Author Topic: Lowest drift, lowest noise voltage reference (ADR1000AHZ)  (Read 176327 times)
Offline branadicTopic starter
  • Sounds like noise
« on: June 27, 2021, 07:25:56 am »
Hi folks,
after I played with the simplified version of W/F7000 schematic with LTZ and achived -0.0106 ppm/K after t.c. trimming for the bare ovenized zener, I now played with a very new, lowest drift, lowest noise voltage reference I had on the desk. Actually I already had it for some months already, but it always takes some time to get all the components and board design ready.
I treated it the same way as mentioned in the linked thread before, thus first setting the z.t.c. temperature to a specific temperature with the oven turned off. Next step was to turn on the oven and adjusting the oven to that specific z.t.c. temperature.
I then captured some low frequency noise data with a 0.1 ... 10 Hz LNA (80 dB, 100 nVpp or 16 nVrms) and an Advantest R9211E used as a DSO.
Afterwards I've measured the remaining t.c. of the board and trimmed the t.c. in multiple steps. The last trimming was slightly overcompensating things., turning the t.c. into opposite direction. The measurement is already limited by the lab temperature stability and the t.c. of the meter. Thus, it needs another run of adjustment and another run in the thermal chamber plus comparison to F7000-2 in the next run, but results are looking good so far.
Compensating the ambient temperature influence, I can find the t.c. already being at -0.03713 ppm/K, but I'm sure we can do better than that.
The board is currently missing the gain in the boost stage, but uses the amplifier as a buffer only. This is part of the next steps that will be performed soon.
-branadic-
« Last Edit: December 22, 2021, 08:17:30 pm by branadic »
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
  • Sounds like noise
Re: Lowest drift, lowest noise voltage reference
« Reply #1 on: June 28, 2021, 03:41:35 pm »
After some tweaking I'm now down to about -10 ppb/K. With some more iterations I'm sure I could get it even smaller, but hey the result is not too bad, or is it?
-branadic-
The following users thanked this post: quarks, Dr. Frank, AlienRelics, notfaded1
Offline Kleinstein
« Reply #2 on: June 28, 2021, 04:16:16 pm »
The temperature effect looks good. However this is the easy part - expecially for a rather limited tempearture range.
The noise performance also looks really good, though the time window is relatively short to really tell. Sometimes popcorn noise happens on a long time scale and some 100 seconds without a jump are well possible even for a LM399. The standard to LF noise charactrization is 0.1 to 10 Hz, because this is a window that can still be measured with AC coupling. However the LF noise that really matters is often the even lower frequency part, like 1-100 mHz. This is especially the case for a precision DC voltage reference. This is really hard to measure - so it is very understandable to measure the sharter time scale. It could still help to have a longer window.
  • Sounds like noise
« Reply #3 on: June 28, 2021, 05:39:00 pm »
At least in the observed time frame, a t.c. measurement takes about 10h 40min, no popcorn noise was present.
The only real measurement setup for very low frequency noise I know of is propably a JVS and a nV-meter. As the LNA already limits the bandwidth to 0.1 ... 10 Hz, longer time frames with this setup don't make any sense at all.
-branadic-
« Last Edit: June 28, 2021, 07:25:44 pm by branadic »
Offline Kleinstein
« Reply #4 on: June 28, 2021, 06:45:32 pm »
I know the problem with the very low frequency noise. One could do a measurement with 2 identical reference, but it is quite some effort.
The TC measurements take quite some time, but the vertical resolution is not as good and the reference in the meter is involved too. So hard to tell which ref. is causing the change.
Offline ramon
  • Country: tw
« Reply #5 on: June 29, 2021, 12:50:34 am »
I usually liked to make measurements for 1 month at ambient temperature.
  • Sounds like noise
« Reply #6 on: June 29, 2021, 03:59:49 pm »
I repeated the measurement with my F7000-2 sitting at room temperature in parallel. I can state with some confidence, that the reference shows no obvious t.c. in the range 15 ... 31 °C. Both, the LDLN reference in the thermal chamber and the F7000-2 at room temperature, track each other.
-branadic-
The following users thanked this post: quarks, Dr. Frank, wolfy007, notfaded1
  • Sounds like noise
Re: Lowest drift, lowest noise voltage reference (ADR1000AHZ)
« Reply #7 on: August 25, 2021, 09:14:05 pm »
I recognized that I've never published results of the final reference. After my last post I've trimmed the output voltage to 10.000 0x V and finally trimmed t.c.
The 10 V boost stage added some small higher order portion to the t.c. as already seen on my similar LTZ reference.
-branadic-
The following users thanked this post: quarks, Dr. Frank, chuckb, iMo, Okertime, jeffjmr, alligatorblues
Offline Castorp
« Reply #8 on: August 26, 2021, 09:21:53 am »
I have taken noise measurements down to lower frequencies, fully in the 1/f region. ADR1000 seems to be at least 2x quieter than good samples of LTZ1000. There's also very little spread among different units. But this is all too preliminary and there are still a number of things that prevent me from publishing the results.
So far I only have two units of HPM7177 with ADR1000, but there will be more soon. What I can tell for sure is that in those units the LF noise at full scale is no longer dominated by the Vref. With LTZ1000 it was, with ADR1000 it's some extra noise coming either from the ADC chip or the AZ op amps (but definitely NOT from the resistor networks). I also have a hint for long-term drift - both units have drifted less than 0.5 ppm in 9 months. They have been powered on and off many times, and one of them was severely abused with EMC tests (conducted RF, EFT, LF magnetic field). The 0.5 ppm/9months was taken with a rock-solid 732B that we calibrate at METAS every year.
I'll mention some of these findings in the coming MM talk.
Offline Castorp
« Reply #9 on: August 27, 2021, 09:41:51 am »
Snapshot of an ongoing test. Once again - that's with a 732B and only a small fraction of the LF noise comes from the ADR.
The following users thanked this post: chickenHeadKnob, TiN, branadic, Andreas, e61_phil, BU508A, chuckb, 2N3055, ch_scr, iMo, czgut, Okertime
Offline Dr. Frank
First Tests ADR1000AHZ
« Reply #10 on: August 27, 2021, 05:54:21 pm »
Hello,
Thanks to ScoobyDoo and branadic, I got 5 samples and some PCBs.
Here's the mockup board, where I made the first tests.
The lower zener voltage compared to the LTZ1000 now allows this Reference Amplifier to be trimmed to zero T.C. for a given oven temperature, by individual selection of the collector resistor R2. Compare this to the +50ppm/K for the unregulated LTZ1000.
That trimming process is the same procedure as for the SZA263 and LTFLU in the different FLUKE instruments.
I measured Uref vs. temperature, which always gives a negative, flat parable. The zero T.C. point varies from 45 to 60°C for a fixed value of R2, and the T.C. varies from about +5 .. -5ppm/K over 24 ... 80°C only.
At a fixed divider ratio R4:R5 for around 52°C oven temperature for all five samples, the individual T.C.s are well below +/- 2ppm/K.
It follows that the attenuation factors for R3 and the most critical R4, R5 are at least 10 times greater than in the datasheet. These resistors now can have much relaxed stability specifications, only R1 and R2 should be PWW or BMF types.
I also measured the low frequency noise 0.1 .. 10Hz, like branadic and Castorp.
I use an LNA from Andreas, and always made "typical" measurements of 10sec long samples, so to have a direct comparison to my LTZ1000s, but also to the measurements done by Andreas.
100s measurement time or averaging give no essentially different results.
Other LNAs will give different absolute noise figures, so these results are not directly comparable .
The noise of the ADRs RefAmp evidently is about 50% of the LTZ1000, which several others also have confirmed.
ADR #1 would be the most quiet sample, but it suffers from  :popcorn: noise, or frequent dips, but with a very moderate amplitude of maybe 400nVpp.
The ADA4522 buffer contributes a lot to the overall noise, so I have to investigate if it's only this specific unit, or what can be done to improve this.
I will add the noise figures of these samples, when measured with my 3458A, maybe now I can see a pronounced difference.
As my 34465A showed much higher stability / lower noise when I replaced its LM399 reference by a LTZ1000 clone PCB, my idea is to replace the LTZ1000 in my 3458A with an ADR1000 (*).
Maybe its noise figures will also go down, best by a factor of 2.
Castorp found that in his DVM other components overwhelmed the noise budget, so let's see what happens.
I would like to remind that TiN some time ago made a similar experiment on one of his 3458A, by implementing 4 parallel LTZ references, which would also yield 50% noise of a single LTZ.
I'm not aware what his final result was.
Anyhow, I think this new device will require quite a big time slot during the Metrology Meeting 2021, and I'm excited to meet the volt-nuts community personally and we'll have some fruitful discussions and comparisons   ;)
Frank
(*) my 3458A already has a modified FW, with CAL? 2,5 = 6.50000V (7V reference lower limit).
Thanks to Poul-Henning Kamp for your advice how to correct the checksum.
« Last Edit: September 06, 2021, 04:21:29 pm by Dr. Frank »
The following users thanked this post: quarks, branadic, Mickle T., Andreas, BU508A, chuckb, 2N3055, ch_scr, iMo, Okertime
Offline Kleinstein
« Reply #11 on: August 27, 2021, 06:59:10 pm »
Directly buffering the reference with an AZ OP can cause problems from input current spike of the OP effecting the reference circuit that uses a relatively slow OP. The usual solution is to have some filtering of the reference voltage. The OP is high input impedance and one can thus use a simple RC low pass, at least to filter out the spikes (MHz range) but also a little more (e.g. 5 K and 1 µF).
I don't think it would be really needed to fine tune the unheated TC. It is enough to start with a typical much smaller value like 5 ppm/K instead of 50 ppm/K. A factor 10 less sensitive resistors is already a big step forward and should be good enough. For the tests it is of cause good to have the unheated TC measured.
Using the ADR1000 in the 3458 may not be so easy, as the voltage is lower and the meter may very well have points to complain about too low a votlage and CAL constants out of range.
  • Sounds like noise
« Reply #12 on: August 27, 2021, 08:20:27 pm »
There is some R-C-R filter in front of the AD4522-1, though Frank uses slightly different values than I do ;)
Once the unheated reference is fine tuned the demand for the resistors with heated references decreases dramatically, same as for all t.c. compensated references (LTFLU, LTZ, 1N829, 2DW23x, ...) with only one exception, LMx99 which is more than just a zener. This is true for ADR either way.
-branadic-
  • Sounds like noise
« Reply #13 on: August 29, 2021, 03:20:57 pm »
I'm constantly monitoring the ADR for 26 days now. Every morning the skript first performs an ACAL, measures the amplified ADR reference with R6581D, the difference to F7000-1 with K2182A-1 and the difference between between F7000-1 and F70002 with K2182A for an hour.
The reference is currently drifting downwards. Allthough some people might think this is due to the resistor network in the boost stage, I tend to believe it's the reference itself. But I don't want to bother the setup and add additional equpiment to prove that.
-branadic-
The following users thanked this post: chuckb, czgut, Okertime
Offline ScoobyDoo
« Reply #14 on: August 29, 2021, 05:06:49 pm »
Hello BRANADIC - You have probably observed the LTD graph from its datasheet.
If you power the ADR1000 without an accelerated ageing process - it will take at least 5-6 months before it settles to an acceptable level. It is possible to age the ADR1000 (as done with LTZ) but ageing process needs modification to get it working ... - the rest is proprietary recipe from the Mastercook.
cheers
ScoobyDoo
Offline Kleinstein
« Reply #15 on: August 29, 2021, 05:24:36 pm »
The curve for aging is with 75 C set point temperature. With a lower temperature the time scale is expected to be even longer.
  • Sounds like noise
« Reply #16 on: August 30, 2021, 08:17:39 am »
I'm aware of that diagram, but since the recipe of accelerated aging is a secret, it doesn't help a lot to know that there is one. The only recipe available by now for LTZ was given by Cern.
https://xdevs.com/doc/CERN/1101699_V1_Burn_in_of_LTZ1000.doc
https://cal.equipment/doc/CERN/1101699_V1_Burn_in_of_LTZ1000.pdf
Modifying this receipt requires knowledge of the die attach being used and its glass transition temperature, an info that is not available at ADI website.
Aging means cycling around its glass transition temperature to reduce internal strain and allow further curing.
-branadic-
« Last Edit: January 06, 2022, 01:16:53 pm by branadic »
The following users thanked this post: 3roomlab
Offline ScoobyDoo
« Reply #17 on: September 05, 2021, 08:17:40 pm »
Hello Folks - this is our ADR1000 EVB - up and running.
We did some basic STDEV tests using 3458A - but running against limits (of 3458A) - 120nV STDEV (K=2) with an output voltage of 6.659V typical for the ADR1000.
The 120nV STDEV (0.018ppm) is just an indication - we expect it to be better using a setup with a null meter - but this would require an averaged ADR1000 buried zener bank or JJA/JVS.
Best regards
ScoobyDoo
« Last Edit: September 05, 2021, 08:41:34 pm by ScoobyDoo »
Offline Kleinstein
« Reply #18 on: September 05, 2021, 09:12:05 pm »
A 2nd ADR1000 refrence and than a low noise meter would be enough. This would measure the sum of the noise of the 2 ADR references plus the noise the meter in a low range (e.g. 100 mV). In the lower range the 3458 could be good enough. It should be good at least for the lower frequency part, as the DMM amplifier has little 1/f noise, but the reference is expected to have 1/f noise in the very low frequency part. A low noise amplifier (no need to be very high impedance) would be an option too.
The following users thanked this post: martinr33
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #19 on: September 06, 2021, 08:27:35 am »
Yes, what is the numerical resolution of the HP 3458A in that measurement? 100 nV? If you want to go beyond what you have as resolution or stability of your meter, you setup a difference/bridge type  measurement.
For example with the two LTFLU references i made last year, i am observing a standard deviation of day-to-day averages of about 270 nV. This is the difference of the two references and comparing one day to the next day. So there are two factors sqrt(2) involved and i estimate the noise of each reference to be about 130 nV. The combination of both LTFLUs should be a reference with 90 nV standard deviation for day averages. I am getting similar numbers for hourly averages.
Of course i would also prefer a direct comparison to a much better reference, but that will be expensive.
Regards, Dieter
Edit: The LTFLU references i mentioned include 10 V gain stages, so it's a difference between two 10 V reference voltages.
« Last Edit: September 06, 2021, 04:04:30 pm by dietert1 »
Offline Dr. Frank
Stability measurements on ADR1000AHZ
« Reply #20 on: September 06, 2021, 05:42:57 pm »
Hello,
I've performed the first two stability measurements on the ADR1000, and updated the 1h noise figures as determined with my 3458A in the table above.
It's been placed inside an aluminum case with proper jacks.
The references show a strong negative drift in the first hours already, so confirming branadics findings.
Room temperature and internal temperature of the 3458A was stable to about +/- 0.1°C, so that's fully the ADRs drift what we see here.
I have then chosen 1h of mostly constant output voltage and determined the noise (SDEV) for 1h and for several 10 minute periods.
I have to emphasize again, that these figures are always a convolution (quadratic sum) of the different noise sources like ADR1000, plus 3458A amplifier, plus its LTZ reference and plus the A/D noise.
The 1h noise (185nV for ADR#3, 168nV for ADR#1) is remarkably lower than measurements on other LTZ based DUTs.
Of course it can't be as low as 50%, as the 3458A contributes a lot to the noise budget.
This also confirms the SDEV~ 120nV measurement by ScoobyDoo, which were probably made on a shorter sampling period.
I've even seen 80..90nV r.m.s. over 1min (16 samples).
In the 2nd graph you can recognize quiet and more noisy 10min periods, ranging from 133nV to 237nV, latter showing sort of small bumps.
This seem to indicate different instability mechanisms, the first being the usual zener noise.
So what might be the 2nd mechanism?
Qualitatively, you'll find that picture identically in differential mode measurements, of course on a bit lower level, as only the instabilities of both references come into play. (10V level, 146nV @10 min., 189nV @ 1h)
So it would be interesting to run two ADRs against each other over 24h, to confirm that the noise is cut by 50%, and to investigate, whether these small bumps vanish.
In this case, the ADR could really improve the short term stability performance of the 3458A, when replacing its LTZ1000A.
(I need to change about a dozen further ACAL limit parameters inside the FW, so that the lower reference voltage is accepted.)
I guess I will also bring this mockup board and Andreas LNA to the Metrology Meeting, on next weekend!
Frank
« Last Edit: September 06, 2021, 05:48:03 pm by Dr. Frank »
The following users thanked this post: 3roomlab, iMo
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #21 on: September 06, 2021, 06:31:10 pm »
I got some 0.5 ppm jumps from first few chips runs, sadly. In these Log 1, Log 2 two socketed ICs tested starting from power on. Circuit and PCB is very same xDevs FX with 13k/1k Tset and 120R.
Logged by 3 x 3458A + 4th (red line) on 10V boosted output.
DSV-files : Log 1, Log 2
Downward drift also present in logs, as others show.
We had crazy heat here in August, so ambient temperature stability was also very poor.
« Last Edit: September 06, 2021, 06:34:50 pm by TiN »
Offline Kleinstein
« Reply #22 on: September 06, 2021, 06:38:17 pm »
There seem to be some popcorn / RTN type noise ( sudden jumps by some 0.05 ppm).  With rare event one can get intervals without and intervals with such jumps. I consider these jumps problematic, especially if they are rare, as averaging ober some 10 minutes is not very effective against such noise.
For short time tests this can be OK, but for a long time experiment you never know whether one is more on the upper of lower values.
It is still the question, if this from the ADRs or maybe the LTZ1000 in the 3458. I don't think it is the amplifier or ADC of the 3458 - though at the 0.05 ppm level INL errors may reach this, much and still be in spec. Still the main suspects are the 2 references.
edit:
TiN's cuves show the jump, at least some comig from the DUT - don't expect 3 times 3458 to jump in sync, even with something like EMI.
