I installed a new AC fan moments ago. The process was simple because the Fluke 5440B has a plug that fits the AC fan terminals. There is no need to tear the entire instrument apart for this job.
Manufacturer: ebm-papst
Part number: 4530Z
Rotron Series: 4000Z
Fan Type: Axial AC Fan
Size: 120 x 120 x 38 mm
Voltage: 115 VAC (60 Hz)
Power: 12 W
Speed: 2000 RPM
Flow: 70.6 CFM (2 m³/min)
Bearing: Sintec
Shape: Square
Static Pressure: 0.070 in H2O (17.5 Pa)
Noise: 32 dBA
Instructions [maximum of 20 minutes to perform]:
1) De-energize the unit, remove the cover (5 screws), use a digital multimeter to drain both rear capacitors by setting the meter to >= 1000 DCV and measuring zero volts.
Note: This is a precautionary step. No charge should exist after the unit has been unplugged.
2) Remove the fan filter (optional, 4 screws), the 2 nuts on the AC fan at the top, then remove the fan by pulling it forward from the top.
Tips: If you apply a little force downwards, it comes off without wearing out the threaded posts. There is no need to remove the unguarded power supply or capacitors for this step.
3) Disconnect the AC fan plug either before removing the fan or afterwards, depending on the accessibility of the plug.
4) Replace the AC fan with one that is equal or greater to the original performance (posted shown above).
Warning: When selecting a fan, do not use a fan with significant performance advantages than the original or the ovenized reference boards may over-compensate for the internal temperature change.
5) Reassemble the parts in the reverse order.
Comments:
- There was a gap (i.e., 1 - 2 mm) between the new fan and the chassis wall at the bottom. Perhaps the fan geometry needs to be more like the Rotron Whisper (WR2A1) to sit properly on the tab. I did try to push it closer, but the fan eventually slipped back to where it was. The gap should not make a significant difference on performance and maybe this is what it does with the original fan too -- I did not check.
- Over the next few days, I plan on testing whether or not the instrument can continuously source voltage on the 22 V range. Hopefully, this fan update prevents the 'power supply fault, check unguarded power supply' error message that my unit has been experiencing. The problem occurred even after I reapplied thermal paste to most of the critical components, replaced all the capacitors and carbon composite resistors, and adjusted the trimmers on the guarded power supply.