Well,
I haven't reported about my ADR #3 for a while, because I switched it off after June 18th.
In between, I got a new baseline comparison @ ab precision, and after several iterations at his lab, the mean value of my 8 references agreed within 0.25ppm with his 732B reference, which itself has 0.3 uncertainty, confirmed by a fresh calibration of the 732B afterwards. You'll see the baseline point in the diagram, which also is another confirmation for the recent ring comparison, branadic had organized around June 2022, with similar uncertainties by each Philipp and ab precision.
At first, the ADR #3 had stabilized before the shut-off. I powered it on again on 30th November 10 p.m., letting it warm up with the 3458A and the 5442A over night, for about 12 hours. So the transfer measurements showed a +5ppm shift, very huge compared to the other shifts, when I switched the ADR1000 off for several days or weeks.
Anyhow, a measurement on December 16th showed, that the ADR 1000 evidently returned to his recent voltage within a few tenth ppm.
All of my LTZ1000 and LTZ1000A, especially in my 3458A which is only switched on for measurements, never ever showed such a shift after being powered down.
They all return within 2h at most to their nominal value.
This effect - if it turns out to be inherent to the ADR1000's technology - is not convenient for usage as a intermittently used reference, like in DMMs or cold travelling standards.
I switched on my equipment on Thursday, 14th, and the next morning, I found my Fluke 5442A got a defect, the processor PCB obviously freaked out, so I have something to do Between The Years. Strangely, the transfer measurement of my other baseline references revealed, that my oldest prototype references Ref_1 and Ref_2, which were still not optimized and not in a tuner box enclosure, like all the others, both show an unexpected +0.4 ppm shift.
Something has happened during the night, damaging the 5442A, and influencing both the references.
By chance, the Solar outbreak just has begun on 14th December, but that influence is quite improbable.
Maybe there was another E.M.C. impact, or a short circuit occurred inside the 5442A, like a burnt tantalum, which impacted the references. We'll see.
Frank