« Last Edit: September 06, 2021, 06:41:17 pm by Kleinstein »
The following users thanked this post: splin
Offline quarks
« Reply #23 on: September 11, 2021, 12:57:20 pm »
thanks a lot for sharing information
Offline MegaVolt
  • Country: by
« Reply #24 on: September 13, 2021, 08:39:32 am »
I seem to have missed the big ADR1000 sale :))) Where do you get them?
Offline YetAnotherTechie
  • Country: pt
« Reply #25 on: September 13, 2021, 07:10:58 pm »
They are already available in alibaba  :-DD
The following users thanked this post: felixd, MaxZ
Offline ScoobyDoo
« Reply #26 on: September 13, 2021, 08:08:55 pm »
Hello everyone - I have an update for all potential ADR1000AHZ buyers:
They can buy through ADI’s distribution arm.   It is called ADX
Analog Devices Express (ADX) | Design Center | Analog Devices
https://www.analog.com/en/design-center/device-programming.html
Also any customer can reach out to Rui Zhou or Carlos Sanchez to inquire about product availability and sales of the ADR1000AHZ.
We can provide them with a datasheet but no extensive technical support.  This is why it is not on the web.
Carlos.Sanchez@analog.com
Rui.Zhou@analog.com
Best regards
ScoobyDoo
The following users thanked this post: quarks, TiN, 3roomlab, doktor pyta, ramon, ch_scr, MegaVolt, syau
Offline ScoobyDoo
« Reply #27 on: September 14, 2021, 07:05:39 pm »
Hello Folks
                good news is coming - Carlos, Rui and ADI team are adding the ADR1000AHZ in the ADI webshop online - part will be visible within a day or so.
You will be able to pay with credit card - no credit letter is needed
To be updated soon !
Best regards
ScoobyDoo
The following users thanked this post: chickenHeadKnob, TiN, 3roomlab, RandallMcRee, ramon, MiDi
Offline Andreas
« Reply #28 on: September 15, 2021, 04:34:21 am »
but who wants the SMD plastic package?
Mouser germany also lists the device (0 Stock limited availability)
https://www.mouser.de/ProductDetail/Analog-Devices/ADR1000AHZ?qs=%2Fha2pyFadugtPWjaDttya%2FKT5U0JA9UT00v1%252BPGnc%252Bo%3D
has someone got already a offer from them?
with best regards
Andreas
Offline quarks
« Reply #29 on: September 15, 2021, 08:48:31 am »
not yet
EDIT: today Mouser wrote me they cannot make me an offer for ADR1000
« Last Edit: September 24, 2021, 09:40:18 am by quarks »
Offline ScoobyDoo
« Reply #30 on: September 16, 2021, 09:11:46 pm »
Hello folks
                The ADR1000AHZ is now available from the ADI online webshop.
The MOQ has been set as low as one to reach all Voltnuts - (MOQ = 1).
The part will be monitored for a period of at least six months to understand if this is a viable (sustainable) path for ADI - so pls. order as much as you can ...
ADI might raise the MOQ to 5 coming days ...
You need to be logged in in order to view the part and price
Best regards
ScoobyDoo
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #31 on: September 17, 2021, 04:10:23 am »
Inspired by chuckb and his light show experiments I've reattached camera onto microscope over my decapitated ADR1000 chipie and turned off all lights in lab. And after some pondering at camera settings (ended up ISO1600 30", so kinda grainy) we got stray photos escaping zener effect region. ADR1000 is very poor LED in that regard. And yes, when I triggered battery operated photoflash (to get countour exposed) I can see a big jump in output 6.611850V voltage on K2002. So it's photodetector now too.
The following users thanked this post: quarks, ManateeMafia, Andreas, e61_phil, picburner, chuckb, RoGeorge, martinr33, denimdragon, MiDi, arcnet
Offline martinr33
« Reply #32 on: September 17, 2021, 05:41:35 am »
ADR1000AHZ is at MOQ of 5 in the US. I may try and call them tomorrow.
Offline chuckb
« Reply #33 on: September 17, 2021, 12:22:27 pm »
TiN
Was the glowing chip at 4 or 8ma?
Offline Noopy
    • Richis-Lab
« Reply #34 on: October 02, 2021, 05:07:25 am »
Of course I had to take some pictures of the ADR1000 too.  ;D
Special thanks to harerod (https://www.harerod.de/) for donating the part.  :-+
In the ADR1000 pin 4 is connected to the case. In the LTZ1000 (https://www.richis-lab.de/REF03.htm) the case wasn´t connected to pin 4 directly. There was just a connection through the substrate.
The die attache is done with some glue like material.
LTZ1000 was soldered which gave you a temperature resistance of 80K/W.
LTZ1000A was glued with some special fluffy polymer you can see in the LT1088 too (https://www.richis-lab.de/LT1088.htm). The temperature resistance was risen to 400K/W.
For the ADR1000 216K/W had to be good enough.
A very clean edge! Perhaps it´s normal. Perhaps they expended effort to get cleaner edges since impurities due to sawing of the wafer can worsen the stability?
Analog applied two layers of metal.
ADR1000 is quite similar to the LTZ1000 but the heater structure is much simpler, no more spare structures.
Let´s take a closer look at the circuit. There are two additional bondpads leading to the reference circuit.
With the two metal layers the circuit is a little harder to read...
The active parts are the same as in the LTZ1000 but here we have the possibility of a 4-wire-connection! VREF and GND are connected a second time directly at the zener-transistor-combo so the bias current and the traces conducting the current don´t interfere with the reference voltage. The sense traces are connected to the two unused bondpads.
Trying to clarify the connections.
And of course I had to light it up (5mA).  ;D
It looks like the light is more uniform than in the LTZ1000. Perhaps here we see one reason for the lower noise figure? A more uniform structure in the buried zener?
https://www.richis-lab.de/REF19.htm
 :-/O
  • Sounds like noise
« Reply #35 on: October 02, 2021, 07:03:30 am »
Thanks for the pictures. I'm somewhat of two minds, on one side I like to see some of the insides, on the other hand all that parts are lost forever :(
That's a mistake, LT used ICA aka conductive adhesive in the LTZ1000, the die isn't soldered. In the LTZ1000A the adhesive is filled with glass beads, a material that hasn't many suppliers and was thought of for ADR1000 too, as said by Eric Modica at MM2021.
Looks like a normal wafer saw cut with diamond blade.
Ah, you missed that part of explaination on that by Eric Modica.
I already spotted that force-sense connections on the pictures provided by TiN. Unfortunately the package is missing two more pins to make use of them. Maybe at some point we will see ADR1000 in a different 10-pin package too?
-branadic-
The following users thanked this post: MiDi, Noopy
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #36 on: October 02, 2021, 08:00:25 am »
At least they could have bonded the second connection to the existing pins. The weak points are the two wire bonds on each wire and redundancy would already help. I'd consider that half ready.
Regards, Dieter
Offline Kleinstein
« Reply #37 on: October 02, 2021, 08:26:36 am »
I would not consider the lack of kelvin sensing a real problem. The temperature is quite constant and thus a constant resistance for the trace and bind wire.
A slightly wider trace my be possible if the extra one for kelvin sensing is made thinner or left out. Even the pin temperature is still quite stable. The current is also constant and thus only a rather constant part of maybe 1 mV or so.
Offline Noopy
    • Richis-Lab
« Reply #38 on: October 02, 2021, 08:42:15 am »
You are right but after all the ADR1000 is still in production.
Thanks for the info!
That means that in the LT1088 they used something different:
https://www.analog.com/media/en/technical-documentation/application-notes/an22.pdf
"air impregnated polymer die attach"
That´s right. Can you enlight me?
Offline iMo
  • Country: va
« Reply #39 on: October 02, 2021, 08:57:11 am »
Ask ADI for chips/dies only. I bet you get them for a small fraction of the ADR1000 price.
And you may experiment - I think bonding 10 wires cannot be a big problem in Germany..
Offline Kleinstein
« Reply #40 on: October 02, 2021, 09:05:28 am »
Just massive glass beads would not help much with thermal conductivity. It would need some hollow glass beads to really make a difference.
Massive beads would only help to get a thicker layer, which may be enough.
For a low power use, there would be the option to use the power of the external heater transistor too. Just spread it out evenly to make good use of it instead of putting the transistor in the most distant corner of the circuit.
Offline harerod
  • ee - digital & analog
    • My services:
« Reply #41 on: October 02, 2021, 11:25:58 am »
I had a similar emotion, when I randomly picked that poor little lamb out of its flock.
Noopy contacted me to buy an ADR1000, with the express intention to provide die pictures for the EE community. The reason why I decided to support his work, is because noopy has been sharing the excellent results of his considerable efforts with the community for years.
I also hope that the high quality of his pictures will keep serving as a basis for technical discussions and keep many others from cracking components open, since they would have a hard time generating better results.
The following users thanked this post: chickenHeadKnob, TiN, Echo88, Andreas, BU508A, RoGeorge, serg-el, wolfy007, MegaVolt, mawyatt
Offline Dr. Frank
« Reply #42 on: October 02, 2021, 12:53:49 pm »
 :-/O
Hello Noopy,
thank you for the die pictures and your analysis.
In the session, I posed the question to Eric Modia, how they achieved these 50% noise level compared to the LTZ.
His answers were
1. "Pure luck"
2. "Due to the lower reference voltage, the zener structure of the ADR1000 operates in the Zener effect mode, compared to the LTZ which operates in the more noisy Avalanche mode"
That sounded reasonable at first, but I can't judge his statement.
Anyhow, after my next question Eric seemed to be surprised, that the ADR1000 can be (and should be) operated with an un-heated TC = 0ppm/K by choosing appropriate zener and collector currents.
His advise though, to use 8mA for the zener, is definitely a wild guess and not necessary at all. See my measurements in this thread.
Frank
« Last Edit: October 02, 2021, 02:18:23 pm by Dr. Frank »
Offline magic
« Reply #43 on: October 02, 2021, 01:51:22 pm »
Comparing the images side by side, I see that dimensions and layout are almost the same as LTZ1000.
Several people who attended the MM mentioned that the chip was under development for many years. Is this a rebranded "LTZ1001" from Linear or a clone that AD worked on before the merger?
  • Sounds like noise
« Reply #44 on: October 02, 2021, 02:43:39 pm »
Eric Modica (ADI) mentioned, that ADI wanted to have their own LTZ1000-like reference for products such as AD5791, that's why they started the development.
-branadic-
The following users thanked this post: magic
Offline ScoobyDoo
« Reply #45 on: October 02, 2021, 05:28:39 pm »
Hello all Voltnuts,
Before there was any internal noise of the LTC M&A - the ADI VP responsible for Voltage reference devices spotted the AD5791 board - he was not thrilled - and wanted to have his own premier Voltage Reference circuit on the board.
The legacy LTZ1000(A) was designed by Carl Nelson and Mark Glanville  (mask designer) hence the initials CN MG on lithography/die.
The ADR1000 was designed by Eric Modica, Paul Henneuse (retired), and Mark Reisiger and had many R&D pitstops - this design initiated 4-5 years before the M&A with LTC.
The VP of ADI would never have launched this project if he had been aware about the M&A talks with LTC.
After the acquisition of LTC the ADI team had access to Carl Nelson (LTC) and heavily leaned on his wide shoulders to bring it to a successful ending ...
Best regards
ScoobyDoo
https://readingjimwilliams.blogspot.com/2011/08/app-note-22.html
« Last Edit: October 03, 2021, 04:22:30 am by ScoobyDoo »
The following users thanked this post: 2N3055, ramon, ch_scr, Noopy
Offline Noopy
    • Richis-Lab
« Reply #46 on: October 02, 2021, 07:45:59 pm »
Would you please share your knowledge?
Worth a try.  :-+
Bonding such small wires is no easy task. If you look too long at the pictures you often think a little probing would be no problem but without special equipment that is no fun.
His answers were
1. "Pure luck"
Answer 2 seems wrong. In the zener mode the tempco would move into the negative area, wouldn´t it?
Offline iMo
  • Country: va
« Reply #47 on: October 02, 2021, 07:59:34 pm »
@Noopy:
1. in the presentation EricM explained why they went with a single duo heating ring..
2. bonding - not easy but in DE there is a lot of manufacturers/r&d_centers where they can bond a chip into a package. They usually use ceramic packages with N-pins for prototyping/measurements, afaik.
« Last Edit: October 02, 2021, 08:30:16 pm by imo »
Offline Noopy
    • Richis-Lab
« Reply #48 on: October 02, 2021, 08:08:35 pm »
@Noopy:
1. in the presentation EricM explained why they went with single heating's ring..
2. bonding - not easy but in DE there is a lot of manufacturers/r&d_centers where they can bond a chip into a package. They usually use ceramic packages with N-pins for prototyping, afaik.
Can I download this presentation somewhere?
Offline Dr. Frank
« Reply #49 on: October 02, 2021, 08:54:54 pm »
His answers were
1. "Pure luck"
Hi Noopy,
please carefully read my question. Here I reproduced his answer concerning the noise, but not about the T.C.!
Or what's your point?
The T.C. of the zener diode is indeed depending on its absolute value, or in other words, if it's in either zener or Avalanche mode, I guess.
The trick with LTFLU, SZA263 and ADR1000 (unintentionally?) is, to inversely match its T.C. with the -2mV/K of the transistor BE diode.
Therefore you have a limited choice of zener voltages which will do the job.
Frank
« Last Edit: October 02, 2021, 08:58:50 pm by Dr. Frank »
  • Sounds like noise
« Reply #50 on: October 02, 2021, 09:14:15 pm »
If I got Eric right he said:
"...First I believe Jim Williams used a similar arrangement to make the LT1088 rms to dc converter, but also when you start to look at the stability of any thermal feedback system the transit delay from the heater to the sensor plays a big role in determining the effective loop phase margin of that loop. So in theory a trade-off exists between the thermal gradient and the thermal stability, that is the closer the rings are to the core the less phase delay is introduced in the feedback, but that would cause a larger gradient between Q2, the zener and Q1. So the fact that this was a kit part makes me think that both, Jim and Carl were using this for the wrong purpose, Jim was using it to create an rms to dc converter, so he just wanted the thing to run fast, so he wasn't using the zener, he was using the temp sensor and the heater. However Carl had to be very concerned with this phase delay effect and also gradients. If you move the heater infinitely far from the sensor what you get is basically a bang bang controller. The heater heats up, it takes to long for the sensor to pick it up and consequently the control loop bangs back and forth inbetween the rails..."
What I hear out of that "kit part" is, both Jim and Carl basically used the same set of masks to create their individual solutions, Jim his thermal converter and Carl his heated zener reference. So that could explain why there are these additional heater rings, we again find in the LT1088.
It's always a mixture of Zener and Avalanche effect (breakdown voltage <4V is kind of a pure zener effect, while with breakdown voltage > 6V the Avalanche is dominant). So decreasing the breakdown voltage by accident, as Eric said by "pure luck", its a little more of a Zener than Avalanche effect.
-branadic-
« Last Edit: October 02, 2021, 10:27:45 pm by branadic »
The following users thanked this post: ch_scr, Noopy, harerod
Offline wutieru
  • Country: cn
« Reply #51 on: October 03, 2021, 01:01:04 am »
Test board:
The following users thanked this post: trophosphere, bsw_m, ramon, ch_scr, syau, bob.cmhk
Offline wutieru
  • Country: cn
« Reply #52 on: October 03, 2021, 01:03:10 am »
Test whole night and datas:
The following users thanked this post: ch_scr, syau
Offline wutieru
  • Country: cn
« Reply #53 on: October 03, 2021, 01:17:01 am »
1000 sample data
The following users thanked this post: syau, miro123
Offline Noopy
    • Richis-Lab
« Reply #54 on: October 03, 2021, 04:01:50 am »
His answers were
1. "Pure luck"
Hi Noopy,
You asked him about noise and for that question his answer is ok. Avalanche breakdown is a rude process.
I was wondering if that wouldn´t destroy the T.C. compensation. The transistor gives you the negative T.C. and the zener has to supply you with a positive T.C. otherwise you won´t get near zero T.C.
Moving the zener diode from avalanche to zener breakdown moves its T.C. from positive to negative. Because of that I was wondering if that is even possible. But probably branadic´s point is important: In this range it´s a mixture of avalanche and zener and probably the lower voltage still gives positive T.C.
Of course in the LTZ1000 and the ADR1000 the T.C. compensation of zener and transistor isn´t as important as in other references but in my view it´s not irrelevant.
OK, it´s just the LT1088 thing.
But that´s an interesting background story.  :-+
Thanks Dr. Frank!
Thanks branadic!
Offline miro123
« Reply #55 on: October 03, 2021, 08:12:56 am »
1000 sample data
Thanks for sharing. Do you have some pictures of full assembly?
Offline Kleinstein
« Reply #56 on: October 03, 2021, 09:15:20 am »
Reading  an external reference directly with a DMM will give the combined noise / drift of the DUT and the meter internal reference. With the ADR1000 based reference chances are most of the noise and a significant part of the drift would be from the meter.
  • Sounds like noise
« Reply #57 on: October 03, 2021, 09:44:45 am »
I disagree to some extend. I'm observing my ADR1000 for 2 months now by directly reading it with a R6581D, differential measurement with a K2182A against my F7000-1 and differential measurement of the F7000-1 against F7000-2 with a second K2182A. Consequently, I take readings for an hour each morning. You can clearly see that the DMM can detect the very same amount of drift that is also visible in the differential measurement.
The ADR reference is not pre-aged, neither by an 168 h bake&burn @150 °C that was mentioned by Eric Modica nor a cycling receipe as proposed by Cern for the LTZ.
