 61.9k, R3 61.9k, R1 100ohm. Did I get this right, or should I change something?
I am also planning on designing my PCB, but I cannot find too much information about layout recommendation.
Hello,
Each single device behaves different regarding ageing.
So you might want to have at least 3-4 stable references to sort out those with the largest ageing.
Whether ADR1000 behaves similar to the LTZ1000 still has to be proven in practice.
At least it seems that the ADR1000 needs longer time to settle than a typical LTZ1000.
The resistor values are those recommended from the data sheet so it should be ok to go with them.
Others try to set the temperature setpoint to the zero T.C. temperature of the ADR1000 to reduce influence from set poin divider resistor drift.
(this is not possible with LTZ1000 due to the higher zener voltage).
For layout it is essential to do star point wiring and especially kelvin sensing of the zener voltage.
(no ground currents from heater or OP-Amps across the ground sensing line).
When you have to integrate the zener in a device like 3458A it might be partially easier to work with ground current compensation.
So the constraints (dual/single supply) and pre-ageing and selecting on device / module level may give different board layouts.
with best regards
Andreas