Thank you for your comments and inputs. Highly appreciated.
Not using scope though, I did measure DC over AC mains (220V) when running my Philips 1600 watts hair dryer.
Two DMMs, left one Fluke 287 showing DC over AC (you can see the bigger digits DC reading over smaller digits AC reading at the LCD display), while Fluke 87V at the right, was set at DC reading, and also Watt meter.
DC readings both were very similar.
I have seen your video (thank you for the nice video and test/demonstration) and I used it as a guideline for my measurements. As I can see it, have done the same as you. The only difference is that I have used a benchtop meter instead of a handheld meter. I have a handheld Fluke 177 as well; I will try to do the same measurements later today with the Fluke.
I have to ask...
- How did you measure your AC line?
- Are you aware that if you clip the crocodile of your probe on one the AC lines you have a 1/2 chance of blowing up your probe/scope/self?
Regarding how I have tested this, my test setup.
I have used a 3-outlet power strip with ON/OFF switch. The outlets are “polarized” and the power cords I have used are “polarized”, only one way to plug the cords in the power strip. I have cut a power cord in half and made flying leads for Live and Neutral. The power cord also has a two pole ON/OFF switch to turn ON/OFF the mains voltage for the flying leads.
First, I turned the power strip OFF. Plugged the multimeter, oscilloscope and test cable in the power strip. Next step was to connect the test probes to the flying leads; connecting the probe’s ground lead to Neutral and the probe tip to Live. For the multimeter, Live was connected to HI input and Neutral to LO input. Then, I turned the power strip ON (test cord switch still OFF), turned on the multimeter and oscilloscope, waited for them to startup and finishing their boot sequence. The last step was to turn the test cord ON so there would be live voltage at the flying leads.
The probe for the oscilloscope was a 100:1 probe.
The meter will clamp its input (likely asymmetrically) to to protect its circuits if you try to read a larger (240 VAC) on a low voltage DC range. You would need to use a ~ 300 VDC range to be sure the result is correct.
to prove your measurement is wrong, swap the meter leads -- you'll get the same DC result (not the -ve of it).
A transformer can't really generate a DC voltage offset, but certain loads (½ wave rectified) may cause a DC component of the load current which can appear as a DC voltage component across the resistance of the wiring.
I am not quite sure whether I understand your comment about the meter clamping the voltage. I have set the multimeter to use the 1000 V range. If I used the 100 V range the meter will only report Overload.