-branadic-
« Last Edit: October 04, 2021, 09:09:50 am by branadic »
The following users thanked this post: Andreas, ch_scr
Offline Andreas
« Reply #58 on: October 03, 2021, 10:04:38 am »
Hello,
- which (model of) instrument did you use?
- do you have any clue where the negative -3 ppm spikes are from?
with best regards
Andreas
Offline syau
  • Country: hk
« Reply #59 on: October 03, 2021, 10:43:47 am »
Test board:
Did you leave any gap between the ADR1000A and the PCB ? Also, what are the value of those film resistors (1K/13K, 120, 70K x 2) ?
I am think of  making up a test board based on Dr. Frank’s PCB using ww resistors.
Offline wutieru
  • Country: cn
« Reply #60 on: October 03, 2021, 03:01:06 pm »
I use HP 3458A for testing.
and the -3 ppm spike maybe come from Vref Pin current ,I made a mistake when wiring.
The following users thanked this post: Andreas, ch_scr, syau
Offline wutieru
  • Country: cn
« Reply #61 on: October 03, 2021, 03:02:20 pm »
I use 1K:12.5K / 70K /70K/ 100R for testing.
Offline iMo
  • Country: va
« Reply #62 on: October 03, 2021, 06:29:22 pm »
OT: @wutieru: I downloaded the MeterKnife Lite (as in your picture above) - is there an english version available too?
Offline wutieru
  • Country: cn
« Reply #63 on: October 03, 2021, 10:52:03 pm »
not only the software but also a hardware named "metercare" you needed. this is a DIY GPIB device by funs.
The following users thanked this post: iMo
Offline syau
  • Country: hk
« Reply #64 on: October 04, 2021, 03:41:17 am »
May I know if you leave any gap between the PCB and the ADR1000A ? If so, what is the distance between the PCB and the ADR1000A ?
Offline syau
  • Country: hk
« Reply #65 on: October 05, 2021, 04:28:32 am »
My version:
ADR1000AHZ
1290079-0
LTZ1000ACH
1290085-1
Those WW resistors were result from a group buy in this forum. Will burn in for few days before taking measurement.
Offline Noopy
    • Richis-Lab
« Reply #66 on: October 05, 2021, 01:54:56 pm »
I found a ADR1000 with datecode 2108.  ;D
Wanna see what´s inside of this batch?  :-/O :popcorn:
The following users thanked this post: Echo88, doktor pyta, chuckb, MiDi, ch_scr, MegaVolt, syau
Offline wutieru
  • Country: cn
« Reply #67 on: October 06, 2021, 05:39:45 am »
Quick start to stable state,only need 30 seconds.
The following users thanked this post: syau, harerod
Offline Andreas
« Reply #68 on: October 06, 2021, 09:23:58 am »
Hello,
you should zoom more in (into the ppm level)
The heater voltage on my scope still changes after 7-8 minutes.
with best regards
Andreas
The following users thanked this post: e61_phil
Offline Kleinstein
« Reply #69 on: October 06, 2021, 10:04:29 am »
The heater power can change slowly even if the temperature is stable. It is normal that this takes longer to stabilize as the whole PCB heats up a little and not just the refrence chip.  The point to look for the actual temperature would be the voltage over R1 (the resistor setting the Zener current).
The chip is small and the temperatur regultion can thus be quite fast.
The longer part is settling in the reference, e.g. thermal stress between the die an the case. This what can cause hysteresis and slow settling over weeks and more. Epoxy glue and similar polymeres can also show internal relaxation in the structure. Above or near the glass temperature the internal structure reaches a kind of equilibrium (temperature dependent) fast (e.g. a few seconds at the glass temperature). At lower temperature this gets slower (like double the time every 3 to 10 K). So with more than some 50-100 K below the glass temperature the structure tends to lag behind the equilibrium.
Offline Noopy
    • Richis-Lab
« Reply #70 on: October 07, 2021, 08:31:28 am »
I stripped the layers of the ADR1000.
Here you can see how I proceeded: https://www.eevblog.com/forum/projects/decapping-and-chip-documentation-howto/msg3733501/#msg3733501
The Linear Technology Application Note 83 shows how a buried zener can look like.
Breakdown occurs between the n+ emitter doping and the p+ isolation sinker.
In my view the picture is not ideal. Under the n+ doping there is no p- but a p+ due to the addition of the isolation sinker doping and the base doping. Furthermore the doping decreases to the edges of the sinker. Only because of that the breakdown occurs in the middle of the bottom of the n+ emitter doping.
In my view the ADR1000 is built like this:
The blue temperature sensing transistors are easy to analyse. There is a buried n+ collector connector. In the outer area there is a sinker connecting the buried collector connector. On top of the collector there is a base area in which you can spot the emitter area.  :-+
The collector of the green inner transistor is connected by four long buried n+ collector connectors. The emitter area is the circle in the middle of the die. The base is a wide p doped circle between the collector and the emitter.
The base circle is connected to the anode of the zener too. Different to the AN the anode is smaller than the cathode. We can be sure about that because the wider ring is connected to the metal layer that carries the cathode potential.
Cutting through the die I assume the structures look like this.
 :-/O
« Last Edit: October 07, 2021, 08:33:29 am by Noopy »
The following users thanked this post: quarks, chickenHeadKnob, doktor pyta, RoGeorge, capt bullshot, ramon, wolfy007, ch_scr, MegaVolt
Offline magic
« Reply #71 on: October 07, 2021, 09:14:30 am »
Mostly the same kind of structure I came up with.
https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg3078768/#msg3078768
One difference: if the connection from pin 4 to the anode/base element were made at the surface, there would be no need to join this element with the P isolation wall and the substrate. I speculate that there is a gap between the anode/base and the P isolation wall, and that the deep anode diffusion is not isolated from the substrate except for those four buried Q1 collector links and that's where the connection to pin 4 lies.
I agree, that drawing seems inaccurate.
Offline Noopy
    • Richis-Lab
« Reply #72 on: October 07, 2021, 10:05:54 am »
I agree with you that this configuration would be possible and would provide us with a Kelvin connection but I don´t see a gap...  :-//
« Last Edit: October 07, 2021, 11:27:30 am by Noopy »
Offline Noopy
    • Richis-Lab
« Reply #73 on: October 07, 2021, 12:00:50 pm »
With the input of magic and a second look at the pictures I would like to change the picture a litte.
« Last Edit: October 07, 2021, 02:51:56 pm by Noopy »
The following users thanked this post: quarks, chuckb
Offline Noopy
    • Richis-Lab
« Reply #74 on: October 07, 2021, 02:02:16 pm »
i am guessing i got the pin numbers right?
pads A and B, any guesses what are the uses? a special ADR with 10 pins?
there is a square with 25 dots, is that some kind of via?
 ;D
« Last Edit: October 07, 2021, 02:51:05 pm by Noopy »
Offline chuckb
« Reply #75 on: October 07, 2021, 02:10:45 pm »
Special thanks to Noopy for braving the dangers of Hydroflouric Acid to show us more details of the chip.   :-+
Offline Noopy
    • Richis-Lab
« Reply #76 on: October 07, 2021, 04:28:42 pm »
With some background information I know can tell you that this picture shows the right arrangement (in principle of course the length and width are not true to scale).  8)
Offline magic
« Reply #77 on: October 07, 2021, 05:16:01 pm »
Hmm, what sort of secret information is that? ;D
That's a fair argument. I can't really find such gap on your photos now.
On pin 2? Not sure why there are dots but it's definitely connection between the upper (south) and lower (north) layer of metal interconnects. The two layers look slightly different and there are several other connections, but without dots.
Offline iMo
  • Country: va
« Reply #78 on: October 07, 2021, 05:28:13 pm »
EricM from ADI presented (MM2021) the cross sections of the heater rings and the reference breakdown region with details (the dopants profiles)..
PS: there are at least XXX people who saw the presentations.. It could be all of them signed an NDA  :D and therefore everything is kept secret :)
PPS: I was told there are only 2 pages - n.13 and n.15 related to the ADR1001 (out of 17) of Eric's presentation marked as "ADI Confidential". His presentation is not marked as "Confidential".
« Last Edit: October 07, 2021, 05:52:41 pm by imo »
Offline Noopy
    • Richis-Lab
« Reply #79 on: October 07, 2021, 08:08:14 pm »
That is the source of my knowledge.  8)
I don´t want to make someone sad or angry. I won´t upload the presentation or the cross section but I´m happy I got quite close in the first place and with a small correction my picture shows the right structures.  8)
You are right in the heater supply there are a lot of small vias while the other vias are big squares. Interesting... Perhaps the smaller vias are better for higher currents. Perhaps there is just no reason for the difference. I can´t tell. :-//
Offline CDN_Torsten
  • Professional electron whisperer
« Reply #80 on: October 07, 2021, 08:26:56 pm »
If memory serves me (sorry haven't routed an IC in decades)...I believe the current carrying capability of a via (in an IC) is a function of the perimeter.    So the multiple small vias would have more perimeter than one big via while occupying a similar area.
Offline chuckb
« Reply #81 on: October 08, 2021, 05:32:28 am »
What has eight legs and an eerie green glow?
My friend Jim has a home lab filled with high end microscopes. Yesterday he discovered why the ADR1000 has such fantastic performance. It is actually a Radioactive Spider at heart. A little Black Light (UV Light) shows its true identity.  It’s a good thing AD keeps them locked up in the metal can.  :)
Nikon Z6, amazing low light performance
Nikon SMZ-U Trinocular, great depth of field.
The following users thanked this post: TiN, ch_scr, syau, Noopy
Offline Noopy
    • Richis-Lab
« Reply #82 on: October 08, 2021, 05:40:27 am »
Looks dangerous!  :scared: ;D
Offline ramon
  • Country: tw
« Reply #83 on: October 08, 2021, 06:56:39 am »
Let guess, Indium?
Not long ago I read some document about bonding stress free die attachment. Sorry, I have lost the source.
Offline antintedo
« Reply #84 on: October 08, 2021, 09:24:06 am »
Might be this one: https://zymet.com/new/TC601_1_070522.pdf
Offline wutieru
  • Country: cn
« Reply #85 on: October 10, 2021, 01:02:06 am »
Does anyone spot this problem on the picture?
First, I thought it was interference caused by the power supply or wiring problems.
I changed another PCB that I used before , but still exists.
Offline rigrunner
« Reply #86 on: October 10, 2021, 02:19:41 am »
If I saw that here I would also be thinking wiring, supply, local disturbance etc.
I have only got one ADR powered up currently, but I haven't observed any jump or spike larger than 5uV so far.
Internet of Things: A solution desperately trying to find its problem
Offline Kleinstein
« Reply #87 on: October 10, 2021, 05:15:44 am »
Mobile phones can sometimes produce quite some high power interference for some sync calls, that happen not very often (e.g. 10 x a days). The output side of the LTZ/ADR1000 circuit is a bit sensitive to loading. Depending on the circuit details it may be more or less sensitive to EMI.
Offline iMo
  • Country: va
« Reply #88 on: October 10, 2021, 08:45:32 am »
The capacitor at the adr1000 output (100nF in the datasheet) should be a ceramic (TH), imho. The foil capacitors do not work well at higher EMI frequencies..
Offline Andreas
« Reply #89 on: October 10, 2021, 09:16:52 am »
Hello,
can be many root causes.
One is popcorn noise (either your instrument or the reference).
The other is EMI influence from outside (USB-cables, refrigerator, mobile phone ...)
I have the impression that the ADR1000 is more sensitive to EMI as the LTZ1000.
But I have at the moment no comparable cirquit where I can prove the impression.
I had (during 1/f noise measurement) 2-4 uV negative spikes all 15.4 seconds
which were related to a unused FTDI RS232 cable which was laying (unintentionally)
on the same desk where I did the 1/f noise measurement.
with best regards
Andreas
Offline Noopy
    • Richis-Lab
« Reply #90 on: October 10, 2021, 09:26:25 am »
In the ADR pin 4 is connected directly to the case.
In the LTZ pin 4 is just connected to the die and there it connects to the substrate.
Does the special polymer die attach electrically connect the substrate and the package?
Either way EMI coupling to the ADR package adds directly to your reference voltage. LTZ looks a little more robust.
Offline iMo
  • Country: va
« Reply #91 on: October 10, 2021, 09:27:58 am »
I think operating a mobile phone close to the vref is a good check.. For such frequencies even 100nF ceramics do not work well. You would need something like quality 1nF ceramic smd. Mind the ceramic smd capacitors generate voltage under mechanical stress..
Offline iMo
  • Country: va
« Reply #92 on: October 10, 2021, 09:48:05 am »
I can see a bond wire from the pin 4 to the case in the LTZ..
Offline Noopy
    • Richis-Lab
« Reply #93 on: October 10, 2021, 09:50:22 am »
Really?  :-//
Offline iMo
  • Country: va
« Reply #94 on: October 10, 2021, 09:51:56 am »
On 1 o'clock..
Offline Noopy
    • Richis-Lab
« Reply #95 on: October 10, 2021, 09:53:33 am »
On 1 o'clock..
Sorry, I don't get it.  :-//
Offline iMo
  • Country: va
« Reply #96 on: October 10, 2021, 09:55:02 am »
Offline Noopy
    • Richis-Lab
« Reply #97 on: October 10, 2021, 09:57:11 am »
 ;D
That's just a part of the package that originated due to the grinding.
Pin 4 is at 4 o'clock.
Offline iMo
  • Country: va
« Reply #98 on: October 10, 2021, 10:04:26 am »
 ;D Now, what would the experts say on the "grounding vs. not grounding the ref case" ? Would be interesting to know..
Offline Noopy
    • Richis-Lab
« Reply #99 on: October 10, 2021, 10:13:19 am »
That's an interesting difference between the LTZ and the ADR.
In an ideal environment that doesn't matter but in real life...  :-/O
Offline iMo
  • Country: va
« Reply #100 on: October 10, 2021, 10:35:22 am »
In the ADR the 100ohm resistor (wired to its case) is somehow blocked against EMI via the opamp's feedback capacitor (100nF in the DS). That would not work for vhf/uhf frequencies, imho. I would try with an 100pF-1nF ceramic smd parallel to the 100ohm resistor.
« Last Edit: October 10, 2021, 10:39:40 am by imo »
Offline wutieru
  • Country: cn
« Reply #101 on: October 10, 2021, 10:47:29 am »
I think it is caused by insufficient aging.
Offline Kleinstein
« Reply #102 on: October 10, 2021, 10:58:36 am »
With the 100 nF capacitor (C5 in ADR1000 DS) directly at the OPs FB, adding capacitance parallel to the 100 Ohms would act like capacitive load to the output. Depending on the OP used this may be bad for the stability.  AFAIR Andreas added more to the circuit (AFAIR an additional resistor between pin4 and the OP) in his EMI enhanced version.
With such changes it is a good idea to check the stability / tendency for oscillation in a simulation. With multiple feedback paths and some gain from the transitor the stability is not so easy.
For a practical compact circuit is may be interesting if C3,C4,C5 could be made somewhat smaller (e.g. 33 nF) to make C0G capacitors more feasable.
I would not expect aging to cause sudden jumps. Aging is mainly about the glue / die attach and this is more like smooth, slow drift.
Offline iMo
  • Country: va
« Reply #103 on: October 10, 2021, 11:03:36 am »
For, say 800MHz - 5GHz (wireless), even a 10pF ceramic smd would help..
Offline wutieru
  • Country: cn
« Reply #104 on: October 10, 2021, 11:29:33 am »
see picture as below:
Offline MiDi
  • Country: ua
« Reply #105 on: October 10, 2021, 11:54:24 am »
Indeed, especially if you are making a call.
You talk about the piezoelectric effect of ceramic capacitors?
It is unrelated to smd, different ceramics show differently pronounced piezoelectric effect, for e.g. C0G type it is negligible.
Offline iMo
  • Country: va
« Reply #106 on: October 10, 2021, 12:04:04 pm »
There is a relation to smd, indeed. Typically the stress on an smd part is higher than with a TH part.
PS: example: the knocking at the bnc connector of an o'scope is a standard test Dave is showing in his teardown videos..
« Last Edit: October 10, 2021, 12:22:18 pm by imo »
Offline Kleinstein
« Reply #107 on: October 10, 2021, 01:38:25 pm »
Most THT parts have some extra decoupling from the board stress. However it is not only capacitors that react to stress. Also resistors react to stress and change there value. With resistor arrays in a plastic case it may be easier to get a change in ratio from stress than from temperature.
Offline 3roomlab
« Reply #108 on: October 10, 2021, 02:47:24 pm »
so this zener seem to only perform best after high temp age "treatment" ? anybody else doing high temp age vs not-high temp aged comparison?
could you talk a little about your 168hr "treatment". like the heat cycling you did?
« Last Edit: October 10, 2021, 02:50:12 pm by 3roomlab »
  • Sounds like noise
« Reply #109 on: October 10, 2021, 03:41:58 pm »
Eric Modica mentioned in his talk at MM2021 - that's where this graph is coming from that wutieru linked in the post above - that them (ADI) but also their customers prefer a bake&burn for 168 h at 150 °C and observed that it removes low frequency noise. No cycling, no secret recipe, just bake&burn.
-branadic-
« Last Edit: December 22, 2021, 08:20:14 pm by branadic »
Offline harerod
    • My services:
« Reply #110 on: October 10, 2021, 03:58:07 pm »
Eric Modica mentioned in his talk at MM2021 - that's where this graph is coming from that wutieru linked in the post above - that them (ADI) but also their customers prefer a bake&burn for 168 h at 150 °C and observed that it removes low frequency noise. No cycling, no secret receipe, just bake&burn.
-branadic-
branadic, to make that statement crystal clear: are we talking about baking the powered-down ADR1000 in an oven for one week @ 150°C ?
Offline Kleinstein
« Reply #111 on: October 10, 2021, 04:47:06 pm »
The graph with the spikes on the low curve for the non baked parts is suspicious. The extra spikes and excursions all happen at the same time for the 5 DUTs. Chances are this is more an effect of the measurement system or EMI and has nothing to do with the treatment. Hard to tell if there is some effect on the noise. That part is hidden under interference.
The drift changes quite a bit. For my feeling it looks like the 168 h at 150 C may be a bit too much / too high, as the dirft changed the direction.
Some high temperature bake is definetly a good idea to let the chemical reaction of a 2 K glue really finish. It help to be above the glass temperature for this.
If the die attach glue stays glassy / amorphous  the rate of cool down from some 140 C to 80 C may have some effect. Not sure if fast or slow cool would be better, both have there merits. A fast (e.g. less than 10 min) cool cold help with stess relieve as the stress can relax before the structure. Slow cool could result in a denser structure and thus slower creep afterwards.
  • Sounds like noise
« Reply #112 on: October 10, 2021, 05:35:30 pm »
Here is a small snippet of the Q&A session:
Question: Is a burn-in possible with the integrated heater?
That's a very good question. I don't know. It is one of the reasons why we build in the capability to raise the chip temperature all the way up to 120 °C, it's not quite as hot as we would run our typical burn-in, but we will certainly try it and if it works we will publish something in the datasheet. Because yes, we did think about that.
Question: At which temperature did you make the burn-in?
I believe it's a 150 °C.
Question: So simply powering the zener and have the ambient at 150°C and then - what was it? - 168 h or something like that?
Yeah, you know and we haven't tested the bounce of that, we basically just copied the bake and burn of the LTZ1000, which is a 168h both bake and burn. We have at least small sample size evidence that you maybe able to get away with just doing an unpowered bake and that's good enough to disipate the charge but I have to say I haven't looked at it very closely.
Question: Simply elevated temperature or cycling during burn-in? Because, some rumor says cycling around glass transition temperature might improve the aging.
Our burn-in process doesn't utilize any cycling.
So you can see, it's a bake&burn process with the circuit powered, otherwise it would be a bake only. Hope that answers your question.
-branadic-
The following users thanked this post: 3roomlab, harerod
Offline Andreas
« Reply #113 on: October 10, 2021, 06:06:16 pm »
Hello,
if you measure the package against 0V you will read about 0.5-0.6V.
So yes: the substrate is somehow connected to the case also in LTZ1000.
with best regards
Andreas
Offline Noopy
    • Richis-Lab
« Reply #114 on: October 10, 2021, 07:20:09 pm »
Hello,
with best regards
Andreas
Thanks! I have donated my (destroyed) LTZ1000 so I can´t measure the path.
Nevertheless 0,5-0,6V seems to be no hard connection so the ADR can be a little more robust against EMI than the LTZ.
Do you know whether there is a different between the LTZ1000 and the LTZ1000A? This strange polymer die attach with the glass beads doesn´t sound very low ohmic.
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #115 on: October 10, 2021, 07:46:59 pm »
Substrate connected to case only on LTZ1000CH. ACH have it isolated.
Offline Noopy
    • Richis-Lab
« Reply #116 on: October 10, 2021, 07:50:15 pm »
That would explain a better EMI behaviour.  :-+
The package of the ACH is still no shield since it isn´t connected to anything but it´s better than connecting it to the reference output.
Offline 3roomlab
« Reply #117 on: October 11, 2021, 12:11:40 am »
maybe after 168hr HT-age, it needs the hot/cold modulation cycling.
  • Sounds like noise
« Reply #118 on: October 11, 2021, 09:11:14 am »
I was thinking the very same and what we can see there seems to be some common mode effect, not low frequency noise.
Eric argued, that the bake&burn removes some surface charges. However, I think bake&burn helps for two major effects:
1. curing / hardening the die attach and remove stress introduced by it --> needs to be around/above above glass transition temperature of the die attach and cycling around glass transition temperature would help even more
2. "healing" of the silicon latice from the doping process --> normally rapid annealing with neon lamps is done after doping steps, which doesn't mean that the latice is already back in perfect condition
I can barely remember some website entry / article from an institute / university discussing such investigation for the LTZ1000, but can't seem to find it at the moment.
-branadic-
Offline iMo
  • Country: va
« Reply #119 on: October 11, 2021, 09:40:50 am »
An annealing process (ie after an ion implantation) is done at around 1000degC for about an hour.
The surface charges might be somehow removed by 150degC (ie gasses/vapors) but they will condense back on the surface afterwards.
So the baking as discussed with Eric is mostly about the epoxy glue, imho.
Offline Kleinstein
« Reply #120 on: October 11, 2021, 10:04:22 am »
A burn in at 150 C would not help much with latice defects in silicon. The temperature is just way to low. For annealing the silicon, think about 700°C and up, so nothing to be done with the ready made part.  Some of the process steps to grow a new oxide layer already works as nealing for much of the earlier damage.
If at all it may be some impurities (e.g. hydrogen) to still move inside the silicon, the silicon itself is fixed.
Surface charges can indeed by a problem, but here it would be more like having a surface that is not susceptiple to catch new one. The glow from hot electrons visible with many zeners or transistors in avalance mode is a thing to avoid, as this could also excite deap states in the oxide.
Water can bond relatively strong to the oxide surface: In vaccum it takes some 100 C to get it off in reasonable time. This water causes stress, kind of making the oxide swell.  I have seen the water in/at the surface oxide bend the silicon (though with thin silicon parts). With the sealed case the water is kind of fixed and nothing to get it out or in (not sure which case is better). The water at the surface could cause some hysteresis, like a new relaxation after oven turn on after off for some time.
I would expect the burn in to be mainly about the epoxy / glue. Here there are 2 processes:
1) stress from mismatch in thermal expansion, of the glue, die and case . At the glass temperature or slightly below the relaxation should the fast enough, that there is no more stress from processes before.
2) relaxation in the structure of the amorpous polymer. Well below the glass temperature the material wants to get a denser structure, but the relaxation gets slow and the actual structure is usually less dense than the equlibrium. The degree of order in the strcuture also effects the strength and speed of the relaxation. Initially it can be fast, but once relaxed it can slow down the process quite a bit.
Chances are a good process would be cure at a high temperature, like 150 C (I don't think this would need a full week and I don't think cycling would help), than cool down relatively fast and to a relatively low temperature (maybe even room temperature) to let much of the stress relax with a still relatively soft structure. This step may take quite some time.  Than heat up to a temperaure a little (e.g. 20-30 K) above the later oven temperature (but below Tg - 50K) for some time to get the structural relaxation done, without adding much new stress. The last step may well be with just the internal heater and the zener running.
Offline wutieru
  • Country: cn
« Reply #121 on: October 11, 2021, 12:02:51 pm »
Noise Test (using 120R),Magnification:100K
Converted to actual value: about 0.9uVpp
Offline branadicTopic starter
  • Sounds like noise
« Reply #122 on: October 11, 2021, 12:53:34 pm »
For a pure bake your statement is right, but for a powered circuit at 150 °C things are different.
-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
Offline Noopy
    • Richis-Lab
« Reply #123 on: October 11, 2021, 02:03:43 pm »
Now let´s take a look into an ADR1000 out of the new batch.
Looks the same like the 1839 ADR1000.  :-+
The following users thanked this post: quarks, Andreas, Kean, ch_scr, CDN_Torsten, syau, leighcorrigall, Jes7
Offline Andreas
« Reply #124 on: October 11, 2021, 05:48:21 pm »
Hello,
can you do a FFT on the aquired data?
there seems to be much non statistical noise on the screen:
either from AC mains frequency or too much bandwith of the scope.
(BW-limiter 20 MHz is active?)
with best regards
Andreas
Offline wutieru
  • Country: cn
Re: Lowest drift, lowest noise voltage reference (ADR1000AHZ)
« Reply #125 on: October 12, 2021, 01:04:30 pm »
Niose with FFT
Offline Andreas
« Reply #126 on: October 13, 2021, 05:31:16 am »
Helllo,
ok this is with bandwidth limiter.
So the fuzzyness of the signal is much better now.
FFT reveals that there is one frequency outside the passband of the LNA which is in amplitude similar than the passband signal.
I am not familiar with this scope. For me it looks like you did a linear frequency display with 157 Hz center frequency and 313 Hz span.
So roughly 31 Hz / div.
I guess that the (red marked) discrete frequency is the mains frequency with 50 Hz.
I usually use a metal cookies box to shield against fields from outside operating the DUT on batteries. But even with the shield I have to put all transformers away with a minimum distance of 1 m. Even my HP34401A internal transformer (shielding around transformer and the housing of the Instrument) has to be at a minimum distance of 0.5 m.
On a quiet place I have the 50 Hz hum  ~18 dB below the pass band for a ADR1000A.
See attachment with double logarithmic display.
with best regards
Andreas
Offline wutieru
  • Country: cn
« Reply #127 on: October 13, 2021, 09:13:22 am »
Yes, there is 50Hz noise Exist, it comes from the test environment, I changed another oscilloscope, it disappeared, it seems that the shielding of KS oscilloscope is not very good.
Offline Andreas
« Reply #128 on: October 17, 2021, 02:25:16 pm »
So according to that what is published the burn in cirquit would look something like this.
(the resitors of course cemented wire wound for the high temperature).
The unused heater is connected to substrate.
any comments?
with best regards
Andreas
  • Sounds like noise
« Reply #129 on: October 17, 2021, 08:31:18 pm »
I can't really comment on that as I simply don't know. However, with respect to the question at MM2021 if the internal heater could be used for a bake&burn, Erics answer was related to the ADR1001, not to the ADR1000.
So one could propably set the oven temperature to a value of 150 °C, which would translate to a divider of 17.74k : 1k with the numbers given for the datasheet circuit and use the build in heater for the bake&burn of the ADR1000?
-branadic-
Offline rigrunner
« Reply #130 on: October 17, 2021, 10:12:52 pm »
I've been experimenting with heater temperature on the ADR1000.
Running the ADR heater at 85C for 48hrs, 110C for another 48hrs and then 150C for a further 48hrs briefly changed the drift direction on my sample of 1.
I'm still playing around with heating and will post should I discover anything useful.
Offline wutieru
  • Country: cn
« Reply #131 on: October 18, 2021, 01:20:58 am »
I think the burning@150° process is irreversible.  Test it several times and see how the curve changes?
Offline wutieru
  • Country: cn
« Reply #132 on: October 18, 2021, 01:28:42 am »
Power off-on testing,first interval 3 hours,second interval 30 minutes.
It seems that the voltage has changed a little  after a long time of power off.
Offline Andreas
« Reply #133 on: October 18, 2021, 05:13:28 am »
The more interesting things would be:
- does the noise actually decrease.
- does the ageing drift (after a settling time) decrease.
Question: how did you check that the temperature was actually 150 deg C?
with best regards
Andreas
Offline Andreas
« Reply #134 on: October 18, 2021, 05:26:46 am »
Hello,
I fear this is not true with my sample.
when calculating from my measurements I get a slope of -2.24 mV/K for the sensing transistor with 70K and pin 6+8 connected. (offset 648.9 mV)
To get 150 deg C temperature I would have to set around 312 mV. which translates with a zener voltage of 6600 mV to a ratio 20.15K : 1K.
But will the Zener really still have 6600 mV at 150 deg C?
with best regards
Andreas
  • Sounds like noise
« Reply #135 on: October 18, 2021, 06:57:51 am »
... which would translate to a divider of 17.74k : 1k with the numbers given for the datasheet circuit
Using the temperature sensor as a diode requires for sure a different divider setpoint.
-branadic-
Offline Vtile
  • Ingineer
« Reply #136 on: October 18, 2021, 04:34:08 pm »
If there is any analogy to metallurgy, glass working and mineral manipulation by annealing one should not only apply heat, but also control the cooldown. Rule of thumb on those mentioned before is that slower is better.
Offline rigrunner
« Reply #137 on: October 18, 2021, 05:08:55 pm »
No noise reduction yet.
It is still drifting at about the same rate.
Measuring the diode voltage across Q2 using -2mV/C
Stepped temperature tests are something I had considered on a fresh ADR. I may try that when I have finished poking ADR#1.
« Last Edit: October 18, 2021, 05:28:27 pm by rigrunner »
Offline rigrunner
« Reply #138 on: October 18, 2021, 05:24:46 pm »
With Q2 configured as a diode I measure around 574mV @ 25.2C for a 71K5 R3. ADR#1 and a random used LTZ are in the same ball park.
« Last Edit: October 18, 2021, 05:26:52 pm by rigrunner »
Offline Andreas
« Reply #139 on: October 18, 2021, 07:09:00 pm »
The question is: how this is measured.
I need sampling with 1 NPLC (faster would be better) and measuring the very first sample (with heater off)
after power on to get 592.3 mV @ 25.3 deg C and 69.8K.
Due to self heating the voltage changes quickly during the first seconds
and finally stabilizes at around -20 mV lower level (+10 deg C).
Ok, in datasheet the pull up resistor is only 62K so we have a slightly higher collector current.
But which datasheet numbers do you refer to? For Q2 I see only the table 7 with some values.
With my cirquit this would be roughly 10% too low. So only around 135 deg C.
This is one reason why I want to use 2*25W-aluminium wirewound resistors as heater
and sensing the case/environment temperature of the ADR with a PT1000 sensor.
The whole in a termo-isolated coffee mug or something like that.
with best regards
Andreas
Offline rigrunner
« Reply #140 on: October 18, 2021, 08:12:28 pm »
I can measure Q2 voltage with the opamp from the heater side of things removed.
I'll need to play around with my logging setup. I currently trigger a sample every minute.
Best I can do at the moment for checking ADR can temperature is take a thermal image.
Offline Kleinstein
« Reply #141 on: October 18, 2021, 08:14:41 pm »
I would expect some reversible effect of the heating to 150 C.
The 150 C are likely above, or at least close to the glass temperature of the die attach glue.
So this would kind of reset the glassy structure to a relatively open (high free volume) state. How much of this is retained depends on the speed of cool down and later aging at lower temperature like 60 C during operation.
The 150 C would be enough for stress from thermal mismatch between the silicon and case to relax. So this would start again from the same level.
The curing of the glues part would be irreversible and probably well fisched after 1 week at 150 C.
Offline rigrunner
« Reply #142 on: October 19, 2021, 02:10:16 am »
3457a setup with NDIG=4, NPLC=.1, AZERO=1,TARM=AUTO,TRIG=SGL, NRDGS=1.
Readings were taken as fast as the pi could manage in a python loop with query(3457,'?')
594.98mV start and drops 13.62mV in 190 seconds to 581.36mV with self heating.
The following users thanked this post: Dr. Frank, Andreas
  • Sounds like noise
« Reply #143 on: October 19, 2021, 10:28:32 am »
I'm refering exactly to that table, from which you can extrapolate the resistor values for a 150 °C oven temperature for the datasheet circuit.
However, things are different when using Q1 as a diode only. By the way, from what I measured on my first unit I get a -2.166 mV/K for Q1 used as a diode.
Running my reference for about 1850 h by now I don't observe any real settling, though not a steady drift either.
The linear drift rate is about -360 nV/day (~13 ppm/a), allthough a square fit matches much better, while a square-root fit shows large deviations and doesn't match well, at least by now.
-branadic-
« Last Edit: October 19, 2021, 04:55:50 pm by branadic »
The following users thanked this post: Mickle T., Andreas
Offline harerod
    • My services:
« Reply #144 on: October 19, 2021, 07:02:04 pm »
I have been asking myself how a gang-burn-in of several ADR1000 might be set up. I have some Vishay RSSD resistors in the lab, which usually serve as loads during EMI tests of power supplies. Those resistors feature a cylindrical ceramic body, which can accommodate several TO-99.
Today's test was about general feasibility. I usually wouldn't touch a breadboard with a ten feet pole, but in these trying times of component shortages, I decided to go retro. It turned out that the only THT OPAMP, which I had at hand, was a veteran that has been with me for a very long time. No component on that breadboard is younger than thirty years.
Since I didn't want to risk an ADR1000 for tests, I decided to use an 1N4148 as target and a KTY81-122 as sensor for a temperature regulator.
I was a bit concerned about flammability, so I took care with the setup:
The two devices go into the RSSD 25x168 100R.
Eager to start the experiment I didn't even bother cleaning out the cookie jar.
The RSSD is suspended by some highly flame resistant steel mounts from the mill router. Some electrical insulation goes under the mounts, not shown in the picture.
The cookie jar is wrapped in cotton cloth. The cotton cloth rests on a thick towel and is covered by another towel.
The whole heap is placed on a metal step ladder next to my desk.
It turns out that in steady state the setup burns 13W in the RSSD. It is autumn here and I need central heating to keep the lab warm, so I am not bothered by the power dissipation.
Now that everybody had a good laugh - could anybody tell how critical that 150°C burn-in temperature is, please?
In which temperature band should the die be kept?
What is the lowest temperature that makes sense?
At which temperature can we expect damage to the ADR1000?
After all, according to the datasheet, 150°C is the junction temperature limit for storage and operation.
Offline Kleinstein
« Reply #145 on: October 19, 2021, 07:22:55 pm »
The main part to worry about, when going high in temperature is the die attach glue.  Epoxy may not like a temperature much above 170 C. With a little patience I would consider some 140 C and maybe even 130 C sufficient. This should still be at least close to the usual epoxy galss temperarure. It just takes longer at a lower temperaure.  If the target is an open structure and thus fast cool down, I would like to start at the glass temperature or above, but not much below.
I may make absolutely sense to later use a step at a lower temperature, maybe some 20 C above the later oven temperature.
The following users thanked this post: harerod
  • Sounds like noise
« Reply #146 on: October 19, 2021, 09:03:39 pm »
It depends on what the mechanism for the initial drift actually is.
Eric Modica argued at MM2021: "... Another important aspect that carries over to the ADR1000 is the benefit of bake and burn process. So when I have first started working with buried zeners and Vrefs in general the conventional wisdom was that bake and burn speed up the aging of the part and that one could accelerate the early life drift to finish up sooner. That’s this area here, that’s the majority of your drift on parts like this, your uncertainty. So I’ve learned enough in my years working in analog design not to make broad generalisations, but what I’ve observed is that bake and burn is helpful primarily for low frequency noise reduction in this context. We have had tier 1 manufacturers tell us the same thing when we send them ADR1000s. Prior to instituting a bake and burn most of the parts look okay from a low frequency noise perspective, but if you look at it enough you find one that is extremly noisy and actually remedies with the bake and burn. So only being a process hobbiest myself, the only explaination I can give there is that you have these dangling hydrongene bonds at the surface of the SI and sometimes that amounts to surface charge and if that ends up in the vacinity of the zener, if you are unlucky enough, then this is the kind of thing that you get. And it’s really the only thing that would move at such low temperatures where you are running a bake and burn, say like a 150 °C...."
I don't know if this is the real deal and wouldn't argue against it, cause I can't prove him neither right nor wrong. However, I believe from what I've learned over the years on packaging of integrated circuits that two mechanisms are involved. One being the heavily doping in the silcon latice and the second coming from the assembly of the die within the package using die attach. All sorts of plastics, thermoplastic and thermoset, do suffer from stress after being (injection) molded or applied and cured. A thermal treatment afterwards can to some extend remove this stress in those and there are many examples where such thermal treatment process is being applied.
I don't know if there were ever studies to rule out what is actually the reason for the initial drift of voltage references in hermetically sealed packages and if they can be seperated from another and what is just stupid hypthesis, rumor or wisdom.
Nevertheless, the 150 °C treatment for 168 h Eric mentioned as well as the cycling that Cern performs on the LTZ1000 gives evidence that at least the stress due to the die attach is a major contributor. I do have build at least one LTZ reference, that was heat-treated using the cycling methode and what I observed on it was, that it settled within about two weeks. I know other forum members, not to mention any names here, that observed similar effects. However, I couldn't get enough people together to perform a double-blind experiment on that to give it more evidence.
So the answer to that die attach contribution would be: storing the Vref above or cycling around the glass transition temperature, whatever that is for the die attach used in the ADR. Chances are, it's in the range of 110 to 120 °C as for many epoxy materials in that field, but we simply don't know and Eric couldn't answer that either.
-branadic-
« Last Edit: October 19, 2021, 09:17:40 pm by branadic »
The following users thanked this post: CDN_Torsten, harerod
Offline Andreas
« Reply #147 on: October 19, 2021, 09:44:17 pm »
Hello,
so what I have heard now 150 deg C is the oven temperature to go? (or is it more the 125 deg C of other processes?)
In a bake & burn configuration we will also have some self heating by the biased zener.
(at room temperature this is ~10-11 deg C).
As already mentioned I have 2 x 25W heater resistors which I will put into a thermally isolated coffee mug or thermo-jar. So I think I will get away with 10-30 W power (or less). A PT1000 temperature sensor just arrived today. I will use one of my ADCs with 4 channel MUX to evaluate temperature. (2 channels for 0 deg C and 150 deg C reference resistors).
The PC software will allow ramping the temperature (e.g. from 145-155 deg C with 0.1K/min slope).
with best regards
Andreas
Offline Dr. Frank
« Reply #148 on: October 19, 2021, 10:41:00 pm »
Hello colleagues,
I recommend not to go near or even above 150°C, as this is the absolute max. rating for storage and chip temperature, and 125°C is the maximum operating temperature, surely for some reason.
I would also chose around 120°C, what branadic has suggested, being a typical glass temperature for die attach.
You might use external heating and additional internal heating.
Such burn-in circuits are often designed with a passive substitution circuit around, i.e. w/o using any OpAmps or other semiconductor components, but appropriate HT resistors instead, due to these high environmental temperatures which are involved.
Frank
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #149 on: October 19, 2021, 10:58:16 pm »
Which means the reference will be soldered after burn-in. No good, better try to run it in its circuit, maybe using an external oven at 85 °C and the internal heater to get the additional 35 °C.
Regards, Dieter
  • Sounds like noise
« Reply #150 on: October 20, 2021, 08:03:58 am »
I do understand that the engineer wants a specific number for an upper temperature limit, allthough there are also some physicists among us. But noone can give you such a number.
You have to understand that the numbers given in the datasheet such as "Specified for −40 °C to +125 °C operation" or "Junction Temperature Range −65 °C to +150 °C" means continuous operation over lifetime. It's not that at 125.1 °C the part goes pop because of some curious effect kicking in, it's not as steep as that. But instead some effects are starting to accelerate, such as electromigration. Having said that, there is no reason to believe that operating the part for a week at 150 °C, either oven temperature or chip temperature, would degrade it permanently. The structures on the silicon are rather large and the burn-in time is limited to cause some serious degradation. Long story short, I wouldn't worry to much about the absolute temperature.
-branadic-
« Last Edit: October 20, 2021, 08:08:15 am by branadic »
Offline DavidKo
  • Country: cz
« Reply #151 on: October 22, 2021, 12:01:25 pm »
Will not be the lifetime at higher temperatures smaller? LEDs have the lifetime only ~500h instead of ten thousands hours when the junction is at 150°C .
Offline Kleinstein
« Reply #152 on: October 22, 2021, 12:36:06 pm »
The shorter life time limit at high temperature is due to slow diffusion or similar processes to be faster with high temperature (about double every 5 to 10 C higher). With the reference we want most the those slow processes to already settle before the refrence is used. For this reason there is an initial burn in / anneal phase at a relatively high temperature to lett those effects settle in reasonable time (e.g. 1 week). Later use of the reference is at lower temperature like 40 C, just warm enough to get good temperature control.
Offline rigrunner
« Reply #153 on: October 22, 2021, 02:22:10 pm »
More torturing of ADR#1  >:D
I decided to try something else with ADR#1. It takes about an hour to get stable with heater temperature at 150C (long legs, no baffle or insulation). With that I decided to see what happens if I power the ADR on for 60 minutes, wait 15 minutes to cool down and go again. This has given the largest shift so far. A little under 200uV upward in 31 hours.
There are a couple of data points missing or blanked out. My power controller doesn't talk to the logging part, so sometimes it screws up and powers down in the middle of a reading etc. Sorry :(
Offline Kleinstein
« Reply #154 on: October 22, 2021, 03:03:57 pm »
Is the temperature regulated at the 150 C , or is it just the maximum possible to reach with the given supply ?
With regulation I would have expected much faster stabilization of the temperature. Normally it should be more like 10 seconds, maybe a minute for a stable temperature.
It may be interesting to also get a stable lower temperature (e.g. 50 C range) and maybe wait longer there to at least get a stable temperature there and get thus a comparable good reading to see changes to the chip.
Offline rigrunner
« Reply #155 on: October 22, 2021, 03:23:51 pm »
150C is regulated. I'm not using the data sheet circuit. The heater is wired up differently. At a more sensible temperature  stability is reached around 20 ish minutes.
I'm planning on dropping the temperature in a little while to see what has changed at the lower settings.
Offline Kleinstein
« Reply #156 on: October 22, 2021, 04:39:51 pm »
Is the 20 minutes the time to reach a stable temperaure or the time to reach a stable ouput voltage ?
These 2 can be different. If still in regulation the temperature should be faster, unless most of the time at full power still before actual regulation starts.
The difference betwenn stable temperature and stable voltage can be interesting, as the extra relaxation to effect the voltage would be one of the effects (e.g. stress between case and die) to look at with the burn in process.
Offline rigrunner
« Reply #157 on: October 22, 2021, 04:59:52 pm »
Good point.
20 minutes is the time to reach stable output voltage.
When I lower the heater temperature next I will hook up another DMM for the heater opamp and log a sample to show the heater settle time.
  • Sounds like noise
« Reply #158 on: October 24, 2021, 09:48:23 am »
Here is an update of my drift measurement, almost 2000 h have past. Again, the reference wasn't pre-aged, only the z.t.c. temperature was measured, the oven adjusted to that temperature, the amplification adjusted to 10.000 00 V and the overall t.c. trimmed. The reference is sitting at the bench, supplied by a low noise lab power supply.
The reference is drifting downwards compared to F7000-1. By the eye the drift is not linear, but slightly curved, a sqr-fit matches best. Although, we are looking at 3.2 ppm of change, which is quite a lot. I'm tempted to stop it here and to perform a one week cooking, using the internal oven. On the other hand I could also build a second one. Not yet sure how to proceed.
Don't take the result too serious, this is only a single specimen and thus without any statistical significance.
-branadic-
The following users thanked this post: TiN, e61_phil, ch_scr, Okertime
Offline Kleinstein
« Reply #159 on: October 24, 2021, 10:21:35 am »
The drift still look pretty linear. There is a little curvature in the curve, but not really much.  Chances are, that just longer waiting would not change very much (excepts the continued linear drift).
Going up in tremperature for some time would be an option. Similar some time (e.g. 1-2 days) of power off may also be an option, to see if there is a hysteresis from power down.  The change to a higher temperature may need to power down anyway. So if possible, one could to the change in the configuration in a way so that one could do 1 day power off, than maybe 1 day with the current temperature (to see if there is a fast settling part) and only than cooking at a higher temperature (with the internal heater likely more 100-120 C). It could help to also get readings at the higher temperature - so keep the regulator active and just change the set point and maybe increase the external temperature for the whole board a little.
To see the change it would be good to get back to the current temperature / config anyway.
Offline Andreas
« Reply #160 on: October 24, 2021, 03:51:40 pm »
It would be interesting also to have the 6.6V output data.
So we do not know wether it is the zener or the output voltage divider.
with best regards
Andreas
  • Sounds like noise
« Reply #161 on: October 24, 2021, 04:42:24 pm »
As the same output stage was used on several other LTZ references before, without showing any drift, I'm convienced, that it's the ADR itself ;)
However, equipment is limited, so you can't track it all at the same time.
-branadic-
Offline miro123
« Reply #162 on: October 24, 2021, 05:13:13 pm »
The graph is almost linear
What about office temperature?
If I characterize reference I'd measure the reference. Re-start and mesayre for 1 week wil have you and idea if the trend is changed
Offline wutieru
  • Country: cn
« Reply #163 on: October 25, 2021, 12:29:26 am »
Noise testing,it seems that in right way.
Offline ScoobyDoo
« Reply #164 on: October 25, 2021, 07:19:52 pm »
Hello Voltnuts,
                           I received following communication from the ADI product marketing team in USA responsible for the ADR1000 release - In a nutshell they made following communication:
The ADR1000 will become available through various sales channels (Digikey - Mouser - Arrow) and an EV-board is in progress - but due to other related product releases (ADR100x) - with higher priority - this process is rolling out more slowly than anticipated - some patience is needed before this part will be available through your preferred distributor sales channel ...
The fact that Digikey and Mouser were mentioned means that for Voltnuts it will be much easier to order these parts in smaller quantitities & w/o VAT reg. hurdles
The ADI product marketing team is hoping to ‘release’ this part on the web by end of year.
Best regards
ScoobyDoo
« Last Edit: October 25, 2021, 07:55:22 pm by ScoobyDoo »
The following users thanked this post: quarks, TiN, branadic, Dr. Frank, Andreas, rigrunner, wolfy007, ch_scr, CDN_Torsten, syau, eplpwr
Offline rigrunner
« Reply #165 on: October 27, 2021, 04:52:38 am »
Thanks for making me take a closer look at the heater settle time.
The heater was making it to around the 85-90C mark pretty quickly and then taking the rest of the cycle to slowly reach 150C. My 11V on board supply was the problem. Running with a higher supply the heater settles much faster now.
« Last Edit: October 27, 2021, 05:12:28 am by rigrunner »
Offline Andreas
« Reply #166 on: November 02, 2021, 07:34:58 pm »
Hello,
after the discussion I plan to do a  up to 125 deg C burn in cirquit.
So I can use standard PCBs (non high TG) and standard (non PTFE) burn in sockets.
Some impressions of the mechanics with security switchoff at 140 deg C.
With best regards
Andreas
The following users thanked this post: quarks, harerod
Offline Andreas
« Reply #167 on: November 13, 2021, 06:24:50 pm »
Hello,
I am trying two different methods of finding the Zero TC point of the ADR reference.
method a)
via setpoint change (using a external DAC to influence the voltage divider for the setpoint) and active internal heater
in this case I can do rather fast temperature sweeps (1K per minute shown) which saves measurement time.
method b)
via external (environment) temperature change in this case the internal heater is de-activated.
In both cases the temperature sensing transistor (Pin 6 + 8 ) connected together is used as X-axis = internal sensor temperature voltage for the diagram.
The y axis is the zener voltage over temperature.
in case a) (heater on) I get 542.4 mV = 47.5 deg C as zero TC temperature
in case b) (heater off) I get 538.2 mV = 49.4 deg C as zero TC temperature
so around 4 mV temperature sensing voltage difference or 2 deg C which I cannot explain  :-//
with best regards
Andreas
« Last Edit: November 13, 2021, 06:29:08 pm by Andreas »
Offline Kleinstein
« Reply #168 on: November 13, 2021, 07:00:05 pm »
The test with the environmental temperature also effects the resistors. So the resistor TC would be included. I have not calculated how much shift ( added linear TC) is needed, but it may not be much.  Another part is possible thermal EMF, that may come in from the heater power. This may cause a little and would be part of the residual TC even if the heater is active.
There is a small contribution from the base current, but this should be only some 0.2 mV of shift.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #169 on: November 13, 2021, 08:37:49 pm »
Also reference voltage at zero TC point shifted by about 18  uV (3 ppm).
If you calculate the residual TC at the average temperature 48.5 °C, i'd guess it will be very small and much better than a LTZ 1000. I mean if you can't decide which oven method you will use.
Our LTFLU references run in small TEC ovens with the support circuit, including the 10 V amplifier. I know they run better when using a mixture of on-chip temperature measurement and oven temperature measurement, since the circuitry around the reference chip imposes its TC times small temperature differences (gradients inside oven). These effects can be separated by their response times. After a temperature step from another temperature to your zero TC point you may see a response with a fast and a slow component that finally compensate each other in steady state.
Regards, Dieter
Offline Andreas
« Reply #170 on: November 14, 2021, 08:20:18 pm »
Hello,
still waiting for some parts for the burn in PCB.
But in the mean time I did some dry-runs with the little oven within the coffee mug.
All tests with ~15V supply voltage (unstabilized).
So with the two 10 Ohms resistors in series I have around 11 W heater power.
The heater is isolated with some cotton cloth.
At 130 deg C inside the outside of the coffee mug is around hand warm.
Startup time from 25 deg C to 110 deg C is less than 25 minutes.
There is enough heater power to do ramps with 2 deg/minute between 110 and 130 deg C.
So this will be also the possible burn in cycle which takes 2 hours. 10 minutes ramp and 50 minutes steady state.
Switch-off of overtemperature switch is done at ~137 deg C. ( Data sheet value = 140 deg C +/- 5 deg C).
But in this case we have a large hysteresis down to 115 deg C until the switch closes again.
Eventually a ramp between 115 and 135 deg is possible without interference of the overtemperature switch.
with best regards
Andreas
Offline RikV
« Reply #171 on: November 21, 2021, 03:20:44 pm »
As of today, 21st of november, the ADR1000 nor ADR1001 are listed on the AD website (https://www.analog.com/en/products) (in Belgium), nor on the ADX site. Mouser.BE doesn't know the device (even not the ADR1399). So, what is all this?
Offline Andreas
« Reply #172 on: November 21, 2021, 09:31:52 pm »
Hello,
first (still unadjusted) TC-Measurement of my ADR1000#1.
I am using long legs.
And the output voltage (10V) is done by a 2:1 statistical output voltage divider (nominal ratio = 1.500000)
Zero TC of the zener is relative low with ~49.5 deg C.
Calculated setpoint temperature is ~51.8 deg C (so slightly above Zero T.C.)
Result:
10V output voltage T.C. ~190 ppb/K.
6.6 -> 10V transfer ratio ~24 ppb/K and ~6 ppm too low against nominal ratio.
(so naked zener would be 166 ppb/K).
Did someone already check the influence of the length of the legs on T.C.?
with best regards
Andreas
The following users thanked this post: Dr. Frank, rigrunner
Offline Andreas
« Reply #173 on: November 22, 2021, 08:08:13 pm »
Hello,
another test.
the idea was: can I influence the slope of the T.C. by using a slightly higher (or lower) setpoint temperature as the Zero T.C. temperature.
Above the zero T.C. temperature the slope is falling so what when setting the setpoint somewhat higer (or lower?)
I adjusted the voltage divider for T.C. setpoint from 11.5 / 1 to values in 5 degree steps.
With ~10-12 deg C self heating of the device by zener current I expected to see the device falling out of regulation when decreasing the setpoint to 46.9 deg C.
Interestingly was that this was not the case. (obviously the zero T.C. is good enough to compensate this for environment temperatures up to 40 deg C.)
I made a further test  with 41.9 deg C which clearly showed that the temperature regulation is only valid up to ~32 deg C.
Result: I need to set the zero T.C. and the final oven temperature somewhat higher (e.g. 52-55 deg C) for a environment temperature up to 40 deg C.
Total result: unfortunately the slope of 190 ppb/K cannot be compensated by setting the heater setpoint slightly around the zero T.C. point.
(although the graphical representation seems to have a sweet spot near the zero T.C. temperature, but maybe this is only stray by the measurement uncertainity).
with best regards
Andreas
Offline Kleinstein
« Reply #174 on: November 22, 2021, 09:06:53 pm »
The small residual TC for the ouput valtage can be from multiple sources.  The TC of the unheated reference and than a not so perfect temperature regulation is only one possible mechanism. As it looks this is more like a small part.
Other contributions are the resistors (though better than with the LTZ1000 as much less effect from the temperature setpoint). There is still the 100/120 Ohms resistor to set the current and the collector resistor(s).
Anther contribution is directly an effect of the temperature gradients on the chip or at the case and thus the heater power. One could check this part by adding / removing some insulation around the chip. The constant part of the gradients are not a problem, but the heater power changes about linear with the external temperature and temperature gradients about scale with the heater power. So the temperature gradients will change about linear with temperature.
Ideally one would get a stable temperature of the reference without much change in the heater power - like having a 2nd oven, even if not very stable.
The board temperature can also change the mechanical stress to the reference, though I don't think this would be a large contribution - at least we should hope it is not, as thus would also come with a humidity effect.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #175 on: November 22, 2021, 09:13:01 pm »
If you run the reference with its own heater inside an oven made to keep the divider at constant temperature, you could use the reference heater signal for TC adjustment, as proposed in the LTZ1000 datasheet. Inside the oven the heater is well protected from ambient temperature changes and you have a fixed operation point, so higher order terms will be negligible. If you bring the heater signal out of the oven, you can fine tune the adjustment without opening the oven. In your case an extremely small contribution will be necessary (maybe 100 MOhm instead of 400 KOhm).
Why not use a TEC oven? You can keep temperatures lower and near usual ambient temperatures, so any kind of power failure or the like will cause less deviation. Our LTFLU ovens restart with less than 0.1 or 0.2 ppm deviation after 24 h without TEC control. I know because one of our Arroyos used to turn off every two or three months (for no good reason). Meanwhile i solved that problem: Send output on command once per minute.
Regards, Dieter
Offline Andreas
« Reply #176 on: November 26, 2021, 08:03:19 am »
Hello,
now I have built 2 burn in PCBs for the ADR1000A.
This one is intended for noise measurements before and after the burn in.
But I fear that T2 base emitter voltage is too low (0.41V) to give any useful noise measurements.
with best regards
Andreas
Offline Kleinstein
« Reply #177 on: November 26, 2021, 08:29:38 am »
The circuit is a bit tricky, as the current is driven from the external 1.8 K resistor on top, and not controlled by the transistor inside the ADR1000. So the TC and noise would not be the same as in the normal circuit. If not carefull there may even be change to damage (at least stress) the transistor from too much base current (e.g. at higher temperarure when VBE goes down). So I don't think the board is really good for burn in.
Why not use the "normal" circuit with maybe lower grade resistors ?
Offline Andreas
« Reply #178 on: November 26, 2021, 07:20:45 pm »
Hello,
sorry I did not find a suitable Op-amp which is specced for up to 150 deg environment temperature.
And what do you mean with "lower grade" resistors. Can there be lower grade resistors than the 5% metal oxide resistors (with 235 deg C maximum temperature?)
In the mean time I measured the noise of the cirquit.
Unfortunately I have typical 7uVpp over a 100 sec time span with this cirquit.
with best regards
Andreas
The following users thanked this post: miro123
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #179 on: November 26, 2021, 08:39:08 pm »
If you have a small p-channel mosfet, it can prevent excessive base current while hot. You connect source to reference output Pin3, gate to the collector of Q2 Pin5 and drain to ground Pin7.
Regards, Dieter
The following users thanked this post: Andreas
Offline Kleinstein
« Reply #180 on: November 26, 2021, 08:55:23 pm »
For protection of the transisitor a resistor at pin 7 would be enough. The temperature / output measurement would than be not from ground, but pin 7 instead.
I did not consider that the whole circuit is planed to run so hot. In this case it would be a bit tricky to build a circuit to work with high temperature parts.
In principle one could consider a circuit with discrete transistors (e.g. 2N2222), maybe as a cascode and than a emitter-follower.
The more logical way would be the OP and most of the resistors at a lower temperature and only the reference chip itself hot. This could possibly be some 60-80 C for most of the circuit and than the ref internal heater to reach the final temperature.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #181 on: November 26, 2021, 09:04:58 pm »
The p-channel mosfet i proposed is a 150 °C capable substitute for an opamp. It brings the reference ciruit into an operational state very similar to the standard circuit and supports the intended noise measurements during burn-in.
Regards, Dieter
Offline Andreas
« Reply #182 on: November 27, 2021, 11:04:45 am »
Hello,
so your suggestion is this way - right?
This will shunt away current from the zener and base as soon as the base current increases.
I do not think that it helps for the noise but it should help against base current overload so I will implement it.
I still have some BSS84p or similar which are not ideal for the job but should work.
with best regards
Andreas
Offline miro123
« Reply #183 on: November 27, 2021, 12:51:09 pm »
Interesting graphs. The noise spectrum flatten around 100Hz.
Is this typical ADR1000 behavior or limitation of measurement setup?
Offline Andreas
« Reply #184 on: November 27, 2021, 01:55:39 pm »
Hello,
uVpp is always 1/f noise measured with a bandwidth limited amplifier (0.1-10 Hz)
Additionally I have a 1000 Hz math filter in the scope to reduce noise of the scope input.
The flattening should be around the quantisation noise of the scope (16 bit mode)
The reason why im showing above 10 Hz is just to see wether I have too much mains frequency (50 Hz) noise in the measurement.
with best regards
Andreas
« Last Edit: November 27, 2021, 02:10:19 pm by Andreas »
Offline mawyatt
« Reply #185 on: November 27, 2021, 02:28:17 pm »
Hello,
with best regards
Andreas
Nice!! What scope and setup did you use for the spectrum plots, those are impressive indeed!!
Best,
~Wyatt Labs by Mike~
Offline Andreas
« Reply #186 on: November 27, 2021, 03:43:51 pm »
Hello,
Setup is a low noise 10000 fold amplifier 0.1-10 Hz bandwidth (cirquit attached as Filt1105w.pdf) here:
https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg834013/#msg834013
together with the DUT (voltage reference) in a metal cookies box (all battery supplied).
Scope is a PicoScope 5444A. (the successor model sold today is a 5444D)
Settings:
20 MHz hardware bandwidth limiter.
10 mV/Div resulting in 1uV/Div with the low noise amplifier.
16 Bit hardware resolution.
Sample rate 20 kS/s 100 seconds (2 Million samples per scan, 19 scans as input for FFTs).
1 kHz digital filter. (far enough away from the 10 Hz hardware filter).
FFT with rectangular window 1 Million points (binning of 2 Million samples)
Averaging of 19 FFT outputs (not input samples) to reduce noise floor of the scope (not the DUT).
without 1kHz filter and without averaging the FFT looks like this:
with best regards
Andreas
The following users thanked this post: splin, mawyatt
Offline Kleinstein
« Reply #187 on: November 27, 2021, 04:39:19 pm »
The simplified circuit can be sensitive to noise in the supply to provide the external current. If the P-MOSFET gets active some (on the order of 1/200) of the noise of the MOSFET can also enter the result. The TC would likely also be quite a bit higher than the normal reference circuit and there can thus be noise from temperature fluctuations.
In the ciruit with the extra feedback (P MOSFET or 2 x NPN) one may also want some capacitance (e.g. at pin 3) to slow down the loop at 1 defined point to make sure it gets stable and is not oscillating.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #188 on: November 27, 2021, 05:58:42 pm »
Andreas, yes that's what i thought. According to datasheet the BSS84p operates up to 150 °C.
Without regulation the power supply noise gets divided by 1K/100R=10. With regulation it will be much less. And i think you get regulation already at room temperature, as (13.9 - 6.43V) / 1K gives about 7.5 mA, not 4.3 mA. So there should be base current at room temperature.
Yes, there is a chance for oscillation, as the two transistors have enough gain to make a nice oscillator. If it happens, it can be solved by removing C4. The mod poses no risk to the rare/expensive reference. But of course one can make a ADR1000 dummy from discrete parts to check a circuit for "fatal threats".
Regards, Dieter
Offline Andreas
« Reply #189 on: November 27, 2021, 06:30:41 pm »
Hello,
its not 1K for the upper resistor but 1K8. (I wanted to stay below 5 mA)
with best regards
Andreas
Offline Andreas
« Reply #190 on: December 02, 2021, 08:00:26 am »
Hello,
pre-test for burn in cirquit with a temperature sweep from 60 to 130 deg C.
Above ~80 deg C the base emitter voltage of T2 decreases by the shunted current of the FET.
So the cirquit works.
But obviously the zener current decreases from initially 4.1 mA (410mV) to around 3 mA (300 mV) at 130 deg C.
So the cirquit is not ideal but should be ok for burn in.
Another effect is due to the massive heat sink capacity of the aluminium bar where the ADR is connected to.
The self heating seems to be only around 2-3 deg C.
(but this is measured with around 8 NPLC of my ADC#25 since my DMMs are occupied by ADR#1)
with best regards
Andreas
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #191 on: December 02, 2021, 09:55:38 am »
While the p-channel mosfet works similar to the usual opamp, there is another variant with a p-channel depletion JFET shunt parallel to R1, i.e. S = U1pin4, G= U1pin5 and D=Gnd. That variant does not shunt zener current at high temperatures. The J175 datasheet says it can be used up to 150 °C.
Regards, DIeter
Offline Kleinstein
« Reply #192 on: December 02, 2021, 11:00:03 am »
In the normal operation with the closed loop the zener current is not constant, but going down. So the change in current is normal. The unusual part is more the constant current at lower temperature.
A problem with the circuit is that the noise and drift may well be more effected by the FET and not so much the reference itself. So one would not get a good real time measurement of the aging / burn in.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #193 on: December 02, 2021, 11:12:48 am »
From previous discussion it was clear that the loop amplifier in this application isn't critical. If one uses the output of the on-chip transistor, it becomes the input stage and determines noise and temperature dependent offset.
Regards, Dieter
Offline MaxTesla
Re: First Tests ADR1000AHZ
« Reply #194 on: December 06, 2021, 01:57:09 pm »
Hello,
Thanks to ScoobyDoo and branadic, I got 5 samples and some PCBs.
Here's the mockup board, where I made the first tests.
The lower zener voltage compared to the LTZ1000 now allows this Reference Amplifier to be trimmed to zero T.C. for a given oven temperature, by individual selection of the collector resistor R2. Compare this to the +50ppm/K for the unregulated LTZ1000.
That trimming process is the same procedure as for the SZA263 and LTFLU in the different FLUKE instruments.
I measured Uref vs. temperature, which always gives a negative, flat parable. The zero T.C. point varies from 45 to 60°C for a fixed value of R2, and the T.C. varies from about +5 .. -5ppm/K over 24 ... 80°C only.
At a fixed divider ratio R4:R5 for around 52°C oven temperature for all five samples, the individual T.C.s are well below +/- 2ppm/K.
It follows that the attenuation factors for R3 and the most critical R4, R5 are at least 10 times greater than in the datasheet. These resistors now can have much relaxed stability specifications, only R1 and R2 should be PWW or BMF types.
I also measured the low frequency noise 0.1 .. 10Hz, like branadic and Castorp.
I use an LNA from Andreas, and always made "typical" measurements of 10sec long samples, so to have a direct comparison to my LTZ1000s, but also to the measurements done by Andreas.
100s measurement time or averaging give no essentially different results.
Other LNAs will give different absolute noise figures, so these results are not directly comparable .
The noise of the ADRs RefAmp evidently is about 50% of the LTZ1000, which several others also have confirmed.
ADR #1 would be the most quiet sample, but it suffers from  :popcorn: noise, or frequent dips, but with a very moderate amplitude of maybe 400nVpp.
The ADA4522 buffer contributes a lot to the overall noise, so I have to investigate if it's only this specific unit, or what can be done to improve this.
I will add the noise figures of these samples, when measured with my 3458A, maybe now I can see a pronounced difference.
As my 34465A showed much higher stability / lower noise when I replaced its LM399 reference by a LTZ1000 clone PCB, my idea is to replace the LTZ1000 in my 3458A with an ADR1000 (*).
Maybe its noise figures will also go down, best by a factor of 2.
Castorp found that in his DVM other components overwhelmed the noise budget, so let's see what happens.
I would like to remind that TiN some time ago made a similar experiment on one of his 3458A, by implementing 4 parallel LTZ references, which would also yield 50% noise of a single LTZ.
I'm not aware what his final result was.
Anyhow, I think this new device will require quite a big time slot during the Metrology Meeting 2021, and I'm excited to meet the volt-nuts community personally and we'll have some fruitful discussions and comparisons   ;)
Frank
(*) my 3458A already has a modified FW, with CAL? 2,5 = 6.50000V (7V reference lower limit).
Thanks to Poul-Henning Kamp for your advice how to correct the checksum.
Hello Frank,
in your post you mentioned that you modified your 3458a's firmware to accept the lower voltage value of the ADR.
Would you mind sharing the process? I like to try it as well...
BR,
Max
Offline Dr. Frank
« Reply #195 on: December 06, 2021, 05:05:26 pm »
Frank
Hello Frank,
BR,
Max
Hello Max,
It was not sufficient to only change the minimum limit for the CAL? 2,5 parameter, i.e. the minimum internal voltage reference, to 6.5000V, but it was also necessary to modify all those lower ACAL limits, which the lower reference voltage also affects.
This can be identified by comparing the auto-calibration constants of a 3458A with two different reference voltages, like a regular LTZ1000, and one with a different LTZ chip, or with an ADR1000 on board, and slightly faked calibration reference values.
TiN kindly helped me by testing my first version on one of his 3458As, and sending me dumps of all his calibration constants. With the ADR1000, he found out, that the ACAL procedure gave system errors, due to several  lower limit violations.
So the procedure is to search for the location of these additional, crucial constants CAL? XXX,5 inside the firmware (=> 13 EA), and then lower them by the approximate difference by the ADR1000 voltage, about 6.667V, and the lower limit of the LTZ1000, i.e. CAL? 2,5 = 7.000V, i.e. each new minimum value has be calculated to about -6% each.
The displayed minimum values CAL? XXX,5 are not represented directly in the firmware, instead they are a calculated product of several internal cal parameters, so that's a bit tricky to find them in the hex file.
For that you need an appropriate hex editor, which is also capable of calculating the DOUBLE numbers back and forth, by taking care for the correct endianess and correct byte-order inside the single or 6 fold EPROMs, so that you can now determine the new lower limits.
I've done that successfully on my 2nd firmware version, which I also run in my own 3458A.
TiN was again so kind to test it and he confirmed that now the basic calibration processes, CAL 0, CAL 10 and CAL 10000 are successful, as well as the ACAL ALL procedure.
Next step would be to optimize the ADR1000 reference board for lowest noise performance, maybe also using less noisy OpAmps inside the 3458A, and then check against another external ADR1000 reference, or another 4x averaged zener reference, if the 3458A then shows a lower noise performance as well.
I don't know, how much further hints or spoilers you want to receive, just let me know.
Frank
Offline MaxTesla
« Reply #196 on: December 06, 2021, 05:29:20 pm »
Frank
Hello Frank,
BR,
Max
Hello Max,
Frank
Hey Frank,
thanks for your detailed answer. This sounds like a lot of changes to the firmware (especially for my limit skillset regarding this kind of stuff). I will definitly have a look into it.
THB I was more on the look for a higher stability spec then lower noise, but for that I first need to finish some boards to test and select some ADRs.
For me it would seem a lot simpler to redesign the A9 board with a small (stable) amp which boosts the ADR output just enough to get in the range of working (i.e. 7 Volts if i remember it correctly).
Obviously noise is concerne in this regard, but this should be manageable (at least from my perspective today) with the right selcetion of ADR/Circuit and would be a "upgrade" path for others without touching any firmware.
To compare agiant I would use another ADR reference, as I am currently designing my take on a 10V reference.
Opions on this "plan" are obviously welcome :)
BR,
Max
Offline Kleinstein
« Reply #197 on: December 06, 2021, 06:19:08 pm »
Adding some 6% of gain the reference circuit would still require very stable resistors, not to increase the drift. Added noise would be the least problem.
It it is just about getting a long term stable reference the old type LTZ1000 reference may be the better choice, maybe with a reduced temperature.
So far the limited reports on the ADR1000 are showing very low noise, but also still some drift. There is still some hope to improve on the drift with burn in or annealing, but I would not absolutely count on this. The advantage of the ADR1000 is more that it can get away with lesser grade resistors and still get good stability. With really good resistors the LTZ1000 may still be more stable.
  • Sounds like noise
« Reply #198 on: December 06, 2021, 07:40:45 pm »
Speaking of drift, I reached 3000 h but I'm not happy with the result. Nevertheless, I guess it is good advice to share such results anyway.
-branadic-
The following users thanked this post: chickenHeadKnob, Andreas, doktor pyta
Offline MaxTesla
« Reply #199 on: December 06, 2021, 08:04:20 pm »
Yes you might be right. But with the limited dataset I still plane on experimenting. I have treated myself with a set of VHP101 for a LTZ (70k,120,13k/1k) and the ADR (acc to the Datasheet). Plane is to put the LTZ in the mix with the ADR in with the same layout and test them side by side. This will take some time though as my ADR resistors are scheduled for march…goal is to compare drift and tempco. This will be done against my well aged 732a so may in a year from now I will have somewhat meaningful results, after that I will have a look at the A9 board project again I think.
Offline MaxTesla
« Reply #200 on: December 06, 2021, 08:07:52 pm »
-branadic-
Wow nice work :)
What temperature setpoint do you use on your ADRs?
Offline Dr. Frank
« Reply #201 on: December 06, 2021, 10:04:44 pm »
Hello Max,
unfortunately, reading every speculation from him, always the complement seems to be correct. No own experience from his own experiments, as far as one can deduce.
A divider being a few % off from "1" is extremely stable.. simply calculate the total differential! A 6% amplification should attenuate all drifts of the implemented resistors by about these 6%. I.e. a resistors T.C. of 2ppm/K will be attenuated to 0.1ppm/K of amplification drift. 5ppm/yr. => 0.3ppm/yr.
Added noise by an OpAmp IS the most severe problem for the ADR1000, as I have demonstrated upwards in this thread, in conjunction with the ADA4522 @ x1 amplification!! Any additional buffer amp might spoil the superior noise performance of the ADR1000.  Therefore the easy replacement of the FW is the best option.. if the 3458A circuit really benefits from the low noise ADR.
The long-term stability of the ADR1000 MIGHT initially be worse compared to the LTZ1000. Only branadic has presented so far a single drift result @10V.
My single sample experiment, measured directly at the raw ADR1000 output, w/o to 10V amplification, shows no remarkable drift so far after 500h. That's also no reliable result at all. Others in the background reported that the ADR needs a longer burn-in than the LTZ. So the speculation which chip is more stable, is at the moment completely baseless and speculative.
Spoiler alert: CERN obviously uses the ADR based circuit as an improved 10mA precision source!
Lower oven-temperature is always a good way to improve timely stability, see P.J. Spreadbury publication from 1990. Chose about 55 .. 60°C in any case. Don't use the ADRs datasheet values for best performance of the ADR. See hints in my post above.
The T.C. of the circuit for both chips have to be trimmed to near zero. There's a lot of potential on the 3458A's LTZ1000A board. Mine has about 0.2ppm/K, half of the overall temperature drift of my instrument.
Use that infamous 200k resistor for T.C. trimming. (or not at all).
The LTZ1000A inside my 3458A drifted -2ppm over about 9 years only, due to the reduced oven temperature, but mainly because the 3458A is running only sparsely. Again, see Spreadbury.
My ADR #3 runs at about 55°C, just at its determined zero TC point of the Reference Amplifier.  Simple thing.
And no, the LTZ1000 does not necessarily need more stable resistors. The timely stability of the circuit is mainly determined by the chip itself, the T.C. drift can be trimmed to near zero. Exactly the same goes for the ADR. So, please, no Vodoo stuff here, again. The ADR has much better noise performance, definitely, and that's the main reason to go for it.
The 3458A is no voltage reference, by all means. Therefore, I might exchange its very stable LTZ1000A by the ADR1000 some day, using the very same PCB with different resistors, and adding Andreas capacitors, for higher noise immunity.
Frank
« Last Edit: December 06, 2021, 10:08:07 pm by Dr. Frank »
Offline MaxTesla
« Reply #202 on: December 07, 2021, 11:31:10 am »
Hello Max,
Frank
Hello Frank,
thanks a lot for the insight and the paper, didn't knew that one.
This is highly interesting, out of my guts i would have suspected stabiltiy to be more of an topic, but as the amplifer would be one rougly 1:1.06 you are right of course.
My idea was to implement the amp with a VHD200, so tempco/stability impact is very small.
Regarding the noise and your buffer setup: This suppries me a lot tbh, I mean the ADA4522 with the right kind of input lowpass should not contribute that much, with the assumption that the datasheet noise values are correct, but as you discribed your samples are much lower noise than the spec. Will be interresting to see what my samples do, I have enought of them to get a little bit more "meat at the bone" so to speak.
Thanks for the tipp with the temp setpoint, I don`t have the appropriate means of measuring it myself (yet). So revised plan for tempco/stability test is the following: 6 ADRs in total, 3 with 12K/1K divider (~60°C) and 3 with 11k6/1k (~53°C) setpoint to get a span, each mounted in a housing with the same board, no preconditioning and just accquiring data form the start. This should give me/us a somewhat good idea what is going on, as three is the statistical minimum. Other resitors are 61.9k and 95.3Ohms. Monitoring ambienten contiditions of course to see any kind of correlation during normal weather shifts, all done automaticly every day with my 1281 and 7168 scanner. One channel reserved for a short, one for the 732a. If all goes well we can see a difference in drift with setpoint temperature in two points.
Thats at least the plan so far :)
After the results comes the tuning...
BR,
Max
Offline Dr. Frank
« Reply #203 on: December 07, 2021, 01:29:39 pm »
Hello Frank,
This is highly interesting, out of my guts i would have suspected stability to be more of an topic, but as the amplifier would be one roughly 1:1.06 you are right of course.
Well, a VHD200 is technical overkill, expensive, and has a long leadtime.
Use two appropriate cheaper single resistors from stock, WW or MBF ones, like 6k8 + 470 Ohm, or so.
Regarding the noise and your buffer setup: This surprises me a lot tbh, I mean the ADA4522 with the right kind of input lowpass should not contribute that much, with the assumption that the datasheet noise values are correct, but as you described your samples are much lower noise than the spec. Will be interesting to see what my samples do, I have enough of them to get a little bit more "meat at the bone" so to speak.
I was also surprised, that the MAX4522 (of course with input LPF) adds so much noise. The ADA4522 has typically 115nV pp noise only, which should not affect the ADR1000 noise performance.
This combined noise level is about the same as an LTZ1000 w/o buffer. Maybe I have to look for a better sample of the ADA4522.
It's not only my 5 samples, which have 50% noise of that of the LTZ1000. Other volt-nuts also reported the same.
Inside the 3458A, both LT1001 on A5 board, U160 and U165, which generate the +/- 12V reference for the A/D will definitely spoil the game.
These have 300..600nV pp noise, i.e. at the same level what I've measured for the ADR1000.
These might have to be replaced also.
Probably you already HAVE the possibility to measure the oven setpoint by yourself. A faster bench DMM is sufficient, to catch the initial diode voltage @ R.T., with oven disabled, and then extrapolate by using -2.2m/K for the temperature dependency. A 34401A is sufficent for that, and your 1281 will definitely do the job.
You may first test, at which collector current of the Reference Amplifier you get zero T.C. for the (not ovenized) RefAmp at your intended oven temperature.
So revised plan for tempco/stability test is the following: 6 ADRs in total, 3 with 12K/1K divider (~60°C) and 3 with 11k6/1k (~53°C) setpoint to get a span, each mounted in a housing with the same board, no preconditioning and just acquiring data form the start. This should give me/us a somewhat good idea what is going on, as three is the statistical minimum. Other resistors are 61.9k and 95.3Ohms. Monitoring ambient conditions of course to see any kind of correlation during normal weather shifts, all done automatically every day with my 1281 and 7168 scanner. One channel reserved for a short, one for the 732a. If all goes well we can see a difference in drift with setpoint temperature in two points.
That's at least the plan so far :)
BR,
Max
Setpoint of the oven can also be trimmed by changing the collector resistor of the temperature sensing transistor inside the LTZ/ADR circuit.
So you may always use a  12k/1k divider and select the collector resistor instead. The stability requirement on latter is relaxed, especially if you run your circuit exactly at the sweet spot, i.e. at zero T.C.
I do not recommend 11k6/1k, that may be too low a temperature.
Select the zener resistor for a bit more than 5mA @ oven temperature, and select both collector resistors as described above.
Your monitoring capabilities sound pretty good.
Frank
« Last Edit: December 07, 2021, 01:56:24 pm by Dr. Frank »
Offline Kleinstein
« Reply #204 on: December 07, 2021, 03:09:02 pm »
The resistors for the 6.6 to 7 V gain would still be the most critical one when it comes to drift. They enter with a 6% factor, or a bout a facto 17 attenuation. The resistors at the ADR have a larger attenuation, like 200 times and higher.
For the gain stage one may just get away with an LT5400-7 as a maybe better available resistor.
Az OPs can react to the input impedance and details in the supply decoupling. So there is a potenetial for them to show more than the datasheet noise. They are also a possible source for EMI so that other parts than give the extra voltage.
As a lower noise OP at the ADC I would consider the OPA202 as a possible candidate.
Offline e61_phil
Re: Lowest drift, lowest noise voltage reference (ADR1000AHZ)
« Reply #205 on: December 07, 2021, 05:31:17 pm »
It is true that the LTZ1000 CAN be trimmed to ZTC, but without trimming the ADR1000 is much better in terms of TC. In one project exactly that was the reason to move to the ADR1000.
Offline MaxTesla
« Reply #206 on: December 07, 2021, 06:34:29 pm »
Yeah the lead time on these VPG resistors is horrible but ok, metrology is no hobby for the hasty ones, trying to improve my patience here :D
I am gathering test gear for this project since nearly two years now (orignally planned with the LTZ of course)...
Thanks for the tipp, didn't thought on that one...in my head I needed a external source (of heat) to do it, not that I don't have a 2510, but I wasn't ready to put the efford in.
So I will think about a proper setup to do it (waiting on some new PCBs anyways which one of them will be modified for that).
Yes I am aware of the trimming, however i would like to get the drift characteristics first (and see how the tempco behaves) and then tweek the refs. Indeed i was wondering about the low setpoint, but I think its worth a shot.
Regarding the exchange of U160 and U165: I assume (not checked) there supply is bipolar +-15V, so a LT1128 would work fine and it has very very low noise...considered it instead of the LT1013 but the bipolar power supply needed to get to (or near) 0V keeped my from implementing it. Didn`t seem necessery at the time, but after this discussion about the really low noise floor this is worth a rethink.
BR,
Max
« Last Edit: December 07, 2021, 06:47:49 pm by MaxTesla »
Offline Kleinstein
« Reply #207 on: December 07, 2021, 08:25:41 pm »
There should be no absolute need to get really zero TC for the unheated reference. If possible it would help to get the TC without the heater to below 20 ppm/K. Something like the 10 ppm/K range should be perfectly OK with a reasonable stable temperature. A slightly higher current for the zener and/or transistor could do the job.
The LT1128 is low noise for a very low impedance ( e.g. 100 Ohms) source, but not with some 10 K like in the 3458 reference amplification. For the ADR1000/LTZ1000 reference the OPs see some 50-70 K and thus even more impedance. The OP is also behind the gain of the transistor and thus less critical. So the LT1013 should be good enough there.
So if at all the LT1128 may have a small chance to work in the 6.6 to 7 V gain stage, but even there I would prefer a lower noise current OP. The chopper artifacts from a chopper OP could be an issue.
The ADC in the 3458 should react a little to reference noise in the 150 kHz range. The noise in this range may depend on the compensation (including OP speed) around the reference. The curve in the DS looks good for the ADR1000, though not too sure which OP is used and there may be limit of the test system.
Trying to get even lower noise from the ADC in the 3458 is a different topic. Only U160 would see the same frequency band as the reference. For U165 and U110 the chopper action of the AZ mode is effective and thus a relevant frequency of more like 2.5 Hz with an often relatively low BW and some extra gain (around 2.25 for the more critcal U110, should be 1.25 for U165).  For the actual use reference noise at lower frequencies than 0.1 Hz may be relevant. The 0.1 to 10 Hz range is more like a range still easy to measure. It would be mainly U160 to worry about for low frequency noise.
Offline Andreas
« Reply #208 on: December 12, 2021, 04:31:30 pm »
Hello,
working on alternatives to the statistical divider with a LTC1043 cirquit.
The main amplification is done by a 3 to 2 (factor 1.5) charge pump.
The remaining part can be easily done with standard resistors.
Either up or in a additional buffer stage with a voltage divider down.
Temperature compensation is done with a diode cirquit.
The typical +220 ppb/K can be compensated with the following cirquit.
R8+R9 are meant to be a 50K trim pot where the T.C. can be trimmed from around -300 to +300 ppb/K.
Unfortunately there is a unwanted offset along with the T.C. trimming where I am looking for a solution.
Any ideas?
with best regards
Andreas
Offline Kleinstein
« Reply #209 on: December 12, 2021, 06:29:38 pm »
The TC adjustment would effect the absolute value in most cases. At best one can eliminate the effect at 1 temperature, but it can be made better than in the circuit shown. Both sides of the trimmer should see about the same voltage at a reasonable temperature.
One could get such a setup with a bridge with 2 diodes and 2 resistors on opposite sides to replace the 2 diodes in series. The main parts is shown in the attachment.
It looks like there should be a little more filtering for the switched capacitor part.
The final circuit should also have supply filtering for the AZ op and LTC1043 to keep the supply clean.
Offline Andreas
« Reply #210 on: December 19, 2021, 01:39:32 pm »
Hello,
current results of ongoing burn in of ADR1000A#02 in burn in socket.
Procedure is temperature cycling from 110 deg C to 130 deg C within 2 Hours.
so 10 minutes rising with 2 deg C/min 50 minutes plateau and 10 minutes falling with 2 deg C/min.
(see previous post).
At around 200 hours I thougt output voltage would stabilize but obviously this is not the case.
Now at nearly 400 hours I am somewhat unshure:
- will there be a stabiisation point
- or will the device be destroyed (because 130 deg C is too high?)
- or is the temperature still too low (as on MM the temperature of 150 deg C was mentioned)
with best regards
Andreas
The following users thanked this post: e61_phil
Offline branadicTopic starter
  • Sounds like noise
« Reply #211 on: December 19, 2021, 02:03:48 pm »
Hi Andreas,
I guess what you look at is an exponential drift curve. Multiple times I though myself the reference would now start to stabilize, but then drift moved on and it was just a plateau in the drift curve for whatever reason - such as variation of the meters reference and t.c..
Meanwhile I'm confident to predict the drift by an exponential function, as can be seen in the image below.
-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
Offline Kleinstein
« Reply #212 on: December 19, 2021, 02:54:20 pm »
If there is a stabilizing point is a good question. There is a chance that there is some process that us not stabilizing but contineously going one direction - kind of aging to wards destruction. One part seems to have stabilized after some 200 hours.
The parts still showing slow drift at some 120 C shoud be quite a bit (e.g. a factor of around 100, maybe more) slower when at a more moderate temperature, like 50 C. So chances are the part should be stable enough by now. So I would not worry too much about the still visible drift.
Another point is that the burn in circuit is more sensitive to the resistors (and maybe the MOSFET). The resistors are not the most stable ones. So the observed drift could also be in parts from the resistors in the burn in circuit and maybe the temperature control part.
It is hard to tell if the 130 C are too low and which annealing process is best. Usually the first and obvious part is to really finish the chemical reaction in the die attach glue. Here the 130 C should be high enough, especially with so much time. For a glassy epoxy I than see 2 possible routes:
A) Get a dense stucture in procedure like this:
 1) anneal near or above Tg to finish the chemical reaction
 2) anneal somewhat (e.g. 30-50K) below Tg to get a dense amorphous structure. This could also be with a slow coold-down, maybe in steps.
B) Get a more open glue stucture to allow better stress ralaxation in procedure like this:
 1) anneal near or above Tg to finish the chemical reaction , so more like short time at a rather high temperature
 2) cool down to room temperature fast, to suppress the densification , so keep the epoxy soft by freezing a more open structure.
 3) wait at room temperature or maybe even slightly below, to allow relaxation of thermal mismatch stress.
   (a lower temperature gives more thermal stress and may this way give faster creep, despite of lower temperture)
 4) anneal at an intermediate temperature (e.g. 70 C) to get the structural relaxation in the glue, but not much new thermal stress.
For a more definitive answer one would need to run the reference in a more normal circuit and with better resistors. To get a shorter time scale, this may be with a slightly elevated temperature (e.g. 20 K higher than normally planed for).  There can also be some hystereseis type effect and thus a new relatively fast part when going to a lower temperature. So a test at let's say 60 C may again show a fast initial part. So this test would again need some time.
Offline Andreas
« Reply #213 on: December 19, 2021, 04:12:45 pm »
It looks like that:
When I make the x-Axis as sqrt(kHrs) and the y-Axis as ppm instead of mV then I get (nearly) a linear curve.
Regression is -62.4 ppm/sqrt(kHr) at 120 deg C.
So with Arrhenius and factor 2 every 10 deg C and 60-70 deg C (including self heating) we would get something like 1-2 ppm/kHr which is still too high for a voltage standard. I want to get down to better than 1-2 ppm/year.
with best regards
Andreas
  • Sounds like noise
« Reply #214 on: December 19, 2021, 04:27:11 pm »
Then run the aging process longer. As you can see with my non-pre-aged reference it takes some time. Fit your drift curve with an exponential curve, predict the drift for the future - say for 8760 h - and monitor, if your prediction changes/stabilizes over time. Once the drift prediction stabilized you will see how long it will take to run this procedure.
We can then compare how long it took your reference to stabilize with my one and see, if this process did anything good at all.  :-+
-branadic-
Offline Andreas
« Reply #215 on: December 19, 2021, 04:35:43 pm »
Hello Branadic,
I calculated roughly -2.3 ppm/kHr for your ageing curve at ~50-60 deg C which is around that what I theoretically would get if lowering the temperature.
Perhaps I will put my setpoint some 5 deg C higher to 115-135 deg C (the maximum I can use with my curent setup) to prove whether the drift increases.
with best regards
Andreas
Offline Kleinstein
« Reply #216 on: December 19, 2021, 06:04:16 pm »
The drift in the burn in circuit can be more than just from the chip. This can also include other parts, like the resistors.
50-60 C would be usually the die temperature, including self heating. Depending on how the temperature is measured during burn in there may be some self heating to add there.
The doubling every 10 K rule is only a first approximation. The actual number can vary depending on the process. The difference can be especially large when it comes to epoxy close to the glass temperature. There is the Arrenius factor and an additional change in structure.
If an exponential fit is suitable depends.  Aging processes can also follow different laws like square root or stretched exponentials or teh suposition of several effects.
The relaxation of stress can be nonlinear as plastic processes often are. Pure viscose creep is more like the exception.
It probably takes more data to really decide if an exponential is suitable. An exponential fit normally corresponds to a single relatively simple process.
Especially the initial phase has usually more contributions, some faster will decay full to later only leave the slowest visible process and other processes that are too slow to see.  An exponential decay is more like an optimistic case.
I don't think the drift measured for the simple burn in circuit is good to estimate the later drift at some 60 C. The circuit is different (more drifting components) and the temperature is very different, and thus quite some extrapolation.
Offline Andreas
« Reply #217 on: December 19, 2021, 08:27:58 pm »
Hello,
since the statistical voltage dividers are in plastic packages I also examined influence of humidity to the voltage ratios.
Setup is as follows.
ADR1000A#01 in my cooling box. Environment is kept to 27 deg C on the heat spreader. (so a little more on the ADR PCB).
Humidity is changed either near 30% rH or up to near 80% rH.
Divider Ratio is measured with the ratio functionality of my HP34401A (100 NPLC).
Results:
The output ratio (nominal 1500 mV/V = 1R/2R built from 6 resistors) changes up to 0.3 ppm for a near 50% rH change within 2 weeks.
The setpoint ratio (nominal 12500 mV/V = 11.5R/1R built from 8 resistors) changes up to 2.2 ppm for the near 50% rH change.
The second ratio drift is larger maybe because the ratio needs several "single" resistors/dividers and so is not really "statistical".
A ratio with 12:1 with 6 resistors in series and 2 resistors in parallel would have the chance to be better in this regard.
And it seems that 2 weeks settling time is still not enough for final settling.
with best regards
Andreas
The following users thanked this post: doktor pyta, miro123
Offline Kleinstein
« Reply #218 on: December 19, 2021, 08:50:48 pm »
A 2.2 ppm change in the temperature set point would not be so bad for the ADR1000, as the drift is attenuated quite a bit (e.g. like 200 times - depends on the unheated TC).
The 0.3 ppm drift of the output gain would be the larger effect. This number is still not so bad, as 50% change in RH is already quite some change.
The actual RH change at the resistors would be smaller, as a higher temperature would reduce the RH (about a factor 0.7 for 5 K higher). It still makes sense to use the environmental RH, as the a similar situation would be common in applications.
80% RH is quite humide and may reach the range where surface film may form, adding extra surface leakage currents. This effect may be highly nonlinear, with a material dependent critical RH for the film to form. This part would be relatively fast and not take weeks, more like minutes.
Offline Andreas
« Reply #219 on: December 29, 2021, 08:37:05 am »
Hello,
news from the burn in cycling of ADR1000A#02
After  ~500 hrs I actually increased the temperature cycling from 110-130 deg C to 115-135 deg C.
But I recognized that the overtemperature switch was actuating at the end of the temperature rising
so I finally adjusted the cycle to 114-134 deg C.
Expectation was that with ~5 deg C increase of the average temperature the ageing would increase by ~factor 1.4
Or somewhat lower (1.32) for ~4 deg C.
The first new measurement trendline (in blue beginning at sqrt(0.5 kHrs) = 0.7) shows something like 200 ppm/sqrt(kHr).
Which is around factor 3 above the 4 deg lower temperature (was 62 ppm/sqrt(kHr)).
So something is different above 130 deg C.
Unfortunately I do not have a good noise measurement "before" with the ADR#02.
Fortunately I remembered that Jason (cellularmitosis) has done a good job (thank you for that) in publishing several LTZ1000 PCBs.
So overcome this for ADR#03 I built a noise measurement cirquit from one of Jasons PX PCBs.
with best regards
Andreas
Offline miro123
« Reply #220 on: December 29, 2021, 10:13:39 am »
Hello Adreas,
Thanks for sharing. I have two questions related to your last two posts.
1. How do you control the humidity by resistor test? - I want to do similar test. My goal is to create resistor divider model. I want to find certain pattern between humidity, temperature and resistor divider ratio. I expect to determine Humidity behavior like - humidity coefficient. humidity time constant & hysteresis. I hope that  a. Observed behavior is repeatable and model can by build b. Expect simple polynom fits the model.
2. At your last post mentioned T 114...134C Is this environmental temperature or Die / Transistor base emitter temperature? I'm afraid that you are going far beyond the specs. Epoxy is just one component in the whole picture. e.g. How SiO2 react to such temperatures, and what about chip multilayered structure, are you casing permanent damage with it? That are all questions that i dont have answer
« Last Edit: December 29, 2021, 11:09:30 am by miro123 »
Offline Kleinstein
« Reply #221 on: December 29, 2021, 10:47:44 am »
The higher temperature speeds up the termally activated porcesses. However there is more change in getting a different equilibrium state.
For the epoxy part, relatively close to or above the glass temperature the structure can reach an equilirium, with a more open structure at a higher temperature. The more open structure also makes things faster, in addition to thermal activation.
With a different equilibrium state (especially for stress) the question is a bit if the aging at the high temperatre actually helps with the later performance at a considerably lower temperature. At least the higher temperature aging can not replace all the burn in time needed also at the final temperature.
There are still hysteresis effects, that than start a new settling / ageing.
Offline Andreas
« Reply #222 on: December 29, 2021, 11:07:55 am »
1. Control of the humidity is done with a 1 kg Dry pack (Silica gel) for the ~35 % rH.
    For the ~80% rH I used acryl polymere crystals like these:
    https://www.amazon.de/GERMANUS-Humidor-Befeuchter-Kristalle-AcrylPolymere/dp/B0796GG83B
    All is within my cooling box. (isolated and humidity tight).
    Essential is a constant environment temperature (here ~27 deg C)
    PWM controlled by heater foils on a aluminium sheet as heat spreader.
2. it is the temperature of the PT1000 sensor which is bound (with strong copper wire) to one of the heater resistors.
   (unfortunately I did no photo). But perhaps you can imagine the setup from the linked pics.
https://www.eevblog.com/forum/metrology/lowest-drift-lowest-noise-voltage-reference/msg3788051/#msg3788051
https://www.eevblog.com/forum/metrology/lowest-drift-lowest-noise-voltage-reference/msg3758498/#msg3758498
   Connection is via the heat spreader (aluminium bar) to the housing of the ADR.
   All is cotton cloth isolated within the thermo mug.
   So the measured temperature is nearly case temperature.
   With self heating of the ADR I think that the chip temperature is ~3 degrees warmer.
with best regards
Andreas
Offline RikV
« Reply #223 on: December 30, 2021, 07:50:40 pm »
All this is very nice, but I would like to play along! Where and how can I get ADR100's? Are they reserved for VIP's? No sign of life on AD's site.
Offline RandallMcRee
« Reply #224 on: December 30, 2021, 10:41:26 pm »
Somewhere in this thread (or was it a different one?) it is explained that you need to type "ADR1000AHZ" directly into the shopping cart search bar. When you do this it pops up with an order for 100, you can then update it to a lesser amount. Not sure, if this will be honored of course since I dare not check out at this time!
Offline Grandchuck
« Reply #225 on: December 30, 2021, 10:49:08 pm »
Tried it.
Some info here:
https://xdevs.com/article/adr1000order/
I guess you need an account.
« Last Edit: December 30, 2021, 10:58:25 pm by Grandchuck »
The following users thanked this post: bob.cmhk
« Reply #226 on: December 30, 2021, 10:57:48 pm »
Tried it.
 :palm:
No, that is the normal search bar, not the shopping cart search bar. The shopping cart search bar is visible in my screenshot; its near the bottom of the page.
Yes, thanks for the link: it does describe the correct procedure. I believe the backorder time to be accurate....nothing until June.
Edit: it allowed me to place an order. Will update y'all in June.  :=\
Edit again: Order confirmation says "In stock - est delivery Jan. 10". So yay!
« Last Edit: December 30, 2021, 11:16:15 pm by RandallMcRee »
Offline Grandchuck
« Reply #227 on: December 30, 2021, 10:59:17 pm »
They really want to sell these? :-//
Offline Andreas
« Reply #228 on: January 04, 2022, 06:47:19 pm »
Hello,
after playing with the suggestion of Kleinstein here the results:
- the T.C. trimming is now symmetrically to a center value near 25 deg C.
- I had to fine trim the suggested 22K resistors which where already a good estimate to 18K4 to achieve this.
- what I had not expected is that the sensitivity is about halved. So I had to reduce R10 from 1 Meg to 470K.
with best regards
Andreas
Offline Andreas
« Reply #229 on: January 06, 2022, 11:24:50 am »
Hello,
news from my ADR1000A#01 (which is built on one of branadics PCBs. See first post).
I now changed the Zero-TC-temperature from around 48 deg C to 55.4 deg. (using method a)
(refer to https://www.eevblog.com/forum/metrology/lowest-drift-lowest-noise-voltage-reference/msg3810545/#msg3810545)
The plan was a zero TC temperature between 52-55 deg C for up to 40 deg C environment temperature.
But since I had no suitable resistors below 15 ppm/K for trimming I finally got 55.4 deg C.
Also the setpoint temperature was changed to the corresponding 524.7 mV at the heater sensing transistor. (got 524.5 mV actually)
With this modifications another T.C. test showed +166 ppb/K temperature slope for the zener voltage.
Zener voltage is now increased by ~2.4 mV.
The next step will be trimming the T.C. by around -178 ppb/K (+166ppb for the zener and half of the 10V output stage T.C. of +20-24 ppb/K).
with best regards
Andreas
Offline Andreas
« Reply #230 on: January 10, 2022, 08:55:56 pm »
Hello,
ADR1000A#01 after T.C. adjustment:
I got near linear T.C. values
6.6V Zener voltage ~     -6 ppb/K
10V Output voltage ~ +20 ppb/K
the 10V output voltage is still too low (~9993 mV) and needs further adjustment (after some ageing drift measurements).
with best regards
Andreas
Offline Andreas
« Reply #231 on: January 15, 2022, 12:17:36 pm »
Update on Burn-In of ADR#02 after 1kHr total.
compare with:
https://www.eevblog.com/forum/metrology/lowest-drift-lowest-noise-voltage-reference/msg3903542/#msg3903542
just after last shown measurement the ageing changed the slope.
At about the same time the output voltage got a much larger hysteresis, which partly explains the larger stray of the drift values.
Slope is now around -44 ppm/sqrt(kHr) (from initial -62 ppm/kHr over -200 ppm/kHr)
next step is to go down again to 110-130 deg C cycling to see if ageing also here has decreased.
with best regards
Andreas
The following users thanked this post: TiN, branadic, Mickle T., doktor pyta, alm
  • Sounds like noise
« Reply #232 on: January 16, 2022, 08:41:41 am »
Thanks for keeping us updated Andreas.
I almost hit 4000 h on my sample, with the setup untouched for the whole time. Attached is an updated graph.
-branadic-
The following users thanked this post: Andreas, doktor pyta, KT88
Offline Andreas
« Reply #233 on: January 16, 2022, 12:05:41 pm »
Hello branadic,
it would be good if we get the exact receipe of burn in for the ADR1000 at Analog Devices.
Is it really the 1 week at  constant 150 deg C under bias?
I now could tweak my burn in cirquit for slightly higher temperatures as I just got a
different overtemperature switch with ~170 deg C.
https://www.conrad.de/de/p/thermorex-tk24-t01-mg01-oe170-s160-bimetallschalter-250-v-16-a-oeffnungstemperatur-5-c-170-c-schliess-temperatur-160-1678193.html
Btw. I have a 2nd temperature sensor measuring the PCB-temperature which is around 30 deg C lower than the heat spreader temperature.
So I can use "normal FR4" for the burn in PCB.
with best regards
Andreas
  • Sounds like noise
« Reply #234 on: January 16, 2022, 12:29:50 pm »
That's at least what Eric Modica said at MM2021.
Question: At which temperature did you make the burn-in?
I believe it's a 150 °C.
Question: So simply powering the zener and have the ambient at 150°C and then - what was it? - 168 h or something like that?
Yeah, you know and we haven't tested the bounce of that, we basically just copied the bake and burn process of the LTZ1000, which is a 168h both bake and burn. We have at least small sample size evidence that you maybe able to get away with just doing an unpowered bake and that's good enough to disipate the charge, but I have to say I haven't looked at it very closely.
Question: Simply elevated temperature or cycling during burn-in? Because, some rumor says cycling around glass transition temperature might improve the aging.
Our burn-in process doesn't utilize any cycling.
-branadic-
« Last Edit: January 16, 2022, 12:33:09 pm by branadic »
Offline Andreas
« Reply #235 on: January 16, 2022, 08:26:07 pm »
So to summarize.
you have around -4ppm in 4000 hours / -2 ppm in 1000 hours
so -2 ppm/sqrt(khr) without burn in and @55 deg C heater temperature (assuming you use zero T.C. temperature as set point).
Extrapolating for the next year (13 kHrs in total) we will have -7.2 ppm total drift or -3.2 ppm for the next year.
The data sheet value of 0.5ppm/year after the first 3000 hours with a setpoint temperature of 75 deg C seems to be somewhat optimistic.
There is no burn in mentioned in data sheet for the ageing drift.
When I take my -44 ppm/kHr now at 124 deg C average for ADR#02 and translate it to 55 deg C setpoint temperature.
(factor 2^7 = 128 when doubling all 10 deg C)  = -0.34 ppm/kHr at 55 deg C.
This gives still -1 ppm/year.
with best regards
Andreas
The following users thanked this post: TUMEMBER
  • Sounds like noise
« Reply #236 on: January 16, 2022, 08:54:31 pm »
I guess you have to think different. The datasheet states 0.5 ppm/year after 3000 h, not sure if this is with or without burn in, but running the oven at 75°C. It excludes the first 3000 h of stabilizing, that's at least how I interprete things.
So going from here the drift of my sample was far less than 1ppm within the last 1000 hours and from here we need to observe how the drift goes on for the next 7760 h (~11 months). And yes, I'm running my oven at 55°C which is the zero t.c. point of this sample. This means, stabilizing will probably take a bit longer than 3000 h @ 75°C oven temperature, but lets assume it reached its stable point already. We will know more after the next 1000 h, I guess.
-branadic-
« Last Edit: January 16, 2022, 09:07:46 pm by branadic »
Offline Kleinstein
« Reply #237 on: January 16, 2022, 09:20:16 pm »
The doubling every 10 K is only a rough approximation. There are some effects (especially slow ones) with a stronger temperature dependence.
Similar for the drift over time there are several possible functions: exponential for simple relaxation processes and square root time for more random walk type processes (or superposition of many uncorrelated random effects) are just 2 examples. Another possible form would be a stretched exponential. So extrapolation is somewhat tricky.
For the high temperature part, I see mainly 3 effects to happen: fully finish the chemicial curing of die attach and possible stabilization (e.g. build up of a surface film, surface "reconstruction" ) on the chip. There may be a tiny bit of diffusion at the contacts, but hardly in the silicon itself.
The stress between the die and case should relax as the die attach would be relatively soft.
A later step would than include relaxation of thermal stress from the different thermal expansion of the silicon and the metal case. This would be near the final use temerature as the stress only comes back on cooling. The die attach can also have some structural relaxation when well below a glass temperature. The 150 C are more like too high for this, more like near the glass temperature where the structure is more like in a local equilibrium.
If I anderstand right the data for the high temperature anneal are for the simplified circuit, and this may show additional drift from the resistors and the FET.
The exponential fit to the data from brandic should give around 0.5 ppm/kh (5 µV for 1000 h) after 3000-4000 hours. Part of the drift could also be from the 6.6 to 10 V stage.  So this would not be so far off from the DS numbers.
Offline Neo2199
« Reply #238 on: January 30, 2022, 04:23:01 am »
Hi all, I am planning on building my first voltage reference. I have hard time getting standard LTZ1000 but I managed to order one ADR1000 from Analog Devices site which should perform similarly. It is quite hard to figure out good resistor values and set the operating point right where good long term stability is. But based on the info in this and other similar threads I think I have puzzled everything together. My target is 60C, 5mA zener current at 100uA collector current. I came up with R4 12k, R5 1k, R2 61.9k, R3 61.9k, R1 100ohm. Did I get this right, or should I change something?
I am also planning on designing my PCB, but I cannot find too much information about layout recommendation. It is interesting to me to see so many differences in layout of KX LTZ1000, reference Analog Design board and HP3458A LTZ1000 board for example.
Offline Andreas
« Reply #239 on: January 30, 2022, 09:39:21 am »
Hello,
Each single device behaves different regarding ageing.
So you might want to have at least 3-4 stable references to sort out those with the largest ageing.
Whether ADR1000 behaves similar to the LTZ1000 still has to be proven in practice.
At least it seems that the ADR1000 needs longer time to settle than a typical LTZ1000.
The resistor values are those recommended from the data sheet so it should be ok to go with them.
Others try to set the temperature setpoint to the zero T.C. temperature of the ADR1000 to reduce influence from set poin divider resistor drift.
(this is not possible with LTZ1000 due to the higher zener voltage).
For layout it is essential to do star point wiring and  especially kelvin sensing of the zener voltage.
(no ground currents from heater or OP-Amps across the ground sensing line).
When you have to integrate the zener in a device like 3458A it might be partially easier to work with ground current compensation.
So the constraints (dual/single supply) and pre-ageing and selecting on device / module level may give different board layouts.
with best regards
Andreas
Offline Andreas
« Reply #240 on: January 30, 2022, 10:37:51 am »
Update on Burn-In of ADR#02 after ~1.2 kHr total.
compare with:
https://www.eevblog.com/forum/metrology/lowest-drift-lowest-noise-voltage-reference/msg3942472/#msg3942472
After last reporting I reverted back to 110-130 deg C cycling hoping that the ageing slope would flatten.
But it increased instead from -44 ppm/sqrt(kHr) to -84 ppm/sqrt(khr) calculated for 120 deg C.
I wanted further reduce the temperature cycling to 100-120 deg and further steps down to see wether there is some hysteresis.
But unfortunately the controlling PC stopped working so I have to stop the bun in heating for a while.
with best regards
Andreas
  • Sounds like noise